Equipment, Software and Processes Validation for Medical Devices - 21CFR 820.75 (a)

C

chasf

Can anyone tell me if it is common to validate the the equipment, software and processes for the manufacturing of a medical device? This will soon become a requirement of one of our customers they tell us it is being fueled by the FDA. My understanding was that Validation was only required under 21CFR 820.75 (a).

Sec. 820.75 Process validation.

(a) Where the results of a process cannot be fully verified by subsequent
inspection and test, the process shall be validated with a high degree of
assurance and approved according to established procedures. The
validation activities and results, including the date and signature of the
individual(s) approving the validation and where appropriate the major
equipment validated, shall be documented.
 

Al Rosen

Leader
Super Moderator
Re: Validation for Medical Devices

Can anyone tell me if it is common to validate the the equipment, software and processes for the manufacturing of a medical device? This will soon become a requirement of one of our customers they tell us it is being fueled by the FDA. My understanding was that Validation was only required under 21CFR 820.75 (a).

Sec. 820.75 Process validation.

(a) Where the results of a process cannot be fully verified by subsequent
inspection and test, the process shall be validated with a high degree of
assurance and approved according to established procedures. The
validation activities and results, including the date and signature of the
individual(s) approving the validation and where appropriate the major
equipment validated, shall be documented.
It is common and FDA will want you to validate your processes. Also look at the GHTF Guidance for Process Validation.
 

Attachments

  • GHTF process validation guidance.pdf
    180.4 KB · Views: 833

yodon

Leader
Super Moderator
Re: Validation for Medical Devices

Note that 820.75 is under subpart G, Production (aka manufacturing) and Process Controls. So, as Al points out, validation will be expected.

One area (I'm sure there are others) that seems to be hit often is soldering. You can't 100% verify the results without destruction so soldering is expected to be a validated process. Another area is sterilization. If the device is sterilized, the sterilization process must be validated.
 

Al Rosen

Leader
Super Moderator
Re: Validation for Medical Devices

Note that 820.75 is under subpart G, Production (aka manufacturing) and Process Controls. So, as Al points out, validation will be expected.

One area (I'm sure there are others) that seems to be hit often is soldering. You can't 100% verify the results without destruction so soldering is expected to be a validated process. Another area is sterilization. If the device is sterilized, the sterilization process must be validated.
There is also a list of processes in the guidance that I posted. It's not comprehensive but it's a start.
 
C

chasf

Re: Equipment, Software and Processes Validation for Medical Devices - 21CFR 820.75 (

Thanks for your input. It makes sense to apply 820.75 to a soldering process but we are being required to validate all processes such as turning, milling and drilling. These processes can be measured and verified. Also the software used in the CNC machinery is to be validated as well. Does this sound reasonable?
 

Al Rosen

Leader
Super Moderator
Re: Equipment, Software and Processes Validation for Medical Devices - 21CFR 820.75 (

Thanks for your input. It makes sense to apply 820.75 to a soldering process but we are being required to validate all processes such as turning, milling and drilling. These processes can be measured and verified. Also the software used in the CNC machinery is to be validated as well. Does this sound reasonable?
If you want to inspect or verify every characteristic rather than validate, you may be able to make an argument, but you should at least validate your CNC software. BTW, reasonable and regulations are sometimes oxymorons. See 21CFR820.70(i).
 
Last edited:
C

chasf

Re: Equipment, Software and Processes Validation for Medical Devices - 21CFR 820.75 (

Verification of characteristicis is done by using a sampling plan during production. Does that sound reasonable? Also I should make it clear that we subcontract for Medical Device companies the product is primarily implantable devices.
 
R

Roland Cooke

Re: Equipment, Software and Processes Validation for Medical Devices - 21CFR 820.75 (

Validation of processes is basically done for one of three reasons:

1) You can't test/inspect for it, so you have to validate. Sterilisation is the obvious example here.

2) You can test/inspect for it, but it is more economical to have a validated process instead. Automated camera inspection systems are a good example.

3) The process doesn't need validation (presumably because of subsequent inspection), but the company opts to validate every process anyway, as a belt 'n' braces approach.
 
C

chasf

Re: Equipment, Software and Processes Validation for Medical Devices - 21CFR 820.75 (

That makes good sense Roland but what I am getting form one customer in particular is that all Processes including all subcontracted processes must be validated. This includes software and installation. Do you have any recommendations?
 
R

Roland Cooke

Re: Equipment, Software and Processes Validation for Medical Devices - 21CFR 820.75 (

Then your customer is an idiot. It happens. :tg:

Do you actually install anything? If not, it makes validating that process really rather hard. :D



Hopefully your customer has merely shortcut what he really means for you to have in place.

What you need to do is to map your processes and identify how you know each of them are going to work to an agreed level of satisfaction. That will either be via inspection/testing, or by validation.

(By the way, software isn't a process. Software might well control a process, and thus may well need to be validated to ensure that it performs the control task correctly. Software may need validation in its own right of course, some medical devices are indeed simply CD-ROMs.)

edit: if you want to know more about validating processes, search this forum, and the FDA website. There is a ton of stuff available.
 
Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
V PQ parameters for computer system validation of configurable software of laboratory equipment Qualification and Validation (including 21 CFR Part 11) 0
C 8.5.1.1 Control of Equipment, Tools, and Software Programs - Questions about the extent of control of NC programs AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
D Software validation in Medical Equipment Other Medical Device and Orthopedic Related Topics 20
S Validation of COTS Equipment plus Software Qualification and Validation (including 21 CFR Part 11) 12
Q Calibrated Equipment Managment Software Suggestions General Measurement Device and Calibration Topics 12
R FDA Requirements - Printing Equipment Software Validation Qualification and Validation (including 21 CFR Part 11) 1
P Software Validation for Equipment - Question ISO 13485:2016 - Medical Device Quality Management Systems 5
C Software Planning Tool for Equipment Qualifications, Process Validations Software Quality Assurance 3
R CLIA Equipment or Software Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
F Aged Inspection Equipment Software has no Audit Trail - Validation of Software Qualification and Validation (including 21 CFR Part 11) 5
S OEE Software - Automating the collection of Overall Equipment Efficiency data Statistical Analysis Tools, Techniques and SPC 8
W Computer Software - ISO 9001 Clause 7.6 Control of Monitoring and Measuring equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibrating Electronic Equipment Tester Software - ISO 9001 Clause 7.6 Benchmarking 1
W Software development company for Medical Device Diagnostic Equipment ISO 13485:2016 - Medical Device Quality Management Systems 5
J How & how often to perform verification of Test Software of Automatic Test Equipment? General Measurement Device and Calibration Topics 3
M ISO 9001 - 4.11.1, Paragraph 3 - Test Equipment and Test Software General Measurement Device and Calibration Topics 1
I Design Verification Testing on Non-Validated Equipment ISO 13485:2016 - Medical Device Quality Management Systems 2
E Class II ME EQUIPMENT and subclause 8.9.1.12 IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
H MTBF industry average for manufacturing automated equipment Manufacturing and Related Processes 0
M Master / Guideline Matrix for measuring equipment Manufacturing and Related Processes 4
K Equipment IQ/OQ/PQ required when you have 100% product inspection? ISO 13485:2016 - Medical Device Quality Management Systems 2
K ISO9001 - Measurement Equipment records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
K ISO 17025 Query Relating to Monitoring of Equipment ISO 17025 related Discussions 3
R Medical equipment selection for refurbishment US Food and Drug Administration (FDA) 1
R CE Certification / PED Certification - Pressure Equipment CE Marking (Conformité Européene) / CB Scheme 4
H DIN 12880:2007-05 Clause 7.2 Test Equipment Description CE Marking (Conformité Européene) / CB Scheme 0
B Painted surface of a protective Class I equipment. CE Marking (Conformité Européene) / CB Scheme 1
K What are the differences between the certificates required for home and medical medical equipment? US Food and Drug Administration (FDA) 1
B ISO 13485 Qualification of R&D Equipment Qualification and Validation (including 21 CFR Part 11) 1
R EU 2017-745 - Dental Equipment CE Marking (Conformité Européene) / CB Scheme 4
T Need advice on what inspection equipment to buy. General Measurement Device and Calibration Topics 2
Q What is considered "Multimedia equipment" (60601-1-2 section 7.1.3) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
V Is it required to recalibrate reference- standard instruments after equipment qualification General Measurement Device and Calibration Topics 0
armani Monitoring and measuring equipment for occupational health and safety purposes Occupational Health & Safety Management Standards 10
P Type of MSA study according to measurement equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Rene Minassian Control of (TMME) Testing and Measurement Equipment Oil and Gas Industry Standards and Regulations 2
M Cameras and Visual Inspection Equipment General Measurement Device and Calibration Topics 3
JoeRandom11 Question about Customer Supplied Equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
C Equipment IQ and OQ ISO 13485:2016 - Medical Device Quality Management Systems 9
D PPE (Personal Protective Equipment) in the UK CE Marking (Conformité Européene) / CB Scheme 1
D Recommendations for equipment used to validate cold storage units General Measurement Device and Calibration Topics 3
T PFMEA vs MFMEA or DFMEA on equipment/machinery FMEA and Control Plans 19
E Brass in medical equipment Other Medical Device and Orthopedic Related Topics 5
G QSB+ : capability of measurement equipment IATF 16949 - Automotive Quality Systems Standard 15
M Gage R&R automated test equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
G Developing our custom test methods and test equipment US Medical Device Regulations 5
E Compressed air quality in ME equipment (used in OR) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M ISO 17025:2017 Clause 6.4 - Measuring Equipment Validity ISO 17025 related Discussions 2
W Strategy for determining which components from a system should be "ME EQUIPMENT" -- home healthcare environment IEC 60601 - Medical Electrical Equipment Safety Standards Series 6

Similar threads

Top Bottom