Equipment Error and Accuracy to be included in Uncertainty Calculations?
N
neoth
#1
hi all,
if i have already include corrections to my equipment , do i need to include the error & accuracy into uncertainty calculations?
regards
Elsmar Forum Sponsor
D
Daniel Walker - 2011
#2
Very good question.
I would include a component for accuracy or tolerance of your standard, distributed rectangularly, along with applying the correction of the standard.
This way you are allowing for a drift over time of your standard. Assuming that this drift stays within the standard's stated accuracy or tolerance limit.
I'm sure there will be more explanations that are better than mine. This is just my initial take on your question.
Hershal
Metrologist-Auditor
Staff member
Super Moderator
#3
Yes.
Readings that give you a standard deviation is called Type A or random. Standard deviation is best done by Excel unless you want the specifics.
The environmental issues are called Type B, or systemic, as is the calibration uncertainty of the standard. Calibration uncertainty of the standard is typically normal and so divided by 2. Others such as temperature or others are typically rectangular and so divided by square root of 3 which is roughly 1.732.
Of course the formula then goes as normal.
Thread starter Similar threads Forum Replies Date
I Gage Error vs. Dimensional Tolerance - Accuracy needed for measurement equipment Measurement Uncertainty (MU) 10
S How to define Permissible Error for Measurement and Test Equipment? Measurement Uncertainty (MU) 3
6 Gage R&R on automated test equipment - No reproducibility error? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J Reproducibility error between instruments for automatic measuring equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
A Canadian Safety Labels for medical electrical equipment Canada Medical Device Regulations 0
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
P Equipment URS, pFMEA and dFMEA ISO 14971 - Medical Device Risk Management 3
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
L Gage RandR on automated equipment. IATF 16949 - Automotive Quality Systems Standard 7
W How long do you keep information about equipment no longer used? Document Control Systems, Procedures, Forms and Templates 2
B 4DINSPEC Inspection Equipment General Measurement Device and Calibration Topics 0
D Equipment Register and PM question ISO 13485:2016 - Medical Device Quality Management Systems 2
R Classification of USB DC operated medical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
J Confusing on Harmonic Emissions/Voltage Fluctuations/Flicker test for CISPR 11 Class A Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C 8.5.1.1 Control of Equipment, Tools, and Software Programs - Questions about the extent of control of NC programs AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Equipment Register related question ISO 13485:2016 - Medical Device Quality Management Systems 1
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
I Equipment service dates ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
M Missing measuring equipment General Measurement Device and Calibration Topics 6
P Equipment 21 CFR 820.70(g) - User Requirements Document for Off the shelf equipment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
K Problems with process equipment ISO 13485:2016 - Medical Device Quality Management Systems 4
D New Equipment form / template Document Control Systems, Procedures, Forms and Templates 5
W Protective earth for Class I ME equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D DQ documents of new equipment - Function specification Qualification and Validation (including 21 CFR Part 11) 0
R MSA studies for Heat Treatment Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
T AMS2570E certification equipment questions AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
J Class Fixed ME Equipment (US Deviation) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
P Manufacturing equipment (e.g.: Dental Welder): medical device or not? US Food and Drug Administration (FDA) 5
R What is meant by Use of COMPONENTS WITH HIGH-INTEGRITY CHARACTERISTICS in ME EQUIPMENT (clause no 4.9) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Infrastructure: Equipment, Work Environment and Contamination Control ISO 13485:2016 - Medical Device Quality Management Systems 4
M Are Electrical Laparoscopic Surgical Devices ME equipment? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
D Definition of "Critical Laboratory Equipment"? General Measurement Device and Calibration Topics 1
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
T How to “validate/qualify” a new measurement equipment? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
D Software validation in Medical Equipment Other Medical Device and Orthopedic Related Topics 20
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
A Handling of contaminated medical equipment Other Medical Device and Orthopedic Related Topics 0
T Confirming The Local Safety Regulations for Equipment - I'm in Malaysia General Measurement Device and Calibration Topics 1
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
E Equipment Qualification - IQ/OQ per ISO 13485:2016 section 7.5.6 Process validation ISO 13485:2016 - Medical Device Quality Management Systems 7
A Contraindication Safety Statement in website - Radiotherapy equipment Other Medical Device and Orthopedic Related Topics 1
B What is 'equipment'? Qualification and Validation (including 21 CFR Part 11) 2
Similar threads
Top Bottom