SBS - The best value in QMS software

EN46002 vs. ISO 13485 - Transition Route

B

Bill Marshall

#1
I've just been talking to a registrar about the transition options available to us and wanted to check out if his approach is different to other people's experience.

We have CE marking via Annex 7 supported by 9002:94 and 46002:96.

He suggested that 9002:94 is not necessary to support CE marking so long as we maintain the 46002 QA system and meet the specific requirements of the Medical Devices Directive. So we could allow the 9002 to lapse.

In parallel with this we should develop systems that comply with 13485:2003 which he says should be complete by July this year. We could then register to this at our next annual visit in the Summer of next year.

This seems like quite a light approach and takes the pressure off our transition programme this year, and gives us until next Summer to move across to the new approach.

Is this a common interpretation ? I'd welcome any feedback about the transition route other people are taking.



Bill Marshall
 
Elsmar Forum Sponsor
A

Aaron Lupo

#2
Bill-

EN46002 was replaced by ISO 13485, the last I heard about the transition period for 13485:2003 was that you needed to be compliant with it by 12/15/03. In addition, you do not need to be EN or ISO anything to maintain a CE mark. If you look at the MDD it states, you need to have a quality system but does not say that it must be ISO or EN compliant. Just out of curiosity who is your notified body?
 
B

Bill Marshall

#3
I think that 46002 has been superced by 13485 by not replaced, nor withdraw. In fact the BSI website says "Current".

The notified body is AMTAC and the bloke I spoke to said that there would be a one transition from 46002:97, and a three year transition from 13485:2001.

He then gave two options :-

Stay with 46002 until 1 year after the new 13485 is published,

Or, we could transfer to 13485:96 now, then we would have 3 years after the new 13485 is published.

The second option gives the longer transition time, but I don't really want to encourage the managment team to go any slower on this than necessary.


regards

Bill Marshall
 
A

Aaron Lupo

#4
Bill Marshall said:

I think that 46002 has been superced by 13485 by not replaced, nor withdraw. In fact the BSI website says "Current".

The notified body is AMTAC and the bloke I spoke to said that there would be a one transition from 46002:97, and a three year transition from 13485:2001.

He then gave two options :-

Stay with 46002 until 1 year after the new 13485 is published,

Or, we could transfer to 13485:96 now, then we would have 3 years after the new 13485 is published.

The second option gives the longer transition time, but I don't really want to encourage the managment team to go any slower on this than necessary.


regards

Bill Marshall
Superseded=replace last time I looked.

ISO 13485 was last published in 1996 not 2001 and the revision is currently in
Stage 50.20

FDIS ballot initiated:
2 months.
Proof sent to secretariat.

EN46002 was last published in 1996 not 1997.

As far as having 3 years to upgrade I will have to double check on that, last I knew they were saying you had until 12/15/03 to upgrade.


In any event obtaining 13485:2003 certification should not be very hard to complete. We are currently in the process of upgrading and the amount of work required is minimal.
 
B

Bill Marshall

#5
I get my documents from BSI and the dates I have quoted reflect this. I appreciate that ISO publishing dates are sometimes earlier.


46002 has not yet been "replaced" as the harmonised standard, which is why it's still used and recognised.

It does seem strange that there are different deadlines for the two standards that the new 13485 will replace, that's partly what inspired the question in the first place.

Although I expect the transition will be straightforward, the timing of our annual visit makes it important that we chose the right path now to minimise the effort over the next 12 months. There's no point in doing two transitions, when only one is necessary.
 
A

Aaron Lupo

#6
Bill it is my understanding that EN 13485/88 replaces EN 46001/2. I have seen this stated on numerous websites, however, you have to go by what your Notified Body tells you.

Just an FYI, I was informed that the release of 13485:2003 has now been pushed back until mid to late 2003 from the anticipated dated of 1st quarter 2003. In addition ISO/TC 210/WG 1 is preparing guidance on the transition period (which will more than likely be 3 years now) which will be made publicly available.
 
