EMC Statement in Instruction Manual
N
nikolaos
#11
Cyberice, the declaration of the manufacturing and the tables in the IFU is part of the certification according to IEC/EN 60601-1-2. This is very well explianed in the standard and all Notified Bodies will accept this.
cyberice
#12
That's right - but we don't put the tables in the IFU - we only put the link in the IFU - that's I want to explain :) It take a while for us to persuade the auditor - that a link only is acceptable!
Because in our mind it doesn't make really sense to put a table with numbers and units in an IFU for the customer - he won't understand it! :)
N
nikolaos
#13
The standard clearly states that the tables must be inserted in the IFU. If the medical device is for hospital use the Medical Engineers should check the tables in the IFU and verify the correct enviroment for proper operation of the device. Table 203 specifies the separation distances between portable a mobile RF equipment and the device. This is an important information for the safe operation of the equipment. A web link can go down anytime and, in my opinion, cannot be used for safety issues
Peter Selvey
#14
Just to clarify one possible important point here:
The EMC tables are not required in the instructions for use, although they commonly included there for convenience.
IEC 60601-1 [2nd ed] requirements are split in two parts:
6.8.2 - instructions for use (i.e. instruction manual)
6.8.3 - technical description (i.e. service manual)
The EMC tables fall under Clause 6.8.3, which means there were intended for a service manual. Of course, if a manufacturer chooses not to provide a service manual then the tables should be in the instruction manual.
It should be noted that both the "instructions for use" and "technical description" fall under "accompanying documents" (defined term) which means they must be physically provided with the equipment to comply with the standard.
Providing EMC tables on the web site might be reasonable (I make no comment on this) but it does not comply with the standard. If you do not comply with a standard there are special provisions under the MDD (EU), FDA, Japanese PAL, CMDR, Australian TGA systems. For example, in Europe, the presumption of conformity no longer exists, and you need to provide the "alternate solution" directly against relevant essential requirements. It is not a matter of whether your Notified Body says it is OK or not. The decision is up to you, but you need special evidence.
Also, take care with legal liability. Let's say an EMC interference problem occurs (equipment position closer than as allowed in the table). If you comply with the standard, it is easier to defend. But if the tables are missing (even if they are on a web-site), it is a non-conformity which becomes more difficult to defend against.
I understand the EU thinking about allowing electronic (website) documentation, but no decision yet.
N
nikolaos
#15
Just to clarify.
The "Instruction for use" (6.8.2.201 of 60601-1-2) and the "Technical description" (6.8.3.201 of 60601-1-2) for us as a medical device manufacturer is what we call "user's manual". In this manual there are the EMC Tables, and for us is falls under the "Accompanying documents".
We also have a "service manual" with detailed technical description, schemanics, diagrams but this is only for our field engineers and not for the user.
Anyway the tables are 3 pages, where is the problem to include them in the "Accompaying documents?
I
iangregson
#16
Is that sufficient ?
MAny IFU I have looked at have a large list of technical data.
I was told you have to demonstrate that your statment meets requirements. I have a similar problem and am struggling to get a definitive answer.
What is the minimum that can be put in IFU ?
N
nikolaos
#17
I have the tables as a doc document and can be edited according the instruction of the IEC60601-1-2 standard and the use of the EMC test report from the EMC testing Lab.
After this you can insert them into the user manual, IFU, operator's manual...
I am not sure if I can attach this file to the forum. The file contains just the tables of the standard.
Can a moderator clarify this?
Thank you
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
L IEC60601-1-11 Home Healthcare + IEC60601-1-2 EMC (Functional Earth allowed?) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M EMC Directive and product safety standards CE Marking (Conformité Européene) / CB Scheme 1
B Evaluation of Basic Safety during EMC Immunity or Climate Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B Does EMC Directive 2014/30/EU cease to apply when the Machine Directive 2006/42/EC is applicable? CE Marking (Conformité Européene) / CB Scheme 4
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F EMC testing for a system that is provided with a computer IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
O EMC - EFT (Electrical Fast Transients) on SIP couples erroneously with other cable CE Marking (Conformité Européene) / CB Scheme 1
K Is a 3V torch classified as medical electrical equipment needing EMI/EMC testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Arguing with the test lab about failing in an EMC test IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
E Airborne Medical Equipment EMI/EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Portugal ISO 13485 Certification and EMC Test Houses Other Medical Device and Orthopedic Related Topics 3
T Question: Do we consider 60601-1-2 (4th ed) EMC testing a mode for a diagnostic port? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
O Medical Device EMC Risk Management CE Marking (Conformité Européene) / CB Scheme 4
O Guide on EMC for Functional Safety CE Marking (Conformité Européene) / CB Scheme 3
T CE Marking & Compliance to MDD, EMC, RoHS, and GPSD REACH and RoHS Conversations 4
J IEC 60601-2-2 - EMC Requirements - Surgical instrument manufacturer IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
O EMC testing per IEC 60601-1-2 for considering new cables CE Marking (Conformité Européene) / CB Scheme 1
O EMC testing under RED and IEC 60601-1-2 CE Marking (Conformité Européene) / CB Scheme 1
D EMC Testing for a Production Fixture CE Marking (Conformité Européene) / CB Scheme 0
Q IVD vs EMC OJ (Official Journal of the European Union) Harmonised Standard Conflicts EU Medical Device Regulations 1
A EMC Emission Issues - Communication between PC and Monitor IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Is EMC + Radio Re-Test Required? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Particular standards, essential performance in EMC Immunity Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F EMC (Electromagnetic Compatibility) Conformity Charts - IFU Requirements Misc. Quality Assurance and Business Systems Related Topics 3
G CE Marking - Low Voltage Directive and EMC Directive CE Marking (Conformité Européene) / CB Scheme 4
T Firmware version in EMC report / DoC CE Marking (Conformité Européene) / CB Scheme 1
W EMC, ESD for battery operated apparatus Other Medical Device and Orthopedic Related Topics 1
N New EMC IEC60601-1-2 4.0 Ed. Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B What is our requirement for maintaining the EMC series of standards? IEC 61000 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P EMC requirements for an accessory to a medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T EMC for Medical Devices in Home Environment Other US Medical Device Regulations 4
J EMI/EMC Test differences for Radiography Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M EMC Testing for a US-only, Home-Based Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Safety & EMC certificate updated when MCU (MIcro-Controller) is changed? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T EMC Tests - While certified power supply is part of device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B EMC Repeat Justification - Reducing NRTL testing required after modification of ME IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C EMC Test Report Requirements to IEC 60601-1-2:2009 ISO 13485:2016 - Medical Device Quality Management Systems 4
O CE TCF (Technical Construction File) Template for LVD/EMC/Pressure Directive CE Marking (Conformité Européene) / CB Scheme 5
P New EMC standard IEC 60601-1-2:2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T IEC 60601-1-2 EMC and WiFi IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 EMC Testing Optional in US? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
S Does Australia/New Zealand require RoHS, WEEE, REACH and EMC? RoHS, REACH, ELV, IMDS and Restricted Substances 4
N Device with Safety & EMC Test Certificates from different Laboratories ISO 13485:2016 - Medical Device Quality Management Systems 22
N User Manual Content Requirements and EMC information ISO 13485:2016 - Medical Device Quality Management Systems 1
B Should EMC (Electromagnetic Compatibility) Testing be performed In-House CE Marking (Conformité Européene) / CB Scheme 10
C An ongoing EMC test in accordance with IEC/EN 60601-2-47 EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E List of Complete EMC Test Requirements consisting of IEC, EN, Cispr, etc. General Measurement Device and Calibration Topics 2