Electropolishing Validation of Orthopedic Parts - What OQ parameter range

#1
We’re a contract manufacturer and now we're doing a EP validation for different type of materials.
We have 440A, 420B, 17-4, 455, 304, 316.

We never get formal input from our customers. So my intension is to define two validation purpose, one is the surface roughness is reduced, and the other is corrosion resistance (ASTM F1089).

I understand some customer requires to test cytotoxicity. I don’t want do it, because I have a wash process to remove manufacturing material after EP. and cleaning process will be validated. Or is there any other purpose of cytotoxicity of EP that I did not see?

Also some customer will have dimension removal requirement, for example the dimension change is no more than 0.0006 inch. Don’t want to that either unless specifically required by customer. Because required dimensions on the print will be measured after EP and in final inspection. Although it will be an more solid evidence to show EP capability is stable…

As for parameter for OQ and PQ, I have time, temperature, and voltage.
My biggest question is how do you choose those parameters for different material and product.
I mean normally we have one validated range of above parameters for all material. Or is there someone using different parameter range for different materials or product size?

My problems is, during EP OQ challenge test, I found “time” varies a lot on different material and product (size of product).
For example, for a 5mm diameter drills made of 420B, it takes 20 seconds to remove deburrs and have corrosion resistance layer. But it takes about 1 minute for a 12mm diameter drill made of 420B. And for a 5mm diameter drills made of 17-4, it only needs 5 seconds, and it got burned if putting a 17-4 in EP solution for 1 minute.

Does anybody have any suggestion on what parameter range do you use for OQ?
Do you have a large range on those parameters?
 
Elsmar Forum Sponsor
K

Karishma

#2
HI,

I think for each material it needs to define OQ & PQ parameter range separately, that's why "change in material" is also a process re validation trigger.
 
Thread starter Similar threads Forum Replies Date
C Electropolishing of 316TI Manufacturing and Related Processes 2
M Procedure for Validating Laser Marking & Electropolishing of Orthopedic Implants Other Medical Device and Orthopedic Related Topics 13
N Validation of special processes - thread closed Oil and Gas Industry Standards and Regulations 3
L Validation without Tolerance Qualification and Validation (including 21 CFR Part 11) 0
shimonv Test Method Validation ISO 13485:2016 - Medical Device Quality Management Systems 10
R Debug mode in software/device validation IEC 62304 - Medical Device Software Life Cycle Processes 2
M Software verification and validation AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
Melissa Process Validation of Rotary Heat Sealer Speeds Design and Development of Products and Processes 4
C Validation of process for releasing the UDI EU Medical Device Regulations 4
C Validation of process for production and servicing 5.7.1.5 API Spec Q1, 9th Edition Oil and Gas Industry Standards and Regulations 5
W Transport validation Qualification and Validation (including 21 CFR Part 11) 4
M Is validation required when consumables are changed Qualification and Validation (including 21 CFR Part 11) 6
C 8.3.4 Verification and Validation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
W Transport validation Qualification and Validation (including 21 CFR Part 11) 12
mustomutlu Process Validation Final Report Other Medical Device and Orthopedic Related Topics 2
P Software verification and validation procedure IEC 62304 - Medical Device Software Life Cycle Processes 6
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 8
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 1
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 3
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
T Laboratory Verification after validation ISO 17025 related Discussions 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 10
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
A Validation of Forced Aeration Process ISO 13485:2016 - Medical Device Quality Management Systems 3
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
M API 4F/7K/8C Design Package Validation Oil and Gas Industry Standards and Regulations 2
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
T Annual Validation as a detection mode on a PFMEA? FMEA and Control Plans 5
B TMV - Selection of TM's for Validation ISO 13485:2016 - Medical Device Quality Management Systems 5

Similar threads

Top Bottom