Equipment for Oven Calibration, Uniformity and System Accuracy Tests

wesatwork

Learning what I can.
#1
Hello,

This whole NADCAP/AMS2750 field is new to me so please forgive my ignorance.

We are performing oven (mostly grieve ovens) controller calibrations, uniformity surveys and system accuracy test in house with some outdated homegrown equipment.

I am looking for a data acquisition system that can graphically display real time thermocouple inputs (about 16). Store the calibration data and fairly easily download it. Ideally an off the shelf system geared for these types of calibrations.

Any recommendations?
 
Elsmar Forum Sponsor
M

MTProcessing

#2
Datapaq and Phoenix are two systems I am aware of. They have products that can go entirely inside the furnace (with thermal barriers to provide several hours of protection to the electronics) so no wires have to be run through the furnace door or wall.

If your setup is fine running wires through an opening you probably have additional cheaper options, I'm just not sure of other name brands.
 

wesatwork

Learning what I can.
#3
MTProcessing thanks for the help. I'm looking at Datapaq (Fluke) now. I have not been able to find the Phoenix system you mentioned.
 
Thread starter Similar threads Forum Replies Date
M Cameras and Visual Inspection Equipment General Measurement Device and Calibration Topics 3
JoeRandom11 Question about Customer Supplied Equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
C Equipment IQ and OQ ISO 13485:2016 - Medical Device Quality Management Systems 6
D PPE (Personal Protective Equipment) in the UK CE Marking (Conformité Européene) / CB Scheme 1
D Recommendations for equipment used to validate cold storage units General Measurement Device and Calibration Topics 3
T PFMEA vs MFMEA or DFMEA on equipment/machinery FMEA and Control Plans 19
E Brass in medical equipment Other Medical Device and Orthopedic Related Topics 5
G QSB+ : capability of measurement equipment IATF 16949 - Automotive Quality Systems Standard 15
M Gage R&R automated test equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
G Developing our custom test methods and test equipment US Medical Device Regulations 5
E Compressed air quality in ME equipment (used in OR) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M ISO 17025:2017 Clause 6.4 - Measuring Equipment Validity ISO 17025 related Discussions 2
W Strategy for determining which components from a system should be "ME EQUIPMENT" -- home healthcare environment IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Do i need to have equipment validation if 100% testing is completed? Qualification and Validation (including 21 CFR Part 11) 6
K Brazil (INMETRO) for Medical Electrical and IVD Equipment? Other Medical Device Regulations World-Wide 0
Y Creating Equipment qualification protocol Qualification and Validation (including 21 CFR Part 11) 2
R Suggest modern equipment/instrument for plastic bottle inspection? Quality Tools, Improvement and Analysis 1
S Certification of equipment after move General Measurement Device and Calibration Topics 14
B Building my own measurement equipment General Measurement Device and Calibration Topics 3
J Process Capability - Determining the process capability of certain equipment Lean in Manufacturing and Service Industries 6
W Classes and Applied Part IEC60601 Electromedical Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A PPAP Alternate Equipment IATF 16949 - Automotive Quality Systems Standard 10
Q Measurement Equipment Revocation - Looking for a Disposal Form with Risk Assessment IATF 16949 - Automotive Quality Systems Standard 10
E Comparison of Personal Protective Equipment Standards EN 14605 vs GB24539 and ISO 13982 vs GB 24539 Occupational Health & Safety Management Standards 0
E Comparison of Respiratory Protective Equipment Standards (China Standard Compare with EU standard of EN149 Occupational Health & Safety Management Standards 2
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
P 7.5.10 Customer property - applies to leased/rented equipment? ISO 13485:2016 - Medical Device Quality Management Systems 10
S Class I Equipment with mobile power inverter IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W Equipment Bracketing Approach US Food and Drug Administration (FDA) 0
H Still having trouble as to what equipment falls under 8.5.1.1 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Register of the monitoring and measuring equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
B Do CTQs need to be equipment specific? FMEA and Control Plans 8
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
C Revalidation of Process Equipment for Equipment Transfered to New Facility. ISO 13485:2016 - Medical Device Quality Management Systems 5
V Clause 7.1.3.1 Plant, facility, and equipment planning Lean in Manufacturing and Service Industries 0
A Canadian Safety Labels for medical electrical equipment Canada Medical Device Regulations 3
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
P Equipment URS, pFMEA and dFMEA ISO 14971 - Medical Device Risk Management 3
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
L Gage RandR on automated equipment. IATF 16949 - Automotive Quality Systems Standard 7
W How long do you keep information about equipment no longer used? Document Control Systems, Procedures, Forms and Templates 2
B 4DINSPEC Inspection Equipment General Measurement Device and Calibration Topics 0
D Equipment Register and PM question ISO 13485:2016 - Medical Device Quality Management Systems 2
R Classification of USB DC operated medical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1

Similar threads

Top Bottom