Eto Sterilisation Re-Validation Requirements

A

AliceW

#1
If a new cleanroom is built to meet the specs of the old cleanroom and the product has not changed, the process has not changed and the operators have not changed - do you need to re-validate the sterilisation process?
 
Elsmar Forum Sponsor
I

iamtroll

#2
Re: Eto sterilisation re-validation

You can probably justify with a paper review by restating all of those factors in detail if you can provide proof that your bioburden is in the same range.
 

Ronen E

Problem Solver
Staff member
Moderator
#3
Re: Eto sterilisation re-validation

...if you can provide proof that your bioburden is in the same range.
Pay attention to both numbers and types of organisms.

If a new cleanroom is built... the process has not changed...
I'm 99% sure that if the operations have moved to a new cleanroom facility, the process must have changed in some ways, not necessarily negligible, regardless of specs, SOPs, personnel etc. ;) If you look close enough you will spot the differences...
 

Ajit Basrur

Staff member
Admin
#4
If a new cleanroom is built to meet the specs of the old cleanroom and the product has not changed, the process has not changed and the operators have not changed - do you need to re-validate the sterilisation process?
Yes, the relocation involves dismantling and reinstalling ... that calls for a requalification and revalidation.

There was an earlier FDA warning Letter issued to Lydall Inc. for ETO relocation -

2. Failure to review and evaluate a process and perform revalidations where appropriate when changes or process deviations occur, as required by 21 CFR 820.75 (c). (FDA 483 item #2) Specifically,

a When your manufacturing process was relocated to North Carolina, the Et0 sterilization process was not revalidated. (FDA 483 item # 1.c).
 
Thread starter Similar threads Forum Replies Date
N ETO Sterilisation Validation - EO Residual Minimum Sample Requirement ISO 13485:2016 - Medical Device Quality Management Systems 2
chris1price EtO Sterilisation Consultant Wanted Job Openings, Consulting and Employment Opportunities 0
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
Q Sterilisation ISO 11135 EtO-in-a-Bag (Novel Non-Traditional EtO Sterilization) Other Medical Device Related Standards 12
P ETO Sterilisation of Single Use Part supplied in Non-Sterile State EU Medical Device Regulations 12
T ETO Product Sterilisation Testing Standard Requirements ISO 13485:2016 - Medical Device Quality Management Systems 4
C ETO Sterilised Class II Medical Device - Required Temperature Storage ISO 13485:2016 - Medical Device Quality Management Systems 1
A Is Aeration Process in ETO Sterilization Validation Required? Manufacturing and Related Processes 7
N Converting ETO Residual Limits from PPM to mg ISO 13485:2016 - Medical Device Quality Management Systems 6
M Is it essential to do the Fractional Test for ETO Revalidation? ISO 13485:2016 - Medical Device Quality Management Systems 1
K Re-validation of EtO sterilization after change of tape Other Medical Device and Orthopedic Related Topics 1
N Decision to file a new 510(k) - ETO (Ethylene Oxide) Sterilization Other US Medical Device Regulations 4
C ETO (Ethylene Oxide) Sterilization Audit Question Other Medical Device Related Standards 9
A How to overcome brown stains after ETO sterilization process ? Other Medical Device Related Standards 2
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
S Fractional cycle fail in Eto sterilization validation ISO 13485:2016 - Medical Device Quality Management Systems 3
R Eto Sterilization Product Adoption SOP ISO 13485:2016 - Medical Device Quality Management Systems 4
M EtO compatible or recommended materials Other Medical Device Regulations World-Wide 1
X EtO Steriliser Software (PACS) Update Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
S IV Catheter Biological Evaluation - EtO Residuals - Permanent Contact Definition Qualification and Validation (including 21 CFR Part 11) 2
S ETO ISO 11135-1 & ISO 17665-2 - Purity and Quality of Steam (Humidity) Qualification and Validation (including 21 CFR Part 11) 3
A Aeration Room Requirements for EtO Sterilization Other US Medical Device Regulations 3
T ETO Sterilization through luer caps Manufacturing and Related Processes 5
F Implementing ISO 11135 - Facilities Conditions - Adding ETO Sterilization Other Medical Device Related Standards 4
N EtO Sterilization Process Waste Treatment System Other Medical Device Related Standards 1
G "Intrinsically Safe" requirement - EtO sterilization of device with Battery Other Medical Device Related Standards 9
D Will just Heat and Humidity be Adequate to Simulate EtO Sterilization? Design and Development of Products and Processes 2
K Sterilization for Hypodermic Needle - EtO or Gamma? ISO 13485:2016 - Medical Device Quality Management Systems 1
D ETO Sterility Requirements for FIM Clinical Study in Europe Other Medical Device Regulations World-Wide 3
Mikilk ETO Sterilization - Annual Report of all Sterilization Validations to send the NB EU Medical Device Regulations 3
L ETO vs. Gamma Irradiation Sterilization of DBM (Demineralized Bone Matrix) Other Medical Device and Orthopedic Related Topics 3
M Effect of double ETO (Ethylene Oxide) Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 13
S AAMI TIR28:2009 Product adoption & process equivalency for ETO sterilization Other US Medical Device Regulations 2
I ETO Sterilizer Validation (ISO 11135:2007) Other Medical Device Related Standards 23
M Eto Certificate of Sterilization (vs. gamma irradiation) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
B ISO 11135:2007 (EtO Sterilization) ISO 13485:2016 - Medical Device Quality Management Systems 5
I Guidelines that specify the frequency of ETO Sterilizer Re-Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
T ETO Sterilization - Process monitoring and out-inspection ISO 13485:2016 - Medical Device Quality Management Systems 3
M Japanese rule on EtO residuals in sterilized medical devices? Various Other Specifications, Standards, and related Requirements 16
G EtO Sterilization - DIN/EN 550 - ISO 11135 Other US Medical Device Regulations 20
Ajit Basrur Audit Checklist for Contract Gamma and ETO Sterilization needed ISO 13485:2016 - Medical Device Quality Management Systems 3
V EMS for Medical Device Facility with Ethylene Oxide (ETO) Sterilization process. Miscellaneous Environmental Standards and EMS Related Discussions 4
V ETO Sterilization of Urological Catheter ISO 13485:2016 - Medical Device Quality Management Systems 5
Q EtO vs. Gamma Sterilization - Stainless Steel and Titanium Orthopaedic Implants ISO 13485:2016 - Medical Device Quality Management Systems 7
C How to verify the SAL (sterilie acceptable level) is 10-6 for ETO Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 6
G Supplier audit of ETO Sterilization Lab - Need a checklist or advice Supplier Quality Assurance and other Supplier Issues 3
P ETO Sterilization Validation Process and an NCR - ISO 13485:2003 CMDCAS audit ISO 13485:2016 - Medical Device Quality Management Systems 1
J EU countries and steam sterilisation protocol EU Medical Device Regulations 0
M Is there an Industry Norm for Expiry of Sterilisation Validity? EU Medical Device Regulations 6

Similar threads

Top Bottom