I've just been talking to a registrar about the transition options available to us and wanted to check out if his approach is different to other people's experience.

We have CE marking via Annex 7 supported by 9002:94 and 46002:96.

He suggested that 9002:94 is not necessary to support CE marking so long as we maintain the 46002 QA system and meet the specific requirements of the Medical Devices Directive. So we could allow the 9002 to lapse.

In parallel with this we should develop systems that comply with 13485:2003 which he says should be complete by July this year. We could then register to this at our next annual visit in the Summer of next year.

This seems like quite a light approach and takes the pressure off our transition programme this year, and gives us until next Summer to move across to the new approach.

Is this a common interpretation ? I'd welcome any feedback about the transition route other people are taking.



Bill Marshall
 
A

Aaron Lupo

#2
Bill-

EN46002 was replaced by ISO 13485, the last I heard about the transition period for 13485:2003 was that you needed to be compliant with it by 12/15/03. In addition, you do not need to be EN or ISO anything to maintain a CE mark. If you look at the MDD it states, you need to have a quality system but does not say that it must be ISO or EN compliant. Just out of curiosity who is your notified body?
 
B

Bill Marshall

#3
I think that 46002 has been superced by 13485 by not replaced, nor withdraw. In fact the BSI website says "Current".

The notified body is AMTAC and the bloke I spoke to said that there would be a one transition from 46002:97, and a three year transition from 13485:2001.

He then gave two options :-

Stay with 46002 until 1 year after the new 13485 is published,

Or, we could transfer to 13485:96 now, then we would have 3 years after the new 13485 is published.

The second option gives the longer transition time, but I don't really want to encourage the managment team to go any slower on this than necessary.


regards

Bill Marshall
 
A

Aaron Lupo

#4
Bill Marshall said:

I think that 46002 has been superced by 13485 by not replaced, nor withdraw. In fact the BSI website says "Current".

The notified body is AMTAC and the bloke I spoke to said that there would be a one transition from 46002:97, and a three year transition from 13485:2001.

He then gave two options :-

Stay with 46002 until 1 year after the new 13485 is published,

Or, we could transfer to 13485:96 now, then we would have 3 years after the new 13485 is published.

The second option gives the longer transition time, but I don't really want to encourage the managment team to go any slower on this than necessary.


regards

Bill Marshall
Superseded=replace last time I looked.

ISO 13485 was last published in 1996 not 2001 and the revision is currently in
Stage 50.20

FDIS ballot initiated:
2 months.
Proof sent to secretariat.

EN46002 was last published in 1996 not 1997.

As far as having 3 years to upgrade I will have to double check on that, last I knew they were saying you had until 12/15/03 to upgrade.


In any event obtaining 13485:2003 certification should not be very hard to complete. We are currently in the process of upgrading and the amount of work required is minimal.
 
B

Bill Marshall

#5
I get my documents from BSI and the dates I have quoted reflect this. I appreciate that ISO publishing dates are sometimes earlier.


46002 has not yet been "replaced" as the harmonised standard, which is why it's still used and recognised.

It does seem strange that there are different deadlines for the two standards that the new 13485 will replace, that's partly what inspired the question in the first place.

Although I expect the transition will be straightforward, the timing of our annual visit makes it important that we chose the right path now to minimise the effort over the next 12 months. There's no point in doing two transitions, when only one is necessary.
 
A

Aaron Lupo

#6
Bill it is my understanding that EN 13485/88 replaces EN 46001/2. I have seen this stated on numerous websites, however, you have to go by what your Notified Body tells you.

Just an FYI, I was informed that the release of 13485:2003 has now been pushed back until mid to late 2003 from the anticipated dated of 1st quarter 2003. In addition ISO/TC 210/WG 1 is preparing guidance on the transition period (which will more than likely be 3 years now) which will be made publicly available.
 
4

45fan

#7
Regarding harmonized standards, when in doubt about dates and such, I go to the Official Journal of the EU (http://europa.eu.int/comm/enterprise/newapproach/standardization/harmstds/reflist/meddevic.html).

Exerpt re MDD QA system is:

CEN EN ISO 13485/8[:1986...became EN on] 2002-07-31 [OJ publication C182]
Reference of the superseded standard: EN 46001/2
Date of cessation of presumption of conformity of the superseded standard: 2004-03-01 (Note 1)

Note 1: Generally the date of cessation of presumption of conformity will be the date of withdrawal ( "dow"), set by the European Standardisation Organisation, but attention of users of these standards is drawn to the fact that in certain exceptional cases this can be otherwise.

