Ethylene Oxide Sterile Implantable Medical Device Shelf Life Extension

D

dallltho

If you have a sterile implantable medical device that is sterilized by Ethylene Oxide (EO aka ETO) and want to extend the shelf life, how many product batches or lots would need to be verified via testing in order to justify the extension? Is there a general best practice? If so, which standard or regulation or guidance would direct us to verify such a number of lots to test (for example 1 lot or 3 lots etc.)?
 

planB

Super Moderator
dalltho,

follow ISO 11607, section 6.4. You have to justify your sample size for your seal- and packaging-integrity tests at the extended time points by a statistical rationale. Typically, you demonstrate batch-wise reproducibility of your sealing process during the PQ of your initial sealing validation at timepoint = 0, and thus, do not have to consider batch variations as a variable during aging/stability studies.

HTH,

Gerhard
 
Thread starter Similar threads Forum Replies Date
A ETHYLENE OXIDE STERILIZATION VALIDATION Manufacturing and Related Processes 4
G Number of Destructively Tested Devices Needed for Ethylene Oxide Validation Other Medical Device Related Standards 4
M Informational USFDA Ethylene Oxide Sterilization for Medical Devices website Medical Device and FDA Regulations and Standards News 0
M Informational FDA Posts Information about Ethylene Oxide Sterilization and Medical Devices Medical Device and FDA Regulations and Standards News 0
T ISO 10993-7 - Ethylene oxide sterilization residues Other Medical Device Related Standards 1
E Ethylene Oxide Sterilization Validation-Reference Load Other Medical Device Related Standards 8
D Ethylene Oxide Sterilization aeration Manufacturing and Related Processes 1
B Product definition for Ethylene Oxide Sterilization Validation Other Medical Device Related Standards 2
D Adding an extra Label on an Ethylene Oxide Sterilized Box ISO 13485:2016 - Medical Device Quality Management Systems 3
D Switching Master Product in EO (Ethylene Oxide) Other Medical Device Related Standards 5
N Decision to file a new 510(k) - ETO (Ethylene Oxide) Sterilization Other US Medical Device Regulations 4
V Ethylene Oxide Testing Requirements as per US-pharmacopoeia Other US Medical Device Regulations 2
C ETO (Ethylene Oxide) Sterilization Audit Question Other Medical Device Related Standards 13
Sam Lazzara Ethylene Oxide Sterilization Validation question Other Medical Device Related Standards 13
A Moving from Ethylene Oxide to Gamma Sterilization - What is required? Other Medical Device Related Standards 10
I EO (Ethylene Oxide) Residuals - EN ISO 10993-7:2008 EU Medical Device Regulations 5
M EO (Ethylene Oxide) and ECH (Ethylene Chlorohydrin) Residual as per ISO 10993-7:2008 ISO 13485:2016 - Medical Device Quality Management Systems 15
A Designing an Ethylene Oxide Sterilization Process Other Medical Device Regulations World-Wide 7
M Effect of double ETO (Ethylene Oxide) Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 13
M Ethylene Oxide Gas Shelf Life ISO 13485:2016 - Medical Device Quality Management Systems 7
M Biological Indicator Ethylene Oxide Sterilization Other Medical Device and Orthopedic Related Topics 11
M Faulty Bioindicator Ethylene Oxide Sterilization Other Medical Device and Orthopedic Related Topics 16
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
V EMS for Medical Device Facility with Ethylene Oxide (ETO) Sterilization process. Miscellaneous Environmental Standards and EMS Related Discussions 4
K Iso 11135:1994 - Validation and Control of Ethylene Oxide Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 5
Soog1459 ECH (ethylene chlorohydrin) residues - Plastic catheter packed in a sterile bag Other Medical Device Related Standards 5
P PMMA IRK304 Resin and EVA (Ethylene Vinyl Acetate) Tape causes a Stain Mark Manufacturing and Related Processes 3
C Does a medical device active (zinc oxide) needs BPR registration in EU? Other ISO and International Standards and European Regulations 6
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Zinc Oxide as a Topical Application EU Medical Device Regulations 4
C Black Oxide Gage Pin Calibration Requirements General Measurement Device and Calibration Topics 4
P How to prove that Cr oxide makes a stent biocompatible (e-polished 316LVM) Manufacturing and Related Processes 9
M Is Dental Impression Paste (Zinc Oxide and Eugenol) a Drug 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
adztesla Sterile packaging validation for medical device - ASTMD4332 and ISTA2A ISO 13485:2016 - Medical Device Quality Management Systems 3
Leigh76 Regulatory Path for Providing Sterile Product in Bulk Non-Sterile Form to Kit Packers US Medical Device Regulations 3
J Is ISO 13485 certification necessary for Class I (non-sterile) medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 7
C Sterile label on secondary packaging when mixed contents EU Medical Device Regulations 4
J Sterile Procedure Pack Labeling Medical Device and FDA Regulations and Standards News 2
F Class A - sterile vs non sterile EU Medical Device Regulations 5
E Control of Sterile Barriers made by the supplier Design and Development of Products and Processes 4
C Legacy device under Class A sterile devices EU Medical Device Regulations 0
A Stability Testing of a device with sterile fluid pathway claim Other Medical Device and Orthopedic Related Topics 1
B Transport Validation For Non-sterile Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 4
L Saudi regulations for Japanese non sterile scopes Japan Medical Device Regulations 0
M Sterile barrier system with outer protective packaging Symbol EU Medical Device Regulations 11
A Sterile Packaging Quick Query EU Medical Device Regulations 5
D Shelf life of non sterile class IIb product. ISO 13485:2016 - Medical Device Quality Management Systems 1
ChrisM Class 1 non-sterile device, need certified QMS or not? EU Medical Device Regulations 2
G Shipping Validation of Non-Sterile Parts? Other Medical Device and Orthopedic Related Topics 9
Edward Reesor Class I non-sterile "Shelf Life" labeling (EU/UK) CE Marking (Conformité Européene) / CB Scheme 4

Similar threads

Top Bottom