Re: Gap analysis for ISO 11135-1994 and 2007

Thank you for the document, this is highly appreciated.
Shira
 
Elsmar Forum Sponsor
M

microb7

#12
Re: Gap analysis for ISO 11135-1994 and 2007

Hello all,
I do not have a formal gap analysis, but I do have a lot of experience with EO sterilization validations. The main difference in the ISO11135 1995 to 2007 version is the addition of a requirement for a Fractional (Sub –lethal) Cycle to your validation runs. Also, this cycle has to be run at least every 24 months along with the annual re-validation. This is a cycle with a shortened EO dwell time duration (Less than a half-cycle time) when compared to the full cycle and with all other parameters constant. This cycle should produce biological indicator growth with no natural product sterility (NPS) growth and will be run to validate the biological recovery process as per ISO 11135-1: 2007. The involvement of this cycle is to fulfill the requirement to demonstrate that the natural product sterility (NPS) samples are less of a challenge to EO sterilization than the biological indicators (BI’s).
 

Doug Tropf

Quite Involved in Discussions
#13
Re: Gap analysis for ISO 11135-1994 and 2007

Hello all,
I do not have a formal gap analysis, but I do have a lot of experience with EO sterilization validations. The main difference in the ISO11135 1995 to 2007 version is the addition of a requirement for a Fractional (Sub –lethal) Cycle to your validation runs. Also, this cycle has to be run at least every 24 months along with the annual re-validation. This is a cycle with a shortened EO dwell time duration (Less than a half-cycle time) when compared to the full cycle and with all other parameters constant. This cycle should produce biological indicator growth with no natural product sterility (NPS) growth and will be run to validate the biological recovery process as per ISO 11135-1: 2007. The involvement of this cycle is to fulfill the requirement to demonstrate that the natural product sterility (NPS) samples are less of a challenge to EO sterilization than the biological indicators (BI’s).
Well said and welcome to the Cove.
 

medwise

Involved In Discussions
#14
Re: Gap analysis for ISO 11135-1994 and 2007

Hello all,
I do not have a formal gap analysis, but I do have a lot of experience with EO sterilization validations. The main difference in the ISO11135 1995 to 2007 version is the addition of a requirement for aFractional (Sub –lethal) Cycle to your validation runs. Also, this cycle has to be run at least every 24 months along with the annual re-validation. This is a cycle with a shortened EO dwell time duration (Less than a half-cycle time) when compared to the full cycle and with all other parameters constant. This cycle should produce biological indicator growth with no natural product sterility (NPS) growth and will be run to validate the biological recovery process as per ISO 11135-1: 2007. The involvement of this cycle is to fulfill the requirement to demonstrate that the natural product sterility (NPS) samples are less of a challenge to EO sterilization than the biological indicators (BI’s).
Hi microb7,

I am very naive to EtO sterilization. Can you please explain me what do you meant by " This cycle should produce biological indicator growth with no natural product sterility (NPS) growth and will be run to validate the biological recovery process as per ISO 11135-1: 2007"?

I am not familiar with the term natural product sterility.

Thanks

Kindest Regards,
Romit:)
 
S

Shira

#15
The meaning is that you should provide objective evidence that the challenge you are using has higher resistance than your product bioburden (i.e. product natural bioburden). If you are using a process challenge device (PCD) you should provide objective evidence that an External PCD (if used) is more resistant than Internal PCD, and that the internal PCD is more resistant than the inoculated product (at the worst case to sterilize locations), which has higher resistant than the natural product bioburden.

Hope it helps :)
 
P

phoexix

#16
Re: Gap analysis for ISO 11135-1994 and 2007

Hello all,
I do not have a formal gap analysis, but I do have a lot of experience with EO sterilization validations. The main difference in the ISO11135 1995 to 2007 version is the addition of a requirement for aFractional (Sub –lethal) Cycle to your validation runs. Also, this cycle has to be run at least every 24 months along with the annual re-validation. This is a cycle with a shortened EO dwell time duration (Less than a half-cycle time) when compared to the full cycle and with all other parameters constant. This cycle should produce biological indicator growth with no natural product sterility (NPS) growth and will be run to validate the biological recovery process as per ISO 11135-1: 2007. The involvement of this cycle is to fulfill the requirement to demonstrate that the natural product sterility (NPS) samples are less of a challenge to EO sterilization than the biological indicators (BI’s).
Hi Microb7, pls see below clause from 1994 version, what's your thought now?

c) When microbial identifications are not performed and the bioburden estimate is high: the
appropriateness of the biological indicator should be determined by exposure to sublethal (partial)
cycles. In these studies, the relative inactivation rates can be compared through testing.

A cycle of short duration from which survivors can be recovered should also be run to demonstrate the adequacy of the recovery technique.
 

harry

Super Moderator
#18
Dear All,

Any body have ISO 11137:2006 Version.

Thanks & Regards
Welcome to the Cove.

I believe some users here do have.

Not sure why you asked but sharing of copyrighted documents is strictly prohibited here. A good place to locate a copy for reference purpose would be your local university library.
 
