EMC Emission Issues - Communication between PC and Monitor
A
AdHocSpecialist
#1
We are currently testing our device on EMC emissions, and unfortunately we are over the limit. We have found the source of our problem which is the communication between the PC and the monitor. The cabled are acting as antennas.
We are currently using an adapter from DVI to optical fibre. This reduces the emission a lot, but just not enough to pass the test. The troublesome frequency is at 742.5 MHz.
Have some of you encountered a similar problem regarding EMC and have you found a solution for the communication between pc and monitor?
Elsmar Forum Sponsor
C
cdewitt
#3
I would look very carefully at the PC board layout of the DVI interface (and the rest of the board also) and the way the shield on the cable is connected to the enclosure shield and PCB ground. DVI is used other places without fiber optic interface so it is possible to pass.
For help on this look at Bruce Archambeault's book "PCB Design for Real-World EMI Control" and Henry Ott's book "Electromagnetic Compatibility Engineering".
Thread starter Similar threads Forum Replies Date
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
L IEC60601-1-11 Home Healthcare + IEC60601-1-2 EMC (Functional Earth allowed?) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M EMC Directive and product safety standards CE Marking (Conformité Européene) / CB Scheme 1
B Evaluation of Basic Safety during EMC Immunity or Climate Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B Does EMC Directive 2014/30/EU cease to apply when the Machine Directive 2006/42/EC is applicable? CE Marking (Conformité Européene) / CB Scheme 4
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F EMC testing for a system that is provided with a computer IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
O EMC - EFT (Electrical Fast Transients) on SIP couples erroneously with other cable CE Marking (Conformité Européene) / CB Scheme 1
K Is a 3V torch classified as medical electrical equipment needing EMI/EMC testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Arguing with the test lab about failing in an EMC test IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
E Airborne Medical Equipment EMI/EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Portugal ISO 13485 Certification and EMC Test Houses Other Medical Device and Orthopedic Related Topics 3
T Question: Do we consider 60601-1-2 (4th ed) EMC testing a mode for a diagnostic port? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
O Medical Device EMC Risk Management CE Marking (Conformité Européene) / CB Scheme 4
O Guide on EMC for Functional Safety CE Marking (Conformité Européene) / CB Scheme 3
T CE Marking & Compliance to MDD, EMC, RoHS, and GPSD REACH and RoHS Conversations 4
J IEC 60601-2-2 - EMC Requirements - Surgical instrument manufacturer IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
O EMC testing per IEC 60601-1-2 for considering new cables CE Marking (Conformité Européene) / CB Scheme 1
O EMC testing under RED and IEC 60601-1-2 CE Marking (Conformité Européene) / CB Scheme 1
D EMC Testing for a Production Fixture CE Marking (Conformité Européene) / CB Scheme 0
Q IVD vs EMC OJ (Official Journal of the European Union) Harmonised Standard Conflicts EU Medical Device Regulations 1
F Is EMC + Radio Re-Test Required? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Particular standards, essential performance in EMC Immunity Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F EMC (Electromagnetic Compatibility) Conformity Charts - IFU Requirements Misc. Quality Assurance and Business Systems Related Topics 3
G CE Marking - Low Voltage Directive and EMC Directive CE Marking (Conformité Européene) / CB Scheme 4
T Firmware version in EMC report / DoC CE Marking (Conformité Européene) / CB Scheme 1
W EMC, ESD for battery operated apparatus Other Medical Device and Orthopedic Related Topics 1
N New EMC IEC60601-1-2 4.0 Ed. Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B What is our requirement for maintaining the EMC series of standards? IEC 61000 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P EMC requirements for an accessory to a medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T EMC for Medical Devices in Home Environment Other US Medical Device Regulations 4
J EMI/EMC Test differences for Radiography Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M EMC Testing for a US-only, Home-Based Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Safety & EMC certificate updated when MCU (MIcro-Controller) is changed? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T EMC Tests - While certified power supply is part of device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B EMC Repeat Justification - Reducing NRTL testing required after modification of ME IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C EMC Test Report Requirements to IEC 60601-1-2:2009 ISO 13485:2016 - Medical Device Quality Management Systems 4
O CE TCF (Technical Construction File) Template for LVD/EMC/Pressure Directive CE Marking (Conformité Européene) / CB Scheme 5
P New EMC standard IEC 60601-1-2:2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T IEC 60601-1-2 EMC and WiFi IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 EMC Testing Optional in US? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
S Does Australia/New Zealand require RoHS, WEEE, REACH and EMC? RoHS, REACH, ELV, IMDS and Restricted Substances 4
N Device with Safety & EMC Test Certificates from different Laboratories ISO 13485:2016 - Medical Device Quality Management Systems 22
N User Manual Content Requirements and EMC information ISO 13485:2016 - Medical Device Quality Management Systems 1
B Should EMC (Electromagnetic Compatibility) Testing be performed In-House CE Marking (Conformité Européene) / CB Scheme 10
C An ongoing EMC test in accordance with IEC/EN 60601-2-47 EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E List of Complete EMC Test Requirements consisting of IEC, EN, Cispr, etc. General Measurement Device and Calibration Topics 2
T GPRS (General Packet Radio Service) and EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Similar threads
Top Bottom