Roland Cooke

#5
I have seen hundreds of these (completed) checklists over the years, and to be honest, the vast majority of them have been a waste of time. They usually don't answer the question, and they usually don't allow me to find the supporting documentation, or at least not all of it, easily.

So while there is nothing wrong with the template, do make sure you use it properly.

What I tend to recommend, if you want to go the extra mile, is to turn the whole thing 90 degrees, so that there is proper space for a detailed response to each ER.

a) justification of non-applicability
b) confirmation of which general harmonised standards are applicable, and if not met in full, justifications for non-applicability of the sections or the whole standard (EN13485, EN14971, EN10993-1, EN980, EN1041 etc)
c) confirmation of which product-specific harmonised standards are applicable, and if not met in full, justifications for non-applicability of the sections or the whole standard
d) Short description (text) of how all the elements of the individual ER have been addressed met, if not already covered above
e) Location of supporting documentation held within the Technical File
f) Location of supporting documentation not held within the Technical File
 
S

Sintolt

#8
There seems to be a new requirement under 7.1:

"- where appropriate, the results of biophysical or modeling research whose validity has been demonstrated beforehand"

What sort of documents / tests can be refernenced here?
 
B

bobthom

#9
Ian, thanks for the document, o my checking it looks preety much close to the mark.

Regards,
Bob:thanx:
 
A

AC Cobra

#10
I had a few problems locating a word format template for an Essential Requirements Checklist which was compliant with MDD 93/42/EC as amended by 2007/47/EC.
After going blind by typing in such a small font size, here it is.
If anyone can see any mistakes, please let me know.

Thanks
Ian
:thanx: I've been looking for this checklist on the web and now I found it!! Thanks a lot! AC Cobra
 
Thread starter Similar threads Forum Replies Date
G Essential Requirements Checklist - (MDD 93/42 / EEC) Misc. Quality Assurance and Business Systems Related Topics 2
C CMDR Essential (Safety & Effectiveness) Requirements Checklist? Canada Medical Device Regulations 9
D Essential Requirements Checklist for Standalone Software CE Marking (Conformité Européene) / CB Scheme 10
N Essential requirements checklist needed? ISO 13485:2016 - Medical Device Quality Management Systems 4
K Essential Requirements Checklist references EU Medical Device Regulations 10
A Regarding Essential Requirements Checklist ISO 13485:2016 - Medical Device Quality Management Systems 9
J Essential Requirements Checklist w/o ISO 13485 CE Marking (Conformité Européene) / CB Scheme 7
S Canada - Essential Requirements Checklist for MDD 93/42/EEC Canada Medical Device Regulations 8
E Developing the Essential Requirements Checklist - Identifying Requirements Other US Medical Device Regulations 3
bio_subbu Essential Requirements Checklist for AIMDD 90/385/EEC CE Marking (Conformité Européene) / CB Scheme 2
A Essential Requirements Checklist for PPE (Personal Protective Equipment) CE Marking (Conformité Européene) / CB Scheme 6
P Essential Requirements checklist as per unrevised MDD (MDD 93/42/EEC) EU Medical Device Regulations 7
T MDD Essential Requirements Checklist in accordance with MDD 2007/47/EC EU Medical Device Regulations 3
A Technical Agreement for the CE Mark - Essential Requirements Checklist EU Medical Device Regulations 7
W MDD (Medical Device Directive) Essential Requirements Checklist ISO 13485:2016 - Medical Device Quality Management Systems 18
K Comparison essential requirements EU compared to those of MDSAP countries Other Medical Device Regulations World-Wide 3
M Brazil and Japan - Essential requirements Other Medical Device Regulations World-Wide 3
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Rincewind Pressure energy - Essential Requirements (MDD/MDR) EU Medical Device Regulations 2
S Medical Device Changes and How to Assess Their Impacts to Essential Requirements EU Medical Device Regulations 2
K New Product Essential Requirements Cheat Sheet EU Medical Device Regulations 6
S Innovation in ER (Essential Requirements) Checklists EU Medical Device Regulations 6
E Question on approval of the EU Essential Requirements Matrix - MDD Requirements EU Medical Device Regulations 1
5 Completed Medical Device Essential Requirements Matrix CE Marking (Conformité Européene) / CB Scheme 1
S MDD 93/42/EEC Essential Requirements and Safety definition CE Marking (Conformité Européene) / CB Scheme 2
E Essential Requirements for each type of Risk EU Medical Device Regulations 11
R ISO IEC EN 62366 Usability Engineering Essential Requirements IEC 62366 - Medical Device Usability Engineering 8
M How to write the Essential Requirements EU Medical Device Regulations 14
I Essential Requirements for Medical Devices - Word copy wanted EU Medical Device Regulations 4
E Essential Requirements - Example of the essential requirements filled in ISO 13485:2016 - Medical Device Quality Management Systems 3
T Essential Requirements Of A Technical File - ISO13485 / Health Canada / MDD ISO 13485:2016 - Medical Device Quality Management Systems 6
J IVD Directive - Review and Update Essential Requirements Documentation EU Medical Device Regulations 5
T Harmonised Standard and Essential Requirements ISO 13485:2016 - Medical Device Quality Management Systems 3
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
normhowe Why is RCA essential? Problem Solving, Root Cause Fault and Failure Analysis 19
Bev D Essential References for Practical Quality Engineering Misc. Quality Assurance and Business Systems Related Topics 0
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C Essential Design Output(s) and Design transfer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Informational TGA – Essential principles checklist (medical devices) Medical Device and FDA Regulations and Standards News 2
M Informational TGA Consultation: Proposed changes to medical device essential principles for safety and performance Medical Device and FDA Regulations and Standards News 0
M Informational TGA Webinar: Essential Principles to software Medical Device and FDA Regulations and Standards News 0
B Essential Performance of a Cone Beam Computed Tomography Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Z In which country is essential to have and IEC 60601 CB Report? Other Medical Device Related Standards 0
D IEC 60601-1 - Essential performance doesn't make sense IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
D IEC 60601-1 - Performance limits for essential performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Essential performance, accompanying documents....confused IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K IEC 60601-1 and Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M IEC 60601 - Limits of agreement as Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Essential Requirement Checklist For Medical Device CE Marking (Conformité Européene) / CB Scheme 4

Similar threads

Top Bottom