Hello! My company is about to run an EtO sterilization validation on medical devices. We were told that our reference load, or dunnage load, needed to be filled with all products that represent the worst case load and that these products will be scrap. For a small company like us, this is just not possible to have that many products (800 something products) going to waste.

Does anyone know if the reference load needs to actually consist of the same product? Can it just be boxes of the same filled with something else like packing peanuts or something? Is there any possible way around this?

Any help with this is greatly appreciated!
 
Elsmar Forum Sponsor

Rincewind

Involved In Discussions
#2
Hello,

you should perform a documented worst case evaluation of your products and run the validation with those worst case products. Only by doing this you can claim that the chosen sterilization method is able to provide a sterile product.

You don't necessarily have to this with all 800 products of yours, only with the ones you deemed to be the most difficult to sterilize (the worst case products).
 

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#3
What Rince said. Give them samples of your highest burden products to make a worst case example. There is a lot of upfront cost in this but once established you can paper adopt new products. Try to have the validation cover a "family" of products where you can (Similar products grouped together)
 
Thread starter Similar threads Forum Replies Date
M Informational USFDA Ethylene Oxide Sterilization for Medical Devices website Medical Device and FDA Regulations and Standards News 0
M Informational FDA Posts Information about Ethylene Oxide Sterilization and Medical Devices Medical Device and FDA Regulations and Standards News 0
T ISO 10993-7 - Ethylene oxide sterilization residues Other Medical Device Related Standards 1
D Ethylene Oxide Sterilization aeration Manufacturing and Related Processes 1
B Product definition for Ethylene Oxide Sterilization Validation Other Medical Device Related Standards 2
N Decision to file a new 510(k) - ETO (Ethylene Oxide) Sterilization Other US Medical Device Regulations 4
C ETO (Ethylene Oxide) Sterilization Audit Question Other Medical Device Related Standards 9
Sam Lazzara Ethylene Oxide Sterilization Validation question Other Medical Device Related Standards 13
A Moving from Ethylene Oxide to Gamma Sterilization - What is required? Other Medical Device Related Standards 10
A Designing an Ethylene Oxide Sterilization Process Other Medical Device Regulations World-Wide 7
M Effect of double ETO (Ethylene Oxide) Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 13
M Biological Indicator Ethylene Oxide Sterilization Other Medical Device and Orthopedic Related Topics 11
M Faulty Bioindicator Ethylene Oxide Sterilization Other Medical Device and Orthopedic Related Topics 16
V EMS for Medical Device Facility with Ethylene Oxide (ETO) Sterilization process. Miscellaneous Environmental Standards and EMS Related Discussions 4
K Iso 11135:1994 - Validation and Control of Ethylene Oxide Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 5
D Ethylene Oxide Sterile Implantable Medical Device Shelf Life Extension Other Medical Device Related Standards 1
D Adding an extra Label on an Ethylene Oxide Sterilized Box ISO 13485:2016 - Medical Device Quality Management Systems 3
D Switching Master Product in EO (Ethylene Oxide) Other Medical Device Related Standards 5
V Ethylene Oxide Testing Requirements as per US-pharmacopoeia Other US Medical Device Regulations 2
I EO (Ethylene Oxide) Residuals - EN ISO 10993-7:2008 EU Medical Device Regulations 5
M EO (Ethylene Oxide) and ECH (Ethylene Chlorohydrin) Residual as per ISO 10993-7:2008 ISO 13485:2016 - Medical Device Quality Management Systems 15
M Ethylene Oxide Gas Shelf Life ISO 13485:2016 - Medical Device Quality Management Systems 7
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
S ECH (ethylene chlorohydrin) residues - Plastic catheter packed in a sterile bag Other Medical Device Related Standards 5
P PMMA IRK304 Resin and EVA (Ethylene Vinyl Acetate) Tape causes a Stain Mark Manufacturing and Related Processes 3
C Does a medical device active (zinc oxide) needs BPR registration in EU? Other ISO and International Standards and European Regulations 1
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Zinc Oxide as a Topical Application EU Medical Device Regulations 4
C Black Oxide Gage Pin Calibration Requirements General Measurement Device and Calibration Topics 4
P How to prove that Cr oxide makes a stent biocompatible (e-polished 316LVM) Manufacturing and Related Processes 9
M Is Dental Impression Paste (Zinc Oxide and Eugenol) a Drug 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
validationspec LDPE for EO Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 2
B Gamma Sterilization Other Medical Device Related Standards 1
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
S Orthopedic Implants (Knee system) Sterilization Indicators Manufacturing and Related Processes 6
chris1price Contract Sterilization Service Providers ISO 13485:2016 - Medical Device Quality Management Systems 5
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
R Scientific Operating Procedures For Sterilization Practices in India PSA/2012/1 Other Medical Device Regulations World-Wide 0
N Looking for Gamma sterilization facility in China ISO 13485:2016 - Medical Device Quality Management Systems 4
I Release checklist EO sterilization process Medical Device and FDA Regulations and Standards News 6
MrTetris Are GLP required for testing cytotoxicity and soil remaining after sterilization of MD? Other Medical Device Related Standards 8
Z EN 868-8 "required tests" - Rigid sterilization containers Other Medical Device Related Standards 0
Judy Abbott Increase the maximum allowable dose for E-beam sterilization Reliability Analysis - Predictions, Testing and Standards 2
H EO Sterilization Validation - Sterility Testing and Load Configuration Other Medical Device Related Standards 3
A Reprocessing, Cleaning, Disinfection and Sterilization Validation EU Medical Device Regulations 4
W EO sterilization and pallets Other Medical Device and Orthopedic Related Topics 4

Similar threads

Top Bottom