Electrosurgical Electrodes - Stability Test
medwise
Hello Experts,
I have undergone an audit and was asked to provide an evidence to demonstrate stability to support the required shelf-life storage requirement of -10 to 55 degrees for 12 months for electrosurgical electrodes (CE marked class IIb medical device). We have conducted cleaning validation test and typically the device is sterilised at 134 degrees. Based on this higher limit temperature exposure we determined that 55 degrees will be the higher limit for storage. Again, we also did shipping stability test as per ISTA. However, our NB is not satisfied and would like a documented justification and or test to demonstrate that our device can be stored at -10 to 55 degrees.
Can this be justified by doing risk assessment only? And as an outcome of that risk based assessment justify that the elements of greatest importance were steam based exposure which is far more challenging on product than low based temperature. Please note device is composed of PVC, Rislan Powder Coating and stainless steel and we can look up the SDS to determine the temperature resistance and 25 years of post-market data.
Look forward to your advise and guidance.
medwise
Ronen E
Problem Solver
Moderator
Yes, I would start with an engineering assessment. Try to analyse / understand the failure mechanisms in your device that act over time, and see if they're enhanced by low temperatures (-10), high temperatures (55), thermal cycles and generally any temperature profile (common and extreme) that the device is likely to be exposed to during its storage, shipping, shelf-life and use-life. Also pay attention to other relevant environmental factors such as RH, sunlight, chemical exposure etc.
Please note that a short term exposure (eg sterilisation) is not necessarily representative of longer terms exposures, even at lower temperatures (eg 55 deg). Some failure mechanisms take time, and your goal is to establish that a 12 months exposure is safe.
If you have 25 years worth of post-market data (relevant to the current design & materials!) then I guess your job should be much easier.
Please PM me if you want to work on this together.
Cheers,
Ronen.
P Electrosurgical Device User Need: Cord Flexibility -> Requirement Other Medical Device and Orthopedic Related Topics 4
N Electrosurgical generator with switchable sockets - creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
W Electrosurgical devices/surgery IFU contraindications EU Medical Device Regulations 2
S Defibrillator Requirement for ME Equipment connected to HF Electrosurgical Generator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC60601 3rd Edition Insulation Thickness for electrosurgical devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Accuracy for electrosurgical analyzer General Measurement Device and Calibration Topics 2
H Clinical Evaluation Template for an Electrosurgical Device or Accessory Document Control Systems, Procedures, Forms and Templates 2
W EN 61000-4-4:2004 Compliance - Electrosurgical Unit Transient Voltage Pulse EU Medical Device Regulations 3
N Electrosurgical Testing - Is there a specific standard for Electrosurgical Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Classification rules for Canada - Electrosurgical and Cryosurgical Equipment ISO 13485:2016 - Medical Device Quality Management Systems 8
A 60601-1 : Integrated Dry ECG Electrodes = 2 Patient connections inside 1 applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A 60601-1-2 : Integrated Dry Electrodes = PATIENT COUPLED PORT? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D Defibrillation protection for patient connection except for electrodes IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S IEC 60601-2-25 ECG 4 electrodes and derived leads IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N ECoG/EEG electrodes - Required IEC 60601-1 tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Are EEG electrodes considered as one applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Standards for testing of adhesive electrodes for use with TENS machines IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Are disposable electrodes in the scope of WEEE Directive? Other ISO and International Standards and European Regulations 2
B Applications required for main accessories (electrodes) of CE Medical Products EU Medical Device Regulations 4
N Must electrodes for Rx/prescription stimulators be Rx only? Other US Medical Device Regulations 2
T Do Disposable Electrodes need CT-Compatible Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J IEC 60601-1 for TENS-electrodes (CE-marking) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
J Sourcing Rubber or Silicone Electrodes for Medical Device Other ISO and International Standards and European Regulations 2
A Classification of ECG/EKG Electrodes according to MDD 93/42/EEC? EU Medical Device Regulations 2
D ECG Electrodes - Duration of Use (How long they are expected to stay on the patient) Other US Medical Device Regulations 2
D Independent Testing of ECG Electrodes in accordance with AAMI standards Reliability Analysis - Predictions, Testing and Standards 5
N CE Mark (CE directives) for Welding Electrodes CE Marking (Conformité Européene) / CB Scheme 5
bio_subbu FDA Issues Guidance Document on Electrocardiograph (ECG) Electrodes Other US Medical Device Regulations 0
J Pull test standard? Wires coming off the electrodes at the weld ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D What is the acceptance criteria/requirement for Stability Study? General Measurement Device and Calibration Topics 2
M Device mounted at IV pole - what about mechanical stability testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Use of OQ cleared batch for stability and clinical study use ISO 13485:2016 - Medical Device Quality Management Systems 0
C Surgical mask stability testing (CE mark) EU Medical Device Regulations 2
G Engineering retest in line - Machine stability issues IATF 16949 - Automotive Quality Systems Standard 8
G Accelerated stability of medical devices Other Medical Device Related Standards 1
R Revalidation machine line and Stability Studies impact. Medical Device and FDA Regulations and Standards News 3
S Stability accelerated vs. real-time Other Medical Device Related Standards 3
S Standards for real time stability for medical devices Other Medical Device Related Standards 2
C Medical Device Stability Studies - Non-validated pilot production Other Medical Device Related Standards 1
B Stability testing of medical devices as per ASTM F1980-16 CE Marking (Conformité Européene) / CB Scheme 0
lilybef Number of lots required for stability testing of a Class II device in the 510(k) Qualification and Validation (including 21 CFR Part 11) 3
V System suitability or stability during continuous usage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
F Extension of a stability study for an API Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
O New stability - Yes or No - Similar product but for different intended use EU Medical Device Regulations 1
A Accelerated Stability Studies - Class III Medical Gloves Other Medical Device and Orthopedic Related Topics 1
ernieto On going Stability for IVDs Other US Medical Device Regulations 5
A Risks related to Method Validation and Stability Studies Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
N Sampling for stability studies - Class II surgically invasive devices Other Medical Device and Orthopedic Related Topics 1
K Food Shelf Life Stability Study Statistical Analysis Tools, Techniques and SPC 1
C IQ and OQ for a "Dormitory" Refrigerator - Temperature Stability Qualification and Validation (including 21 CFR Part 11) 4