Electropolishing Validation of Orthopedic Parts - What OQ parameter range

#1
We’re a contract manufacturer and now we're doing a EP validation for different type of materials.
We have 440A, 420B, 17-4, 455, 304, 316.

We never get formal input from our customers. So my intension is to define two validation purpose, one is the surface roughness is reduced, and the other is corrosion resistance (ASTM F1089).

I understand some customer requires to test cytotoxicity. I don’t want do it, because I have a wash process to remove manufacturing material after EP. and cleaning process will be validated. Or is there any other purpose of cytotoxicity of EP that I did not see?

Also some customer will have dimension removal requirement, for example the dimension change is no more than 0.0006 inch. Don’t want to that either unless specifically required by customer. Because required dimensions on the print will be measured after EP and in final inspection. Although it will be an more solid evidence to show EP capability is stable…

As for parameter for OQ and PQ, I have time, temperature, and voltage.
My biggest question is how do you choose those parameters for different material and product.
I mean normally we have one validated range of above parameters for all material. Or is there someone using different parameter range for different materials or product size?

My problems is, during EP OQ challenge test, I found “time” varies a lot on different material and product (size of product).
For example, for a 5mm diameter drills made of 420B, it takes 20 seconds to remove deburrs and have corrosion resistance layer. But it takes about 1 minute for a 12mm diameter drill made of 420B. And for a 5mm diameter drills made of 17-4, it only needs 5 seconds, and it got burned if putting a 17-4 in EP solution for 1 minute.

Does anybody have any suggestion on what parameter range do you use for OQ?
Do you have a large range on those parameters?
 
Elsmar Forum Sponsor
K

Karishma

#2
HI,

I think for each material it needs to define OQ & PQ parameter range separately, that's why "change in material" is also a process re validation trigger.
 
Thread starter Similar threads Forum Replies Date
C Electropolishing of 316TI Manufacturing and Related Processes 2
M Procedure for Validating Laser Marking & Electropolishing of Orthopedic Implants Other Medical Device and Orthopedic Related Topics 13
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 1
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
D Test Method Validation Qualification and Validation (including 21 CFR Part 11) 4
T Laboratory Verification after validation ISO 17025 related Discussions 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 6
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
A Validation of Forced Aeration Process ISO 13485:2016 - Medical Device Quality Management Systems 3
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
M API 4F/7K/8C Design Package Validation Oil and Gas Industry Standards and Regulations 2
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
T Annual Validation as a detection mode on a PFMEA? FMEA and Control Plans 5
B TMV - Selection of TM's for Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 6
P Human Factors / Usability validation in the time of COVID Human Factors and Ergonomics in Engineering 14
C Template for Excel Validation Reliability Analysis - Predictions, Testing and Standards 6
M IT validation for a paper based MD repair company QMS ISO 13485:2016 - Medical Device Quality Management Systems 6
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
M Is Validation of Plating Processes required and who is responsible? Qualification and Validation (including 21 CFR Part 11) 11
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
Stoic Manual soldering processes - 100% verifiable, or always requiring validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Rees System Validation Qualification and Validation (including 21 CFR Part 11) 1
K PQ validation qualification - Asked to write a PQ protocol ISO 13485:2016 - Medical Device Quality Management Systems 6
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 9
Y Retrospective Validation - Class I device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
M Production approval testing - Alternative ideas for Validation Reliability Analysis - Predictions, Testing and Standards 4
M Validation of two nearly identical products Other Medical Device Regulations World-Wide 5
J Requested Validation plan and reports Manufacturing and Related Processes 4
S Validation Records - Very young QMS Qualification and Validation (including 21 CFR Part 11) 2
M Test method validation - Is MSA (MSA1, MSA2, MSA3 and linearity) a good solution? Medical Device and FDA Regulations and Standards News 1
G Devices from IQ, OQ or PQ process to be used for verification, validation and summative? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
K ERP System Software Validation - ISO13485 2016 4.1.6 Design and Development of Products and Processes 8

Similar threads

Top Bottom