EN 45502 - Protection from HARM to the patient caused by heat
Dear all,
for the following EN 45502 requirement:
In the absence of external influence, no outer surface of an implantable part of the ACTIVE IMPLANTABLE MEDICAL DEVICE not intended to supply heat to the PATIENT shall be greater than 2 °C above the normal surrounding body temperature of 37 °C when implanted, and when the ACTIVE IMPLANTABLE MEDICAL DEVICE is in normal operation or in any single component failure (see 19.3).
How shall I demonstrate the compliance with this requirement : test, fault tree analysis, ...
Thanks in advance
Best regards
Tidge
I don't work with ACTIVE IMPLANTABLE medical devices, but this strikes me as the sort of basic safety requirement that requires an actual test to determine the effectiveness of the design to meet the requirement.
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
S Defibrillator Protection applicability When using Multiple applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
V Setup for testing against ISO14708 clause 16 (protection of the patient from herms caused by heat) Other Medical Device Related Standards 0
J Surge Protectors and other power protection Manufacturing and Related Processes 7
M Data Protection and Privacy Policy - looking for a template/example EU Medical Device Regulations 1
S Battery powered device - electrical protection requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 18
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
F General Data Protection Regulation (GDRP) CE Marking (Conformité Européene) / CB Scheme 6
B Operator protection - When to apply table 7, Dielectric strength test voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D Defibrillation protection for patient connection except for electrodes IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Applicability of Means of Protection, working voltage in an Automated External Defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L IEC 60601-1-11 IP22 Compliance - Ingress protection testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
MrTetris GDPR - General Data Protection Regulation - Only applicable to EU data? Other ISO and International Standards and European Regulations 6
M Informational USFDA Final Rule – Human Subject Protection; Acceptance of Data From Clinical Investigations for Medical Devices Medical Device and FDA Regulations and Standards News 0
S Are Defibrillator Protection and Pacemaker pulse detection features mandatory as per IEC 60601-2-25 or 2-27...? IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
S Defibrillator protection test IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
K Is Defib Protection mandatory for wearable ECG? IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
V Software as control or protection will lead to different Software Safety Class? IEC 60601 - Medical Electrical Equipment Safety Standards Series 18
L GDPR scope - "Personal data" definition - General Data Protection Regulation EU Medical Device Regulations 5
S GDPR (General Data Protection Regulation) - My company is ISMS certified IEC 27001 - Information Security Management Systems (ISMS) 3
D IEC 60601 Defibrillation protection test - Metal foil IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W EU GDPR General Data Protection Regulation - What we need to update for our QMS EU Medical Device Regulations 14
T GDPR - General Data Protection Regulation (EU and UK 2018) Other ISO and International Standards and European Regulations 7
M IEC 60601-1 and Active Current Protection Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
K Thoughts on the impact of the General Data Protection Regulation? Medical Information Technology, Medical Software and Health Informatics 5
R Internal Audit of Information Security and Data Protection Internal Auditing 6
B Primary lithium battery protection and vented gas IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P Question on Table 6 "Test voltages for solid insulation forming a MEANS OF PROTECTION IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K ESD Protection in the Assembly Bay ISO 13485:2016 - Medical Device Quality Management Systems 9
R Determination of IP (International Protection Marking) Rating IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
U USB Connector Operator Protection (MOPP) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A IEC 60601-1 cl.15.4.2.1 - Tubular Heating Element (Protection against Overheating) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
F How to provide 2 MOPP (Means of Patient Protection) - Heating Rod IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
F MOP (Means Of Protection) - Creepage and Clearance questions IEC 60601 - Medical Electrical Equipment Safety Standards Series 19
F Evaluation of MOP (Means of Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
L Intellectual Property Protection - Storage, Access, Distribution, etc ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
B Use of Y1 capacitors for MOPP (means of patient protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R Security Standard referred to as TAPA (Transported Asset Protection Association) Other ISO and International Standards and European Regulations 1
J Data Protection - Documents on Desktops ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
S 510(K) Ownership and Protection from Use by Others 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S How to implement protection for multi-winding transformer? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W Medical Device Single fault test-protection against fire IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
R Protection of Records from Unauthorized Access or Alteration Records and Data - Quality, Legal and Other Evidence 4
L ESD Protection and Danger for Operators Manufacturing and Related Processes 3
E Data Protection Compliance Plan example wanted Misc. Quality Assurance and Business Systems Related Topics 6
M Can I continue using IEC60601-1 Ed2 PSU for MOOP (Means Of Operator Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What is "protection" as in section 7.5.5 "Preservation of product" ? IATF 16949 - Automotive Quality Systems Standard 7
K Employee ESD Protection Methodologies - Need input. Training - Internal, External, Online and Distance Learning 8
G Placing a dollar value on our QMS (we are looking at IP protection) Quality Manager and Management Related Issues 3