Dear Sirs, what is the environmental control that we should have at packaging material warehouse for medical devices, for example pouch, lid, tray for sterile packaging? We are in Malaysia which is high humidity and temperature contry, how significant is their effect to these packaging materials?
Thank you.
G
George Weiss
Your customer's have requirements, as the product , and product's location develop.
A good plan would be to:
1. Ask your customer(s), (The end-products will have storage guidelines, FDA or other).
2. Do some self process analysis.
FDA Basics @
http://www.fda.gov/MedicalDevices/ResourcesforYou/Industry/ucm126252.htm
21 CFR 820.150 @
http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfcfr/CFRSearch.cfm?fr=820.150
Your product(s) will develop into an end product(s), which should have a storage statement,
Example would be an IN VITRO Diagnostic Product, 21 CFR 809.10 @
http://www.fda.gov/downloads/MedicalDevices/DeviceRegulationandGuidance/GuidanceDocuments/UCM081366.pdf
MIREGMGR
Obviously higher temperature and higher RH is friendlier to microbiological contamination...but you are going to sterilize. Certainly you do need to protect materials from contamination to such an extent that it's visually distressed, or physically deteriorated, or has toxic properties. But "controlled" doesn't have to mean 25C and 40%RH.
You're required to conduct all operations that potentially affect the quality of sterile-when-marketed devices in a "controlled" environment. Logically that includes storage of sterile-barrier-packaging materials prior to use. You're also required to validate your processes based on the typical-worst-case environmental conditions applicable to them...and you know that worst case from your controls.
So, if you can validate your sterile-barrier-packaging and sterility processes with materials that have been stored in whatever controlled-but-practical conditions you cause to exist at your location, and if those conditions don't cause other problems that sterilization doesn't remediate, you're good to go.
Ajit Basrur
In addition you should maintain the "integrity" for all materials ie. the packaging materials should be prevented from external dirt and foreign material.
Rolande Hall
Thank you.
It is worth noting that some heat seal coatings used on Tyvek lids are adversely effected by storage at "elevated" temperatures. I have had instances where satisfactory heat sealing was not possible after accidental storage at circa 40 celsius for a few weeks.
Good luck with your project.
P
pangchiaboon
Thank you Sir.
Is there any technical information / article available from Tyvek supplier or any from the market / internet?
arios
Adding to Ajit's great comment, also please consider the prevention of mix-ups, particularly with printed materials, such as labels, IFU's, preprinted lids, etc... You could achive this by maintaining orderly areas, good identification and inventory control.
MIREGMGR
In regard to the particular materials you use, rather than relying on information obtained here, you should contact your supplier for specifically valid data.
ScottK
Not out of the crisis
Hi there Pangciaboon
As has been pointed out the best thing to do is contact your supppliers.
With that in mind: my company prints labels, lidding, and flexible packaging for drug and device companies in the US.
We recommend 70 degrees F at 50% relative humidity for flexible packaging and lidding as well as pressure sensitive labels.
For Thermoplastic Labels we are similar, but with small range: 56-70 deg F, 45%-50% RH.
This is all based on data from our suppliers.
hope that helps some.
