Hi all,

We want to test the electrosurgery interference according to the standard 60601-2-27 section 202.6.2.101. According to the test a HF surgical equipment with working frequency of 400 kHz ± 10% is required. However, the HF surgical equipment we have available has a working frequency of 300 kHz. What is the reason behind the frequency requirement in the standard?

Thanks!
 
Elsmar Forum Sponsor
#2
Hi Ana,

If you check the rationale for that clause (in the informative annex AA) it states -
There is no ideal test method to generate electrosurgical interference in a test laboratory but the ones given...has been shown to give similar results to ...surgical practice and to be reproducible.
(I only have the outdated 2006 version - so this may have been altered).

The specifications given are there to try and create a reproducible test - always a significant issue in EMC testing, especially so for elctrosurgical interference.

So can you use a 300kHz generator instead of a 400kHz unit?
I think that this depends on what you hope to achieve by your testing.

If you are doing "pre-compliance" testing as part of your development work, then I would suggest the 300kHz unit could be used and could give strong confidence (90% +) that your device copes with the interference. Then, when you do the full compliance tests with a 400kHz unit you are unlikely to have any 'nasty surprises'.

If you are doing an in-house test programme to prove compliance I might suggest that a 'strong positive' test result form the 300kHz unit would be sufficient, but with the deviation of test equipment clearly noted in your test results.

However, if you are hoping to get full certification by one of the internationally recognised test houses, and use their test mark, then I would expect that they would only accept test results using a 400kHz test unit.

Welcome to the forums!
 
Last edited:
Thread starter Similar threads Forum Replies Date
K FDA 510k electrosurgery 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
R Electrosurgery Immunity Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Electrosurgery Emission Compatibility - Patient monitoring equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Informational US FDA – Programmable CSF Shunts and Magnetic Field Interference with Implanted Hearing Devices – Letter to Health Care Providers Medical Device and FDA Regulations and Standards News 0
J Using Gage Pins - Inside diameter - Interference fit General Measurement Device and Calibration Topics 4
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 2
P 7.5.10 Customer property - applies to leased/rented equipment? ISO 13485:2016 - Medical Device Quality Management Systems 10
S Class I Equipment with mobile power inverter IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W Equipment Bracketing Approach US Food and Drug Administration (FDA) 0
H Still having trouble as to what equipment falls under 8.5.1.1 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Register of the monitoring and measuring equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
B Do CTQs need to be equipment specific? FMEA and Control Plans 8
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
C Revalidation of Process Equipment for Equipment Transfered to New Facility. ISO 13485:2016 - Medical Device Quality Management Systems 5
V Clause 7.1.3.1 Plant, facility, and equipment planning Lean in Manufacturing and Service Industries 0
A Canadian Safety Labels for medical electrical equipment Canada Medical Device Regulations 2
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
P Equipment URS, pFMEA and dFMEA ISO 14971 - Medical Device Risk Management 3
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
L Gage RandR on automated equipment. IATF 16949 - Automotive Quality Systems Standard 7
W How long do you keep information about equipment no longer used? Document Control Systems, Procedures, Forms and Templates 2
B 4DINSPEC Inspection Equipment General Measurement Device and Calibration Topics 0
D Equipment Register and PM question ISO 13485:2016 - Medical Device Quality Management Systems 2
R Classification of USB DC operated medical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
J Confusing on Harmonic Emissions/Voltage Fluctuations/Flicker test for CISPR 11 Class A Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C 8.5.1.1 Control of Equipment, Tools, and Software Programs - Questions about the extent of control of NC programs AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Equipment Register related question ISO 13485:2016 - Medical Device Quality Management Systems 1
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
I Equipment service dates ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
M Missing measuring equipment General Measurement Device and Calibration Topics 6
P Equipment 21 CFR 820.70(g) - User Requirements Document for Off the shelf equipment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
K Problems with process equipment ISO 13485:2016 - Medical Device Quality Management Systems 4
D New Equipment form / template Document Control Systems, Procedures, Forms and Templates 5
W Protective earth for Class I ME equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D DQ documents of new equipment - Function specification Qualification and Validation (including 21 CFR Part 11) 0
R MSA studies for Heat Treatment Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
T AMS2570E certification equipment questions AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
J Class Fixed ME Equipment (US Deviation) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
P Manufacturing equipment (e.g.: Dental Welder): medical device or not? US Food and Drug Administration (FDA) 5
R What is meant by Use of COMPONENTS WITH HIGH-INTEGRITY CHARACTERISTICS in ME EQUIPMENT (clause no 4.9) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3

Similar threads

Top Bottom