Thread starter Similar threads Forum Replies Date
A Is Aeration Process in ETO Sterilization Validation Required? Manufacturing and Related Processes 7
K Re-validation of EtO sterilization after change of tape Other Medical Device and Orthopedic Related Topics 1
N Decision to file a new 510(k) - ETO (Ethylene Oxide) Sterilization Other US Medical Device Regulations 4
C ETO (Ethylene Oxide) Sterilization Audit Question Other Medical Device Related Standards 9
A How to overcome brown stains after ETO sterilization process ? Other Medical Device Related Standards 2
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
S Fractional cycle fail in Eto sterilization validation ISO 13485:2016 - Medical Device Quality Management Systems 3
R Eto Sterilization Product Adoption SOP ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Sterilisation ISO 11135 EtO-in-a-Bag (Novel Non-Traditional EtO Sterilization) Other Medical Device Related Standards 12
A Aeration Room Requirements for EtO Sterilization Other US Medical Device Regulations 3
T ETO Sterilization through luer caps Manufacturing and Related Processes 5
F Implementing ISO 11135 - Facilities Conditions - Adding ETO Sterilization Other Medical Device Related Standards 4
N EtO Sterilization Process Waste Treatment System Other Medical Device Related Standards 1
G "Intrinsically Safe" requirement - EtO sterilization of device with Battery Other Medical Device Related Standards 9
D Will just Heat and Humidity be Adequate to Simulate EtO Sterilization? Design and Development of Products and Processes 2
K Sterilization for Hypodermic Needle - EtO or Gamma? ISO 13485:2016 - Medical Device Quality Management Systems 1
Mikilk ETO Sterilization - Annual Report of all Sterilization Validations to send the NB EU Medical Device Regulations 3
L ETO vs. Gamma Irradiation Sterilization of DBM (Demineralized Bone Matrix) Other Medical Device and Orthopedic Related Topics 3
M Effect of double ETO (Ethylene Oxide) Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 13
S AAMI TIR28:2009 Product adoption & process equivalency for ETO sterilization Other US Medical Device Regulations 2
M Eto Certificate of Sterilization (vs. gamma irradiation) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
B ISO 11135:2007 (EtO Sterilization) ISO 13485:2016 - Medical Device Quality Management Systems 5
T ETO Sterilization - Process monitoring and out-inspection ISO 13485:2016 - Medical Device Quality Management Systems 3
Ajit Basrur Audit Checklist for Contract Gamma and ETO Sterilization needed ISO 13485:2016 - Medical Device Quality Management Systems 3
V EMS for Medical Device Facility with Ethylene Oxide (ETO) Sterilization process. Miscellaneous Environmental Standards and EMS Related Discussions 4
V ETO Sterilization of Urological Catheter ISO 13485:2016 - Medical Device Quality Management Systems 5
Q EtO vs. Gamma Sterilization - Stainless Steel and Titanium Orthopaedic Implants ISO 13485:2016 - Medical Device Quality Management Systems 7
C How to verify the SAL (sterilie acceptable level) is 10-6 for ETO Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 6
G Supplier audit of ETO Sterilization Lab - Need a checklist or advice Supplier Quality Assurance and other Supplier Issues 3
P ETO Sterilization Validation Process and an NCR - ISO 13485:2003 CMDCAS audit ISO 13485:2016 - Medical Device Quality Management Systems 1
chris1price EtO Sterilisation Consultant Wanted Job Openings, Consulting and Employment Opportunities 0
C ETO Sterilised Class II Medical Device - Required Temperature Storage ISO 13485:2016 - Medical Device Quality Management Systems 1
N ETO Sterilisation Validation - EO Residual Minimum Sample Requirement ISO 13485:2016 - Medical Device Quality Management Systems 2
N Converting ETO Residual Limits from PPM to mg ISO 13485:2016 - Medical Device Quality Management Systems 6
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
M Is it essential to do the Fractional Test for ETO Revalidation? ISO 13485:2016 - Medical Device Quality Management Systems 1
M EtO compatible or recommended materials Other Medical Device Regulations World-Wide 1
X EtO Steriliser Software (PACS) Update Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
S IV Catheter Biological Evaluation - EtO Residuals - Permanent Contact Definition Qualification and Validation (including 21 CFR Part 11) 2
S ETO ISO 11135-1 & ISO 17665-2 - Purity and Quality of Steam (Humidity) Qualification and Validation (including 21 CFR Part 11) 3
A Eto Sterilisation Re-Validation Requirements Other Medical Device Related Standards 4
D ETO Sterility Requirements for FIM Clinical Study in Europe Other Medical Device Regulations World-Wide 3
P ETO Sterilisation of Single Use Part supplied in Non-Sterile State EU Medical Device Regulations 12
I ETO Sterilizer Validation (ISO 11135:2007) Other Medical Device Related Standards 23
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
I Guidelines that specify the frequency of ETO Sterilizer Re-Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
M Japanese rule on EtO residuals in sterilized medical devices? Various Other Specifications, Standards, and related Requirements 16
T ETO Product Sterilisation Testing Standard Requirements ISO 13485:2016 - Medical Device Quality Management Systems 4
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2

Similar threads

Top Bottom