EN 62368 (In an Annex - M.4.2.1) Secondary Lithium Battery Charging Safeguards
We have a system with a secondary lithium battery which needs to meet the charging safe guard requirements of EN 62368.
One of the requirements of EN 62368, tucked away in an annex, is M4.2.1, which states that your device must stop charging a battery at the highest operating temperature of the device.
(Image attached).
Our test lab wants to single fault the battery charging circuit and try to mess with the thermistor input.
Our reading of the requirement is that you single fault the system to ensure no over current / over current voltage occurs.
The second paragraph doesn't indicate you single fault the system for the temperature cut-off. It just says the battery charging circuit must stop charging and is ambiguous as in what states it must work.
Is this selective reading on our side? Do you think this section M.4.2.1 implies you apply faults for the temperature cut-off as well?
(Reason I ask is that, most of the temperature cut-offs are dependent on a thermistor and if you single fault some charging current to the thermistor input it seems like you will cause a malfunction on any modern battery charging IC).
Benjamin Weber
There are two SFC I can think of for a thermistor: open and short. Usually e thermistor is an NTC, so a short will present a very high temperature to the system, an open circuit will present a very low temperature. In the short-case, the charging should be stopped immediatley, independent from the ambient temperature. This should not be a problem. The open circuit on the other hand is more a problem, since the charging will not be stopped when the limit tmeperature is reached. This a SFC that would fit to cl. B.4.6 (short/open of passive components).
But there are some exemptions fro particalr components, maybe these are applicable for you?
CharlieUK
I think it depends
M.4.2.2 Compliance criteria
Compliance is checked by measuring the charging voltage, the charging current and the temperature of each individual cell of the secondary lithium battery under normal operating conditions, abnormal operating conditions and single fault conditions. The cell temperature shall be measured at the points specified by the battery manufacturer. Single fault conditions that may affect the charging voltage or charging current or the temperature shall be applied in accordance with Clause B.4.
B.4.2 Temperature controlling device
Except for temperature controlling safeguards, according G.3.1 to G.3.4, any single device or component of a circuit controlling the temperature during temperature measurement shall be open-circuited or short-circuited, whichever is more unfavourable.
Temperatures shall be measured according to B.1.6.
Is your thermistor a certified component (e.g. UL or TUV?) to IEC 60730-1 ?
E EN 62368-1 or EN 60730-1 for controller? CE Marking (Conformité Européene) / CB Scheme 2
L IEC 62368-1 / Electrical Energy Source vs. Power Source? CE Marking (Conformité Européene) / CB Scheme 2
S EU MDR Annex I Chapter III 23.4 z EU Medical Device Regulations 2
C EU MDR - Annex II 6.1 Pre-clinical and clinical data EU Medical Device Regulations 4
K Section 13.6 Annex 1 MDD 93/42/EEC for Glove EU Medical Device Regulations 0
D Is PMCF really a continuous activity per Annex XIV,Part B? EU Medical Device Regulations 5
R MDR, Annex I, 23.1 Interpretation - IFU on the website EU Medical Device Regulations 5
P MDR/IVDR Annex III - Technical documentation on PMS EU Medical Device Regulations 1
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
K Applicability of Cybersecurity EU MDR 2017/745 Annex 1 23.4(ab), 14.2(d) CE Marking (Conformité Européene) / CB Scheme 3
K Applicability of eIFU as per EU MDR 2017/745 Annex 1 23.1 CE Marking (Conformité Européene) / CB Scheme 0
I IVD - 98/79EC - class IIa - Annex to the EC from the NB states the CM as one of the facilities? CE Marking (Conformité Européene) / CB Scheme 2
