ETO Sterilised Class II Medical Device - Required Temperature Storage

#1
Hi all,

I've recently inherited an existing product to manage. It currently has a specified temperature storage range of 10-30 degrees. This is unfortunately causing some problems with storage and monitoring of temps.

Upon speaking to people, no one seems to know how we came to this temperature range. A lot of our products certainly require specialist storage due to chemical components but this has none. Also, whilst speaking with our sterilisation manufacturer they commented that very few products that they sterilise in pouches by ETO are labelled with a temperature storage range?

Is a storage temperature required to be on the label or has someone been over zealous in the past and restricted our storage options?

The product is simply stainless steel and polyester. I'm hopefully looking too increase temp range or if justifiable remove it completely. Alas having trouble finding concrete specs.

Thanks for your answers.
 
Elsmar Forum Sponsor

planB

Trusted Information Resource
#2
Chris,

Storage temperature may be driven not by (only) the packaged product, but (more often) by the sterile barrier system and its associated stability / shelf life validation according to ISO 11607-1. This may also be the case for your product. Specifically for your product, you should have a shelf-life validation report on file, that demonstrates that your product can be stored at 30 °C (which generally implies that any lower temperature, including your min of 10 °C is also fine) over the labelled shelf life.

Since you cannot store pouches at arbitrarily high temperatures you always to have to label at least an upper storage temperature. Typically, there is no justification to completely remove temperature labelling on sterile barrier systems.

Regarding the comment of your contract steriliser: this is worrisome, since these products in pouches are actually mislabelled. (Typically, a lot of low-risk products are sealed in pouches, that have little or no regulatory oversight.) Another option could be that final labelling for distribution only happens after sterilisation on return to the manufacturer. This way your contractor would not see the full picture.

You can extend your labelled temperature range if your shelf life validation covers the desired, extended temperature range.

HTH,
 
Thread starter Similar threads Forum Replies Date
A Is Aeration Process in ETO Sterilization Validation Required? Manufacturing and Related Processes 7
N ETO Sterilisation Validation - EO Residual Minimum Sample Requirement ISO 13485:2016 - Medical Device Quality Management Systems 2
N Converting ETO Residual Limits from PPM to mg ISO 13485:2016 - Medical Device Quality Management Systems 6
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
M Is it essential to do the Fractional Test for ETO Revalidation? ISO 13485:2016 - Medical Device Quality Management Systems 1
K Re-validation of EtO sterilization after change of tape Other Medical Device and Orthopedic Related Topics 1
N Decision to file a new 510(k) - ETO (Ethylene Oxide) Sterilization Other US Medical Device Regulations 4
C ETO (Ethylene Oxide) Sterilization Audit Question Other Medical Device Related Standards 9
A How to overcome brown stains after ETO sterilization process ? Other Medical Device Related Standards 2
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
S Fractional cycle fail in Eto sterilization validation ISO 13485:2016 - Medical Device Quality Management Systems 3
R Eto Sterilization Product Adoption SOP ISO 13485:2016 - Medical Device Quality Management Systems 4
M EtO compatible or recommended materials Other Medical Device Regulations World-Wide 1
X EtO Steriliser Software (PACS) Update Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
S IV Catheter Biological Evaluation - EtO Residuals - Permanent Contact Definition Qualification and Validation (including 21 CFR Part 11) 2
S ETO ISO 11135-1 & ISO 17665-2 - Purity and Quality of Steam (Humidity) Qualification and Validation (including 21 CFR Part 11) 3
A Eto Sterilisation Re-Validation Requirements Other Medical Device Related Standards 4
Q Sterilisation ISO 11135 EtO-in-a-Bag (Novel Non-Traditional EtO Sterilization) Other Medical Device Related Standards 10
A Aeration Room Requirements for EtO Sterilization Other US Medical Device Regulations 3
T ETO Sterilization through luer caps Manufacturing and Related Processes 5
F Implementing ISO 11135 - Facilities Conditions - Adding ETO Sterilization Other Medical Device Related Standards 4
N EtO Sterilization Process Waste Treatment System Other Medical Device Related Standards 1
G "Intrinsically Safe" requirement - EtO sterilization of device with Battery Other Medical Device Related Standards 9
D Will just Heat and Humidity be Adequate to Simulate EtO Sterilization? Design and Development of Products and Processes 2
K Sterilization for Hypodermic Needle - EtO or Gamma? ISO 13485:2016 - Medical Device Quality Management Systems 1
D ETO Sterility Requirements for FIM Clinical Study in Europe Other Medical Device Regulations World-Wide 3
Mikilk ETO Sterilization - Annual Report of all Sterilization Validations to send the NB EU Medical Device Regulations 3
L ETO vs. Gamma Irradiation Sterilization of DBM (Demineralized Bone Matrix) Other Medical Device and Orthopedic Related Topics 3
M Effect of double ETO (Ethylene Oxide) Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 13
S AAMI TIR28:2009 Product adoption & process equivalency for ETO sterilization Other US Medical Device Regulations 2
P ETO Sterilisation of Single Use Part supplied in Non-Sterile State EU Medical Device Regulations 12
I ETO Sterilizer Validation (ISO 11135:2007) Other Medical Device Related Standards 23
M Eto Certificate of Sterilization (vs. gamma irradiation) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
B ISO 11135:2007 (EtO Sterilization) ISO 13485:2016 - Medical Device Quality Management Systems 5
I Guidelines that specify the frequency of ETO Sterilizer Re-Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
T ETO Sterilization - Process monitoring and out-inspection ISO 13485:2016 - Medical Device Quality Management Systems 3
M Japanese rule on EtO residuals in sterilized medical devices? Various Other Specifications, Standards, and related Requirements 16
G EtO Sterilization - DIN/EN 550 - ISO 11135 Other US Medical Device Regulations 20
Ajit Basrur Audit Checklist for Contract Gamma and ETO Sterilization needed ISO 13485:2016 - Medical Device Quality Management Systems 3
V EMS for Medical Device Facility with Ethylene Oxide (ETO) Sterilization process. Miscellaneous Environmental Standards and EMS Related Discussions 4
T ETO Product Sterilisation Testing Standard Requirements ISO 13485:2016 - Medical Device Quality Management Systems 4
V ETO Sterilization of Urological Catheter ISO 13485:2016 - Medical Device Quality Management Systems 5
Q EtO vs. Gamma Sterilization - Stainless Steel and Titanium Orthopaedic Implants ISO 13485:2016 - Medical Device Quality Management Systems 7
C How to verify the SAL (sterilie acceptable level) is 10-6 for ETO Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 6
G Supplier audit of ETO Sterilization Lab - Need a checklist or advice Supplier Quality Assurance and other Supplier Issues 3
P ETO Sterilization Validation Process and an NCR - ISO 13485:2003 CMDCAS audit ISO 13485:2016 - Medical Device Quality Management Systems 1
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 3

Similar threads

Top Bottom