tallu
FDA suggests using the 4th edition of IEC 60601-1-2 for Home Healthcare environments. In 4th edition the test levels for power frequency magnetic fields increased tenfold, to 30 A/m.
If company is seeking FDA clearance for class II device (tested against 3 A/m) and does not perform testing against 30 A/m they need to have (very strong?) rationale why the test was not performed.
Somehow the company should proof why testing against 3 A/m is acceptable for their device (through a study of published literature or environmental measurements).
What kind of rationale could safisty the FDA?
Any comments?
If company is seeking FDA clearance for class II device (tested against 3 A/m) and does not perform testing against 30 A/m they need to have (very strong?) rationale why the test was not performed.
Somehow the company should proof why testing against 3 A/m is acceptable for their device (through a study of published literature or environmental measurements).
What kind of rationale could safisty the FDA?
Any comments?