We are developing a single-use endoscope with additional internal electronics including a small solenoid in the handle.
There is no electrical isolation between the insertion tube and the handle and the display unit. The display unit has 2 MOPP from mains due to our selected power supply and 2 MOPP from a USB port on the display unit.
1) Based on this, do we consider the entire system to be a BF applied part, even though only the insertion tube necessarily comes in physical contact with the patient during normal use?
2) For temperature limits, we have concerns over the solenoid causing small areas of the handle, not normally touched by the operator, to exceed the 43 degree C temperature limit when a single-fault condition occurs where the controlling mosfet is damaged (shorted) or the MCU is damaged (pin stuck high). This assumes a contact duration of over 10 minutes, even though these areas would not even contact the patient and only be touched by the operator unintentionally. Would we be able to apply less strict temperature requirements for these parts that do not come into contact with the patient and the operator despite it being considered part of the same applied part electrically speaking?
Any help on this matter would be greatly appreciated.
Best regards
There is no electrical isolation between the insertion tube and the handle and the display unit. The display unit has 2 MOPP from mains due to our selected power supply and 2 MOPP from a USB port on the display unit.
1) Based on this, do we consider the entire system to be a BF applied part, even though only the insertion tube necessarily comes in physical contact with the patient during normal use?
2) For temperature limits, we have concerns over the solenoid causing small areas of the handle, not normally touched by the operator, to exceed the 43 degree C temperature limit when a single-fault condition occurs where the controlling mosfet is damaged (shorted) or the MCU is damaged (pin stuck high). This assumes a contact duration of over 10 minutes, even though these areas would not even contact the patient and only be touched by the operator unintentionally. Would we be able to apply less strict temperature requirements for these parts that do not come into contact with the patient and the operator despite it being considered part of the same applied part electrically speaking?
Any help on this matter would be greatly appreciated.
Best regards