R Identify Medical Device characterstics as Annex C of ISO 14971 Risk Management ISO 14971 - Medical Device Risk Management 5
K EU MDR Annex 1 Chapter III: Information in the Instructions for Use-23.4 (e) the performance characteristics of the device; EU Medical Device Regulations 1
Rincewind MDR - GSPR - Annex I - Chapter II EU Medical Device Regulations 4
R An indication that the device is a medical device (MDR, Annex 23.2q) - applicable for accessories? EU Medical Device Regulations 5
S Difference between EU-MDR Annex IX and the Annex-combo X&XI EU Medical Device Regulations 3
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
S EU MDR Annex XIV - Clinical Evaluation Plan - What do these methods mean? EU Medical Device Regulations 16
K Annex XVI Device Transition Timeline under the MDR EU Medical Device Regulations 2
M EU MDR - Annex II 3b - What documents fall under this requirement? EU Medical Device Regulations 3
C IEC 60601-1-8, difference between table 4 and annex D IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A EU MDR Annex I section 11.3 - Medical Devices labelled as having a specific microbial state EU Medical Device Regulations 7
J MDR Annex VIII, Rule 6 Classification - Implication for lower risk CV products? CE Marking (Conformité Européene) / CB Scheme 3
Z 21 CFR and Annex 11 mapping document Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
A EU-MDR Annex I Requirement 10.1(d) EU Medical Device Regulations 3
T Post-Market Surveillance Plan - Requirements outlined in Annex III (1.1b) EU Medical Device Regulations 2
D Applicable MDR Annex's for Class IIa Medical Devices EU Medical Device Regulations 0
J MEDDEV 2.12/1's new MIR & IMDRF terms & codes (IMDRF Annex) EU Medical Device Regulations 1
M MDR Annex IX Chapter I, 2.2 (c) - Device identification procedures during manufacture. EU Medical Device Regulations 1
D Example for Accessories of "Annex XVI non-medical devices" and how to classify it now with respect ot EU MDR Other Medical Device Regulations World-Wide 1
P Performance evaluation (IVDD Annex VIII) & Design and Development Validation Studies EU Medical Device Regulations 1
H Critical Supplier Agreement acc. to NBOG 2010-1 Annex II EU Medical Device Regulations 5
L EU-MDR-Annex I - 10.4 (CMR & Endocrine-disrupting properties) - Labeling Content EU Medical Device Regulations 5
T EU-MDR-Annex I - 10.4 (CMR & Endocrine-disrupting properties) EU Medical Device Regulations 0
M Annex II - Technical Documentation. V&V, Performance and Safety EU Medical Device Regulations 3
S MDR Annex II Design & Manufacturing info - 'Adjuvants' definition EU Medical Device Regulations 3
J IVDR and MDR ANNEX I Requirements Template EU Medical Device Regulations 7
M Annex I - General Safety and Performance Requirements. Precise identity - how provided EU Medical Device Regulations 6
S Class II Medical Device Conformity Assessment Route according to Annex V only CE Marking (Conformité Européene) / CB Scheme 7
F IMDRF opened a Consultation on Annex E & F and the link to ISO 14971 ISO 14971 - Medical Device Risk Management 4
Douglas E. Purdy ISO/IEC 17025:2017 Clause 8 & Annex B ISO 17025 related Discussions 9
W EU GMP Annex XI - What is an "appropriate QMS"? Other Medical Device Related Standards 2
C Annex II EU MDR gap analysis EU Medical Device Regulations 7
W MDR Annex I chapter 23.2 (q) - Information required on the label EU Medical Device Regulations 17
I Formula in the Annex of IEC 61000-4-3 - How is k=7 determined? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Meaning of Annex II, excluding point Clause 4 conformity CE Marking (Conformité Européene) / CB Scheme 4
Paul Simpson Informational The role of Annex SL - High Level Structure of ISO MSS's - Revision Update February 2019 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
N What should be included in the supply contract - Annex 2 list B IVD ISO 13485:2016 - Medical Device Quality Management Systems 6
R Background of Annex X section 1.1d of MDD 93/42/EEC EU Medical Device Regulations 74