Ethylene Oxide Sterile Implantable Medical Device Shelf Life Extension

D

dallltho

#1
If you have a sterile implantable medical device that is sterilized by Ethylene Oxide (EO aka ETO) and want to extend the shelf life, how many product batches or lots would need to be verified via testing in order to justify the extension? Is there a general best practice? If so, which standard or regulation or guidance would direct us to verify such a number of lots to test (for example 1 lot or 3 lots etc.)?
 
Elsmar Forum Sponsor

planB

Super Moderator
#2
dalltho,

follow ISO 11607, section 6.4. You have to justify your sample size for your seal- and packaging-integrity tests at the extended time points by a statistical rationale. Typically, you demonstrate batch-wise reproducibility of your sealing process during the PQ of your initial sealing validation at timepoint = 0, and thus, do not have to consider batch variations as a variable during aging/stability studies.

HTH,

Gerhard
 
Thread starter Similar threads Forum Replies Date
G Number of Destructively Tested Devices Needed for Ethylene Oxide Validation Other Medical Device Related Standards 4
M Informational USFDA Ethylene Oxide Sterilization for Medical Devices website Medical Device and FDA Regulations and Standards News 0
M Informational FDA Posts Information about Ethylene Oxide Sterilization and Medical Devices Medical Device and FDA Regulations and Standards News 0
T ISO 10993-7 - Ethylene oxide sterilization residues Other Medical Device Related Standards 1
E Ethylene Oxide Sterilization Validation-Reference Load Other Medical Device Related Standards 8
D Ethylene Oxide Sterilization aeration Manufacturing and Related Processes 1
B Product definition for Ethylene Oxide Sterilization Validation Other Medical Device Related Standards 2
D Adding an extra Label on an Ethylene Oxide Sterilized Box ISO 13485:2016 - Medical Device Quality Management Systems 3
D Switching Master Product in EO (Ethylene Oxide) Other Medical Device Related Standards 5
N Decision to file a new 510(k) - ETO (Ethylene Oxide) Sterilization Other US Medical Device Regulations 4
V Ethylene Oxide Testing Requirements as per US-pharmacopoeia Other US Medical Device Regulations 2
C ETO (Ethylene Oxide) Sterilization Audit Question Other Medical Device Related Standards 9
Sam Lazzara Ethylene Oxide Sterilization Validation question Other Medical Device Related Standards 13
A Moving from Ethylene Oxide to Gamma Sterilization - What is required? Other Medical Device Related Standards 10
I EO (Ethylene Oxide) Residuals - EN ISO 10993-7:2008 EU Medical Device Regulations 5
M EO (Ethylene Oxide) and ECH (Ethylene Chlorohydrin) Residual as per ISO 10993-7:2008 ISO 13485:2016 - Medical Device Quality Management Systems 15
A Designing an Ethylene Oxide Sterilization Process Other Medical Device Regulations World-Wide 7
M Effect of double ETO (Ethylene Oxide) Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 13
M Ethylene Oxide Gas Shelf Life ISO 13485:2016 - Medical Device Quality Management Systems 7
M Biological Indicator Ethylene Oxide Sterilization Other Medical Device and Orthopedic Related Topics 11
M Faulty Bioindicator Ethylene Oxide Sterilization Other Medical Device and Orthopedic Related Topics 16
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
V EMS for Medical Device Facility with Ethylene Oxide (ETO) Sterilization process. Miscellaneous Environmental Standards and EMS Related Discussions 4
K Iso 11135:1994 - Validation and Control of Ethylene Oxide Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 5
S ECH (ethylene chlorohydrin) residues - Plastic catheter packed in a sterile bag Other Medical Device Related Standards 5
P PMMA IRK304 Resin and EVA (Ethylene Vinyl Acetate) Tape causes a Stain Mark Manufacturing and Related Processes 3
C Does a medical device active (zinc oxide) needs BPR registration in EU? Other ISO and International Standards and European Regulations 5
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Zinc Oxide as a Topical Application EU Medical Device Regulations 4
C Black Oxide Gage Pin Calibration Requirements General Measurement Device and Calibration Topics 4
P How to prove that Cr oxide makes a stent biocompatible (e-polished 316LVM) Manufacturing and Related Processes 9
M Is Dental Impression Paste (Zinc Oxide and Eugenol) a Drug 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
S Classification of a Sterile Procedure Pack EU Medical Device Regulations 0
L "Shelf-Life" Class I Non-Sterile Products Expiration Date? CE Marking (Conformité Européene) / CB Scheme 3
M Changing Investigational Device from Non-Sterile to Sterile - what are the implications? Other Medical Device and Orthopedic Related Topics 3
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
L Water requirement for Non-sterile topical OTCs Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
M Bioburden Frequency (sterile MD) Other Medical Device Related Standards 2
shimonv Single lot release for sterile packaging 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
K Sterile barrier system symbols - Label redlines for a Class III device (Kit Box) EU Medical Device Regulations 0
C Where to place CE/label in Sterile IVD EU Medical Device Regulations 1
K ISO 15223-1 Sterile barrier system - The 3-layer symbol EU Medical Device Regulations 1
J Sterile medical device expiration date 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
B Bioburden monitoring for surgical instrument provided non-sterile EU Medical Device Regulations 3
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7

Similar threads

Top Bottom