Essential Performance of Mechanical Device?

R

reyemka

#1
Looking for opinions on this question:

I have a prototype where all the intended use is provided via an approved mechanical device (no electronics). We've added electronics to the device to capture usage data for a clinical study. The electronics do not contribute to the intended use, and do not affect the clinical performance of the device.

We are reviewing 60601-1 to ensure the device is safe to use, and I'm wondering if, according to the standard, it would be considered to have essential performance?

I'm inclined to say no, but looking for your thoughts.

Thanks!
 
Elsmar Forum Sponsor
T

Thukira

#2
Extract from IEC 60601-1, 3.1 edition (EP, cl. 3.27*) definition
* ESSENTIAL PERFORMANCE
performance of a clinical function, other than that related to BASIC SAFETY, where loss or degradation beyond the limits specified by the MANUFACTURER results in an unacceptable RISK .
NOTE ESSENTIAL PERFORMANCE is most easily understood by considering whether its absence or degradation would result in an unacceptable RISK.

EP process (Cl. 4.3*)
· During RISK ANALYSIS, the MANUFACTURER shall identify the EP of the MEE&S (1st para of 4.3)
· The MANUFACTURER shall specify performance limits between fully functional & total loss of the identified performance in both the NORMAL CONDITION & SINGLE FAULT CONDITION
· Follow full process of Cl. 4.3

And also;
· EP will impact test plan (Safety, EMC/I test etc)
· Changes in intended use͟ can change EP of device
· Particular Standards (60601-2-XX & 80601-2-XX) tend to have EP

RISK MANAGEMENT PROCESS (RMP) (4.2.2)
· RMP complying with ISO 14971 shall be performed & required
· Ubiquitous in standard.
· Therefore you can’t comply with IEC 60601-1 ed. 3.0 or 3.1 without complying with ISO 14971
 

Peter Selvey

Staff member
Moderator
#5
Based on the original description there is no essential performance. Someone could try to argue that the feedback data could lead to improvements in future device design hence they could be essential, but that's clearly a step too far with the interpretation. Essential performance should be limited to the actual device on the market (in use), not possible future devices.

The catch is design creep. I have seen cases where monitoring functions are deemed not essential, just diagnostic feedback for designers. Later someone says hey how about we do this, and it eventually evolves into a important clinical function. But, the designers continue to deem it non-essential based on the original decision.

Keep that in mind and you should be OK.
 
Last edited by a moderator:

VinceTech

Involved In Discussions
#6
Based on the original description there is no essential performance. Someone could try to argue that the feedback data could lead to improvements in future device design hence they could be essential, but that's clearly a step too far with the interpretation. Essential performance should be limited to the actual device on the market (in use), not possible future devices.

The catch is design creep. I have seen cases where monitoring functions are deemed not essential, just diagnostic feedback for designers. Later someone says hey how about we do this, and it eventually evolves into a important clinical function. But, the designers continue to deem it non-essential based on the original decision.

Keep that in mind and you should be OK.


Further discussion. Assuming the electronics failure can lead to the mechanical device connects to supply mains, the protection for this failure mode is earthing the mechanical device. In this case, is the proper earthing of device is an essential performance? Thanks

Best regards,
 
Thread starter Similar threads Forum Replies Date
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Informational TGA Consultation: Proposed changes to medical device essential principles for safety and performance Medical Device and FDA Regulations and Standards News 0
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Essential Performance of a Cone Beam Computed Tomography Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D IEC 60601-1 - Essential performance doesn't make sense IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
D IEC 60601-1 - Performance limits for essential performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Essential performance, accompanying documents....confused IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K IEC 60601-1 and Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M IEC 60601 - Limits of agreement as Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D IEC 60601-2-44: 202.101 Immunity Testing of Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L "Potential" Essential Performance in IEC 60601-2-54 (Definition) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
E Particular standards, essential performance in EMC Immunity Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M "Basic Safety" versus "Essential Performance" IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
I Defending "No Essential Performance" for a Surgical Video Camera IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
L ISO 14971 - Annex C question C.2.34 "Essential performance" ISO 14971 - Medical Device Risk Management 2
A Essential Performance in IEC 3rd edition of the 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
U Defining Essential Performance to Achieve Freedom from Unacceptable Risk IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
N Why is RCA (Root Cause Analysis) essential? Problem Solving, Root Cause Fault and Failure Analysis 23
Bev D Essential References for Practical Quality Engineering Misc. Quality Assurance and Business Systems Related Topics 0
C Essential Design Output(s) and Design transfer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
K Comparison essential requirements EU compared to those of MDSAP countries Other Medical Device Regulations World-Wide 3
M Informational TGA – Essential principles checklist (medical devices) Medical Device and FDA Regulations and Standards News 2
M Brazil and Japan - Essential requirements Other Medical Device Regulations World-Wide 3
G Essential Requirements Checklist - (MDD 93/42 / EEC) Misc. Quality Assurance and Business Systems Related Topics 2
M Informational TGA Webinar: Essential Principles to software Medical Device and FDA Regulations and Standards News 0
Z In which country is essential to have and IEC 60601 CB Report? Other Medical Device Related Standards 0
C CMDR Essential (Safety & Effectiveness) Requirements Checklist? Canada Medical Device Regulations 9
D Essential Requirements Checklist for Standalone Software CE Marking (Conformité Européene) / CB Scheme 10
Rincewind Pressure energy - Essential Requirements (MDD/MDR) EU Medical Device Regulations 2
N Essential requirements checklist needed? ISO 13485:2016 - Medical Device Quality Management Systems 4
S Medical Device Changes and How to Assess Their Impacts to Essential Requirements EU Medical Device Regulations 2
A Essential Requirement Checklist For Medical Device CE Marking (Conformité Européene) / CB Scheme 4
A Policies Mandatory or essential for ISO 27001 implementation IEC 27001 - Information Security Management Systems (ISMS) 6
K Essential Requirements Checklist references EU Medical Device Regulations 10
K New Product Essential Requirements Cheat Sheet EU Medical Device Regulations 6
A Regarding Essential Requirements Checklist ISO 13485:2016 - Medical Device Quality Management Systems 9
M Is it essential to do the Fractional Test for ETO Revalidation? ISO 13485:2016 - Medical Device Quality Management Systems 1
D What are the Essential Quality Books in the last 15 years Book, Video, Blog and Web Site Reviews and Recommendations 3
G Audit of essential QMS processes General Auditing Discussions 11
somashekar GMP (Good Manufacturing Practices) essential information URL Manufacturing and Related Processes 1
R Minimum Essential Receiving Inspection (M.E.R.I.) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
J Project Specific Documentation - Completion and filing of all essential records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
J Essential Requirements Checklist w/o ISO 13485 CE Marking (Conformité Européene) / CB Scheme 7
S Innovation in ER (Essential Requirements) Checklists EU Medical Device Regulations 6
S Canada - Essential Requirements Checklist for MDD 93/42/EEC Canada Medical Device Regulations 8
E Question on approval of the EU Essential Requirements Matrix - MDD Requirements EU Medical Device Regulations 1
5 Completed Medical Device Essential Requirements Matrix CE Marking (Conformité Européene) / CB Scheme 1
AnaMariaVR2 The Essential Guide to Combination Products cGMPs Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
A Essential Design Output - Still a problem 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7

Similar threads

Top Bottom