Hello, relatively new to 60601-1 with regards to making use of AC power supply. Please bear with me while I fumble through this question:

The device is classified as medical weigh scale. The scale is separated into 2 components:
1) Display Fixture that is embedded into the wall
2) Weighing platform Frame consisting of load cells that is embedded into the floor

There are no motors, or heating devices.



The device has a UL purchased power supply which drops it to 12Vdc which is used to power the load cells, display electronics, and printer.

I am questioning the cable that goes from the display fixture to the weighing platform frame which is contained within conduit. The voltage in the wire is less than 12V and the current is less than 12mA. The cable is a circular DIN style connector on each end.

What type of insulation voltage rating is required to meet the EN 60601-1 specification?


Thank you for your time

Qu1nn
 
Elsmar Forum Sponsor
T

temujin

#2
Re: 60601-1: Cable requirement low voltage DC

Hi and welcome to the Cove,

The question you need to ask is if the cable can be classified as an Accessible Part, i.e. will it come into contact (unintentionally) with the patient? (..which I believe it will).

THere is a table in 60601-1 with isolation thicknesses for different voltages, though I do not remember where. (I do not have the standard any more).

Another thing you might need to look up is the flammability class of the isolation.

If you already have a Notified Body, they will be able to help you out.

regards
t.
 

qu1nn

Involved In Discussions
#3
Re: 60601-1: Cable requirement low voltage DC

Thank you for your response, it is appreciated.

according to the definitions contained within EN60601-1:2006

Accessible Part (3.2)
"part of electrical equipment other than an APPLIED PART that can be touched by means of the standard test finger"

This would really not apply as the cable in question is behind the wall at the display, in the conduit to the scale, and under the scale platform. The cable would not be even visible unless the platform was removed... So I don't think that it would be considered Accessible (the only time it would be accessible is during the initial installation of the product... that is unless that counts??)

I would think that the part would more closely match the definition of SIGNAL INPUT/OUTPUT PART (3.115) - "part of ME EQUIPMENT, not being an APPLIED PART, intended to deliver or receive signals to or from other electrical equipment, for example, for display, recording or data processing."

where the APPLIED PART (3.8) - "part of ME EQUIPMENT that in NORMAL use necessarily comes into physical contact with the PATIENT for ME EQUIPMENT of an ME SYSTEM to perform its function."

Do you agree that the APPLIED PART in this case would be the platform that the user stands on in order to take a weight measurement?


I have some more to look up right now.
Thanks again, look towards more comments.

qu1nn
 
T

temujin

#4
Re: 60601-1: Cable requirement low voltage DC

Hi again,
This would really not apply as the cable in question is behind the wall at the display, in the conduit to the scale, and under the scale platform. The cable would not be even visible unless the platform was removed... So I don't think that it would be considered Accessible
If the cable is hidden in the wall it is definitely not an Accessible Part.

I would think that the part would more closely match the definition of SIGNAL INPUT/OUTPUT PART (3.115) - "part of ME EQUIPMENT, not being an APPLIED PART, intended to deliver or receive signals to or from other electrical equipment, for example, for display, recording or data processing."

I don´t think you need to classify it as SIP/SOP if it is only within your own equipment. It is not defined as an external interface for other displays to connect to (as far as I understood). That apart, SIP/SOP would be the interface, not the cable itself.

Do you agree that the APPLIED PART in this case would be the platform that the user stands on in order to take a weight measurement?
Yes.


As for the cable, I don´t think this belong in any particular category...it´s just an integral part of your equipment. Though, take a look at the requirements for flammability / fire safety in 60601-1 - I think those will be the most important ones for such a cable.

regards
t.
 

qu1nn

Involved In Discussions
#5
Stumbled on another item that I am confusing myself on:

would you agree that the definition of PATIENT CONNECTION does NOT apply as there is no physical wires (making an individual point) that are getting attached to the patient:

or

that it DOES matter because in a single fault condition a current could flow between the platform and the patient? (single point being entire platform)

I am leaning towards the latter.


PATIENT CONNECTION
(3.78)
"individual point on the APPLIED PART through which current can flow between the PATIENT and the ME EQUIPMENT in NORMAL CONDITION or SINGLE FAULT CONDITION."



Thank you for your time
Qu1nn
 
R

royal77

#6
Hi,
The patient connection need not be exactly wires. Any conductive part will act as a patient connection. for example, the X-ray table on which a patient lies is also a applied part. I think, you should provide two levels of patient protection for the highest voltage available at that applied part.
 
T

temujin

#7
Hi again,

that it DOES matter because in a single fault condition a current could flow between the platform and the patient? (single point being entire platform)
I would rather make sure current cannot flow in single fault condition by applying more means of protection, as pointed out by royal77.

regards
t.
 
Thread starter Similar threads Forum Replies Date
O Is Patient Cable Exempted from EFT Tests? (IEC 60601-1-2) CE Marking (Conformité Européene) / CB Scheme 1
I IEC 60601-2-52 - HD22.10 H05-BQ-F type Power Cable for Medical Beds IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Accredited vs. non-accredited labs for 60601 compliance in the US IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Accredited vs. non-accredited labs for 60601 compliance in the US Other Medical Device Related Standards 0
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
Ed Panek 60601-1-2:2015/A1:2021 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E 60601-1 - Tilt testing - Tensile safety factor IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A 60601-1 : Integrated Dry ECG Electrodes = 2 Patient connections inside 1 applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A 60601-1-2 : Integrated Dry Electrodes = PATIENT COUPLED PORT? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N 60601-2-2 Ed 6 Clause 201.8.8.3.102 - HF instrument leakage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Maximum mains voltage used when applying 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S Applicability of 60601 vs. 60335 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A 60601-1 understanding of stationary device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A ANSI/AAMI versions of 60601-1-2 and related testing requirements Other Medical Device Related Standards 3
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
shimonv Minimum 60601 requirements for a clinical trial IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom