ETO Sterilizer Validation (ISO 11135:2007)
I
ISO MAN
Dear All,
I had been asked by one of my clients to conduct validation of large ETO sterilizer. As I do not have enough sensors to cover the sterilizer volume, we discussed that we can repeat each physical performance qualification twice to cover the volume of the chamber. I remember that this practice was acceptable in old version of may be of ISO 11135:1994 or EN:550, but I did not find it in the current version. Can anyone help by confirming whether this approach is correct or wrong, is there any guideline for this issue.
:thanks:
somashekar
Based on your ETO sterilizer design and your loading pattern, can you scientifically decide which is the challenge zone in the entire volume ? Would you have enough sensors to assess this zone. If so, you have done a major part of your validation step.
I
ISO MAN
Thank you for your reply. What I want to do is to cover the complete chamber, but divided into two zones, each zone is covered in seprate cycle. Is there any guidline or standard prevent this approach
somashekar
During the use for sterilization, if the load covers the complete chamber, I would perform the validation per a load cycle which means the complete chamber. Dividing into zones are for the convenience of practically representing the complete zone volume into identified sectors to compare results across the complete chamber.
bio_subbu
An ETO sterilization chamber must be validated depending upon the size of usable volume. For eg. If the ETO chamber usable internal volume of 25 cubic meters and you can use 25 sensors for a performance qualification study. I don’t think so it is acceptable to test twice for the physical performance qualification to cover the entire volume.
Regards
S. Subramaniam
bio_subbu
can anyone in this thread help me with a reference version of ISO 11737?
Priya
There are two parts in ISO 11737, Part 1 (Determination of a population of microorganisms on products) and part 2 (Tests of sterility performed in the validation of a sterilization process).
ISO 11737-1:2006 specifies requirements and provides guidance for the enumeration and microbial characterization of the population of viable microorganisms on or in a medical device, component, raw material or package.
ISO 11737-2:2009 specifies the general criteria for tests of sterility on medical devices which have been exposed to a treatment with the sterilizing agent that is a fraction of the specified sterilization process. These tests are intended to be performed when validating a sterilization process.
Priya, can you please brief what information you looking for?
Regards
S. Subramaniam
medwise
Dear All,
Can anyone please let me know the changes in the regulatory and industry expectation in the validation and ongoing monitoring of EtO process?
Basically I want to know the difference between ISO 11135:1994 and ISO 11135-1:2007.
I have read the 2007 version but since 1994 was not available to me I cant identify the difference between them.
Any help will be much appreciated.
Kindest regards
Romit:)
bio_subbu
Dear All,
Refer AAMI TIR14:2009 (Contract sterilization using ethylene oxide), AAMI TIR15:2009 (Physical aspects of ethylene oxide sterilization) and AAMI TIR16:2009 (Microbiological aspects of ethylene oxide sterilization). The AAMI (Association for the Advancement of Medical Instrumentation) has released an updated technical information report TIR15 on April 6, 2010, that offers guidance on developing sterilization cycles for ethylene oxide (EO) sterilization. For more details find the press release from AAMI website and refer AAMI Database to find and purchase the TIR reports and other standards.
I Guidelines that specify the frequency of ETO Sterilizer Re-Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
chris1price EtO Sterilisation Consultant Wanted Job Openings, Consulting and Employment Opportunities 0
C ETO Sterilised Class II Medical Device - Required Temperature Storage ISO 13485:2016 - Medical Device Quality Management Systems 1
A Is Aeration Process in ETO Sterilization Validation Required? Manufacturing and Related Processes 7
N ETO Sterilisation Validation - EO Residual Minimum Sample Requirement ISO 13485:2016 - Medical Device Quality Management Systems 2
N Converting ETO Residual Limits from PPM to mg ISO 13485:2016 - Medical Device Quality Management Systems 6
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
M Is it essential to do the Fractional Test for ETO Revalidation? ISO 13485:2016 - Medical Device Quality Management Systems 1
K Re-validation of EtO sterilization after change of tape Other Medical Device and Orthopedic Related Topics 1
N Decision to file a new 510(k) - ETO (Ethylene Oxide) Sterilization Other US Medical Device Regulations 4
C ETO (Ethylene Oxide) Sterilization Audit Question Other Medical Device Related Standards 9
A How to overcome brown stains after ETO sterilization process ? Other Medical Device Related Standards 2
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
S Fractional cycle fail in Eto sterilization validation ISO 13485:2016 - Medical Device Quality Management Systems 3
R Eto Sterilization Product Adoption SOP ISO 13485:2016 - Medical Device Quality Management Systems 4
M EtO compatible or recommended materials Other Medical Device Regulations World-Wide 1
X EtO Steriliser Software (PACS) Update Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
S IV Catheter Biological Evaluation - EtO Residuals - Permanent Contact Definition Qualification and Validation (including 21 CFR Part 11) 2
S ETO ISO 11135-1 & ISO 17665-2 - Purity and Quality of Steam (Humidity) Qualification and Validation (including 21 CFR Part 11) 3
A Eto Sterilisation Re-Validation Requirements Other Medical Device Related Standards 4
Q Sterilisation ISO 11135 EtO-in-a-Bag (Novel Non-Traditional EtO Sterilization) Other Medical Device Related Standards 12
A Aeration Room Requirements for EtO Sterilization Other US Medical Device Regulations 3
T ETO Sterilization through luer caps Manufacturing and Related Processes 5
F Implementing ISO 11135 - Facilities Conditions - Adding ETO Sterilization Other Medical Device Related Standards 4
N EtO Sterilization Process Waste Treatment System Other Medical Device Related Standards 1
G "Intrinsically Safe" requirement - EtO sterilization of device with Battery Other Medical Device Related Standards 9
D Will just Heat and Humidity be Adequate to Simulate EtO Sterilization? Design and Development of Products and Processes 2
K Sterilization for Hypodermic Needle - EtO or Gamma? ISO 13485:2016 - Medical Device Quality Management Systems 1
D ETO Sterility Requirements for FIM Clinical Study in Europe Other Medical Device Regulations World-Wide 3
Mikilk ETO Sterilization - Annual Report of all Sterilization Validations to send the NB EU Medical Device Regulations 3
L ETO vs. Gamma Irradiation Sterilization of DBM (Demineralized Bone Matrix) Other Medical Device and Orthopedic Related Topics 3
M Effect of double ETO (Ethylene Oxide) Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 13
S AAMI TIR28:2009 Product adoption & process equivalency for ETO sterilization Other US Medical Device Regulations 2
P ETO Sterilisation of Single Use Part supplied in Non-Sterile State EU Medical Device Regulations 12
M Eto Certificate of Sterilization (vs. gamma irradiation) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
B ISO 11135:2007 (EtO Sterilization) ISO 13485:2016 - Medical Device Quality Management Systems 5
T ETO Sterilization - Process monitoring and out-inspection ISO 13485:2016 - Medical Device Quality Management Systems 3
M Japanese rule on EtO residuals in sterilized medical devices? Various Other Specifications, Standards, and related Requirements 16
G EtO Sterilization - DIN/EN 550 - ISO 11135 Other US Medical Device Regulations 20
Ajit Basrur Audit Checklist for Contract Gamma and ETO Sterilization needed ISO 13485:2016 - Medical Device Quality Management Systems 3
V EMS for Medical Device Facility with Ethylene Oxide (ETO) Sterilization process. Miscellaneous Environmental Standards and EMS Related Discussions 4
T ETO Product Sterilisation Testing Standard Requirements ISO 13485:2016 - Medical Device Quality Management Systems 4
V ETO Sterilization of Urological Catheter ISO 13485:2016 - Medical Device Quality Management Systems 5
Q EtO vs. Gamma Sterilization - Stainless Steel and Titanium Orthopaedic Implants ISO 13485:2016 - Medical Device Quality Management Systems 7
C How to verify the SAL (sterilie acceptable level) is 10-6 for ETO Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 6
G Supplier audit of ETO Sterilization Lab - Need a checklist or advice Supplier Quality Assurance and other Supplier Issues 3
P ETO Sterilization Validation Process and an NCR - ISO 13485:2003 CMDCAS audit ISO 13485:2016 - Medical Device Quality Management Systems 1
G UVC sterilizer CE Mark CE Marking (Conformité Européene) / CB Scheme 5
K Our local sterilizer is closing - Record Retention Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1