4

45fan

#7
Regarding harmonized standards, when in doubt about dates and such, I go to the Official Journal of the EU (http://europa.eu.int/comm/enterprise/newapproach/standardization/harmstds/reflist/meddevic.html).

Exerpt re MDD QA system is:

CEN EN ISO 13485/8[:1986...became EN on] 2002-07-31 [OJ publication C182]
Reference of the superseded standard: EN 46001/2
Date of cessation of presumption of conformity of the superseded standard: 2004-03-01 (Note 1)

Note 1: Generally the date of cessation of presumption of conformity will be the date of withdrawal ( "dow"), set by the European Standardisation Organisation, but attention of users of these standards is drawn to the fact that in certain exceptional cases this can be otherwise.

Let me see if I have it right...

EN 46001/2 are in effect until March 1, 04. We can choose to use EN ISO 13485 NOW, if we like, but its not mandatory. When the 2003 rev becomes an EN is anybody's guess, so using the standards to demonstrate conformity for CE mark likely won't change for a while.

So if we're just concerned about our ISO certificate, we will also have until early 2006 to comply with 13485:2003 because of the three-year transition period.

I don't see any reason to rush the transition, unless you have the time and resources.
 
Thread starter Similar threads Forum Replies Date
Marc Systems Requirements Mandated by EN46001 and EN46002 ISO 13485:2016 - Medical Device Quality Management Systems 4
Le Chiffre Online training available for ISO/IEC 17021-1: Requirements for bodies providing audit and certification of management systems Training - Internal, External, Online and Distance Learning 3
B ISO 6508 and portable hardness measurement instruments General Measurement Device and Calibration Topics 0
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 2
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
Sidney Vianna Release of ISO 10013:2021, Quality management systems – Guidance for documented information Other ISO and International Standards and European Regulations 0
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
R Remote Audits for ISO 9001 (or any other standard) General Auditing Discussions 29
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
D ISO 17021 Certificate Registrars and Notified Bodies 1
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
T Relationship between ISO 9001 and ISO – IEC BS EN 870079- 34 2020 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
S Sequence of ISO 9001:2015 Implementation Steps ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
qualprod Business Continuity Planning in ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
T Help with BS EN ISO - IEC 80079-34 2020 (Explosive atmospheres QMS) Other ISO and International Standards and European Regulations 0
C Sampling - ISO 16269-6 vs ISO 2859 Other ISO and International Standards and European Regulations 0
Brizilla Employee Data Privacy Policy - ISO 9001:2015 requirement(s)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 1
S ISO 9001:2015 Internal Auditing Internal Auditing 8
Q Process: Knowledge Section 7.1.6 of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
N ISO 27001 for Jumb Burger - Risk Assessment sheet IEC 27001 - Information Security Management Systems (ISMS) 10
H What ISO certification is for an IT department? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
P ISO 9001 certification with zero customers? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
M ISO 14971 Determination of Competent Persons ISO 14971 - Medical Device Risk Management 4
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
Q The scope of ISO 21534 Other Medical Device Related Standards 0
A What must be recorded? (ISO 9001:2015, subclause 10.2) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
M ISO 14971:2019: Criteria for overall residual risk ISO 14971 - Medical Device Risk Management 4
B Updated IATF 16949 - Will IATF 16949 get revised when ISO 9001:202X is released? IATF 16949 - Automotive Quality Systems Standard 4
S ISO 9001:2015 vs 21 CFR Part 211 matrix Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
N Quality Compliance Officer - ISO 13485, London Job Openings, Consulting and Employment Opportunities 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
S ISO 9001 implementation in a Gold exporting business ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Does the ISO 9001:2015 standard require a disaster recovery plan or emergency response plan ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
A Environmental Compliance obligations and risks (ISO 14001:2015 6.1.3) ISO 14001:2015 Specific Discussions 3
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A Tips and Tricks to understand ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
M ISO 9001 Major Nonconformance Internal Audit Schedule/COVID-19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18

Similar threads