S Question about environmental control - Pest control management procedure ISO 13485:2016 - Medical Device Quality Management Systems 5
N ISO 14001 Control vs. Influence - Environmental Aspects Identification ISO 14001:2015 Specific Discussions 6
D How to use temperature and humidity environmental control data? Statistical Analysis Tools, Techniques and SPC 18
B Analytical Lab Environmental Condition (Temperature & Humidity Control Limit) General Measurement Device and Calibration Topics 5
C 21 CFR 820.70 (c) Environmental Control and Pest Control ISO 13485:2016 - Medical Device Quality Management Systems 5
S Environmental Control Standards - Sterile medical device manufacturing environment ISO 13485:2016 - Medical Device Quality Management Systems 3
D On-site Calibration Environmental Control - Is there a correction factor? General Measurement Device and Calibration Topics 5
I Environmental Program - Create a Environmental Control Plan - Example Attached Miscellaneous Environmental Standards and EMS Related Discussions 4
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
F Environmental Requirements E-Lab General Measurement Device and Calibration Topics 0
A Environmental Compliance obligations and risks (ISO 14001:2015 6.1.3) ISO 14001:2015 Specific Discussions 3
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
S Laboratory Environmental Conditions Outside of limits ISO 17025 related Discussions 4
A How to address Environmental & OHS Opportunities? ISO 14001:2015 Specific Discussions 6
T Advice needed - Environmental MS - unwritten but customer requests policy document ISO 14001:2015 Specific Discussions 5
I Environmental Policy Content - ISO 14001 Clause 5.2 ISO 14001:2015 Specific Discussions 2
dhuot1990 Issues with REES Environmental Monitoring System Manufacturing and Related Processes 1
K How to verify the Storage and Transport environmental condition for Medical Device Miscellaneous Environmental Standards and EMS Related Discussions 3
D Templates for Environmental, Safety and quality Management System wanted Document Control Systems, Procedures, Forms and Templates 3
M What is Environmental intelligence specific to IVDs or Medical devices? Miscellaneous Environmental Standards and EMS Related Discussions 1
A Environmental Management System Lead Auditor Competence ISO 14001:2015 Specific Discussions 2
W Organizational Responsibility for Environmental Compliance of Products REACH and RoHS Conversations 4
M PDCA cycle - Process mapping - Environmental ISO 14001 ISO 14001:2015 Specific Discussions 1
V Environmental aspects in a bank ISO 14001:2015 Specific Discussions 7
C Legal or Regulatory Requirements as identified in our Environmental Aspects ISO 14001:2015 Specific Discussions 9
R India - EMS 14001:2015 Environmental product declarations ISO 14001:2015 Specific Discussions 2
C Validation of Applications in a Cloud, CFR 21 part 11 (Environmental Monitoring) Other US Medical Device Regulations 3
A Difference between Environmental Impact and Risk ISO 14001:2015 Specific Discussions 7
R Excel Dashboard for Health Safety and Environmental KPI's for Multiple Sites Occupational Health & Safety Management Standards 4
L Environmental Gas Leak Detection Equipment Recomendations Miscellaneous Environmental Standards and EMS Related Discussions 2
G Sample size for IEC 60601-1-11 environmental tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R ISO 14001 Environmental targets - Office only ISO 14001:2015 Specific Discussions 9
T Guidance - Setting up an integrated quality & environmental management system Career and Occupation Discussions 12
V Looking for a consolidated Manual for QMS, FSMS and Environmental system Food Safety - ISO 22000, HACCP (21 CFR 120) 1
V Looking for a Supplier Evaluation template for Environmental Performance Miscellaneous Environmental Standards and EMS Related Discussions 1
R Mobile Auto Valeting Services - Waste Disposal (Environmental Perspective) Miscellaneous Environmental Standards and EMS Related Discussions 3
G Supplier Significant Environmental Aspects - Audit Finding ISO 14001:2015 Specific Discussions 21
U HSE (Health, Safety and Environmental) Documents and Documentation Occupational Health & Safety Management Standards 1
B Environmental Monitoring for Biopharmaceutical Manufacturing of Bulk Drug Substances Other ISO and International Standards and European Regulations 1
S MIL-HDBK-217 - Understanding the various Environmental Conditions Reliability Analysis - Predictions, Testing and Standards 1
somashekar Industry examples of Significant Beneficial Environmental Impacts Miscellaneous Environmental Standards and EMS Related Discussions 3
G ISO 14001 & Product Design Process Environmental Aspect Review Requirements ISO 14001:2015 Specific Discussions 7
S Format for Significant Environmental Aspects Study ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
B EMP (Environmental Management Program) for an Office Miscellaneous Environmental Standards and EMS Related Discussions 8
M Biological Monitoring Program Environmental Bioburden Test Method Needed ISO 13485:2016 - Medical Device Quality Management Systems 1
R ISO is certified to ISO 14001:2004 - What are their Significant Environmental Aspects ISO 14001:2015 Specific Discussions 2
R Environmental Monitoring in Pharmaceuticals Manufacturing in EU Benchmarking 2
C Environmental Temperature Chamber Testing & A2LA requirements IATF 16949 - Automotive Quality Systems Standard 4
T How to meet the requirements of Environmental Conditions in a Laboratory General Measurement Device and Calibration Topics 11
I Feedback on an Environmental Aspects and Impacts Register Miscellaneous Environmental Standards and EMS Related Discussions 5