Hello,
If a particular standard defines essential performance as providing therapy above limit X or generation of a technical alarm, and in the same particular standard, under electromagnetic disturbances addition general requirements section (typical 202.8.1.101) the standard states,
"The device may exhibit temporary degradation of performance (e.g. deviation from the performance indicated in the instructions for use during IMMUNITY testing) that does not affect BASIC SAFETY or ESSENTIAL PERFORMANCE".
Is it acceptable during EMC immunity testing that the device fails to deliver therapy above limit X but the device raises a technical alarm?
Thanks a lot of any input!
Elsmar Forum Sponsor
Benjamin Weber
Trusted Information Resource
#3
If the particular standard does define the EP only as "output shall not exceed limit X", this has to be passed during the EMC testing in my oppinion.
Other particular standard explicitely say "value shall not be below/above limit X OR an alarm shall be raised". Regarding the EP of the x-ray output I think the alarm is not sufficient, because the x-ray output can lead to severe harm of the patient in very short time. Alarms for degradation of EP usually are OK for performances, where after the alarm is raised the operator can act quickly in order to prevent the potential severe/catastrophic harm.
david316
#4
Thanks. In this case the committee have defined EP as providing therapy above limit X or generation of a technical alarm. Given the definition of EP, one would assume that if you meet this EP the risk is acceptable. One would also assume that an alarm is required in order to make the risk acceptable if the device fails to deliver therapy above limit X.
But, the test regarding EMC immunity is worded vaguely. It states...
My understanding is EMC immunity testing is considered a "normal condition". So if during EMC immunity testing the device fails to provide therapy above level X but alarms, and this is considered a PASS as EP is still maintained, doesn't this imply it is acceptable to fail to deliver therapy as long as you alarm under "normal conditions"?
Thread starter Similar threads Forum Replies Date
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
E Particular standards, essential performance in EMC Immunity Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Link Between Essential Performance Requirements and Essential Design Outputs 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Informational TGA Consultation: Proposed changes to medical device essential principles for safety and performance Medical Device and FDA Regulations and Standards News 0
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Essential Performance of a Cone Beam Computed Tomography Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D IEC 60601-1 - Essential performance doesn't make sense IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
D IEC 60601-1 - Performance limits for essential performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Essential performance, accompanying documents....confused IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K IEC 60601-1 and Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M IEC 60601 - Limits of agreement as Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Essential Performance of Mechanical Device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
D IEC 60601-2-44: 202.101 Immunity Testing of Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L "Potential" Essential Performance in IEC 60601-2-54 (Definition) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
M "Basic Safety" versus "Essential Performance" IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
I Defending "No Essential Performance" for a Surgical Video Camera IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
L ISO 14971 - Annex C question C.2.34 "Essential performance" ISO 14971 - Medical Device Risk Management 2
A Essential Performance in IEC 3rd edition of the 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
U Defining Essential Performance to Achieve Freedom from Unacceptable Risk IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
P Essential MDR documents IEC 62366 - Medical Device Usability Engineering 0
L Essential Design Outputs for SAMD 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
normhowe Why is RCA essential? Problem Solving, Root Cause Fault and Failure Analysis 19
Bev D Essential References for Practical Quality Engineering Misc. Quality Assurance and Business Systems Related Topics 0
C Essential Design Output(s) and Design transfer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
K Comparison essential requirements EU compared to those of MDSAP countries Other Medical Device Regulations World-Wide 3
M Informational TGA – Essential principles checklist (medical devices) Medical Device and FDA Regulations and Standards News 2
M Brazil and Japan - Essential requirements Other Medical Device Regulations World-Wide 3
G Essential Requirements Checklist - (MDD 93/42 / EEC) Misc. Quality Assurance and Business Systems Related Topics 2
M Informational TGA Webinar: Essential Principles to software Medical Device and FDA Regulations and Standards News 0
Z In which country is essential to have and IEC 60601 CB Report? Other Medical Device Related Standards 0
C CMDR Essential (Safety & Effectiveness) Requirements Checklist? Canada Medical Device Regulations 9
D Essential Requirements Checklist for Standalone Software CE Marking (Conformité Européene) / CB Scheme 10
Rincewind Pressure energy - Essential Requirements (MDD/MDR) EU Medical Device Regulations 2
N Essential requirements checklist needed? ISO 13485:2016 - Medical Device Quality Management Systems 4
S Medical Device Changes and How to Assess Their Impacts to Essential Requirements EU Medical Device Regulations 2
A Essential Requirement Checklist For Medical Device CE Marking (Conformité Européene) / CB Scheme 4
A Policies Mandatory or essential for ISO 27001 implementation IEC 27001 - Information Security Management Systems (ISMS) 6
K Essential Requirements Checklist references EU Medical Device Regulations 10
K New Product Essential Requirements Cheat Sheet EU Medical Device Regulations 6
A Regarding Essential Requirements Checklist ISO 13485:2016 - Medical Device Quality Management Systems 9
M Is it essential to do the Fractional Test for ETO Revalidation? ISO 13485:2016 - Medical Device Quality Management Systems 1
D What are the Essential Quality Books in the last 15 years Book, Video, Blog and Web Site Reviews and Recommendations 3
G Audit of essential QMS processes General Auditing Discussions 11
somashekar GMP (Good Manufacturing Practices) essential information URL Manufacturing and Related Processes 1
R Minimum Essential Receiving Inspection (M.E.R.I.) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
J Project Specific Documentation - Completion and filing of all essential records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Essential Requirements Checklist w/o ISO 13485 CE Marking (Conformité Européene) / CB Scheme 7
S Innovation in ER (Essential Requirements) Checklists EU Medical Device Regulations 6
Similar threads
Top Bottom