Let me see if I have it right...

EN 46001/2 are in effect until March 1, 04. We can choose to use EN ISO 13485 NOW, if we like, but its not mandatory. When the 2003 rev becomes an EN is anybody's guess, so using the standards to demonstrate conformity for CE mark likely won't change for a while.

So if we're just concerned about our ISO certificate, we will also have until early 2006 to comply with 13485:2003 because of the three-year transition period.

I don't see any reason to rush the transition, unless you have the time and resources.
 
Thread starter Similar threads Forum Replies Date
Marc Systems Requirements Mandated by EN46001 and EN46002 ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 9001 Internal Audits - No production right now due to furloughs Internal Auditing 3
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 0
M Customer Property - ISO 13485:2016 Clause 7.5.10 ISO 13485:2016 - Medical Device Quality Management Systems 9
pbojsen ISO 13485 Requirements versus FDA product classification and GMP exemptions - Audits ISO 13485:2016 - Medical Device Quality Management Systems 3
D "certified" in ISO 19011, as well as IATF required? IATF 16949 - Automotive Quality Systems Standard 4
S ISO/IEC 15408 - Is this is Certifiable Standard? Other ISO and International Standards and European Regulations 2
D Lead time to schedule an ISO 13485 audit Auditing Quality and Environmental Management Systems 2
T Do we need an SOP for ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
K ISO 9001 Auditing in a Healthcare setting Auditing Quality and Environmental Management Systems 15
S Does anyone have a checklist to prepare for ISO 13485, Stage I audit? ISO 13485:2016 - Medical Device Quality Management Systems 1
H QMS ISO 13485:2016 - ISO14971 IEC60304 etc ISO 13485:2016 - Medical Device Quality Management Systems 2
Y How can i integrate ISO 13845 into ISO 27001? ISO 13485:2016 - Medical Device Quality Management Systems 4
vickyva ISO 14155:2020 CIP CIR templates Other Medical Device Related Standards 0
C ISO 9001:2015 8.3.2. h) Design and Development Planning - What is required? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
B Employee Handbook in ISO 9001:2015 Section 7 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
B Operational Procedures for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
Q ISO 9001/IATF 16949 Audit Finding Question - Document Retention IATF 16949 - Automotive Quality Systems Standard 10
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
G Trying to get a financially reasonable ISO Certification Body Registrars and Notified Bodies 8
L ISO 45001:2018 - Clause 5.4: Consultation and Participation of Workers Process Maps, Process Mapping and Turtle Diagrams 1
E ISO 13485 in Clinical Trial conduct: Applicable or No ISO 13485:2016 - Medical Device Quality Management Systems 2
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
Richard Regalado Informational ISO/IEC DIS 27001:2021, to be published soon. IEC 27001 - Information Security Management Systems (ISMS) 0
Q Audit report template ISO 9001/14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
Q Process matrix examples of ISO 9001 & 14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Need ISO 15189:2012 Documentation toolkit. Document Control Systems, Procedures, Forms and Templates 0
chris1price Archiving of paper records - ISO 9001 7.5.3.1b Records and Data - Quality, Legal and Other Evidence 4
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
D Common practices in ISO 9001 deployment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
Q ISO 9001-2015 Internal audit finding Internal Auditing 12
P Audit check for IT company (ISO 9001) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Label Making & Printing Standards ISO / ASTM ISO 13485:2016 - Medical Device Quality Management Systems 5
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance document to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
S ISO 2768-mk print call out Other ISO and International Standards and European Regulations 11
T ISO 17024, clauses 4.3.8. and 5.1.1. Other ISO and International Standards and European Regulations 4
C ISO 14001:2015 6.1.3 Compliance Obligations - Legal requirements monitoring ISO 14001:2015 Specific Discussions 0
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
W First time being audited (ISO 9001), asking for advice ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
John C. Abnet ISO 26262 ISO 26262 - Road vehicles – Functional safety 3
Marc ISO 26262- Road vehicles – Functional safety ISO 26262 - Road vehicles – Functional safety 0
John C. Abnet ISO 26262 IATF 16949 - Automotive Quality Systems Standard 0
A ISO/DIS 15223-1:2020 - Country of manufacture label (IEC 60417 No. 6049) - Which national law requires this symbol? Other Medical Device Related Standards 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
K PDCA cycle and ISO processes alternative model Quality Management System (QMS) Manuals 14

Similar threads

Top Bottom