EU Harmonized Standards to which a company has declared compliance
DCONN1
I'm in the process of reviewing a list of standards to which a company has declared compliance for their product. I am unsure if some standards noted as harmonized on this list are currently harmonized in the EU as they are not listed on the relevant page on the EC website (which contains a disclaimer stating that the list may not be complete or fully up to date anyway).
Medical devices - European Commission
As examples, EN ISO 11135:2014, EN ISO 11138-1:2006, and EN ISO 15223-1:2012 are absent from the European harmonized standards list published on the EC website. Are these standards currently harmonized? Is the EC page with standards listing the best source for determining which standards are currently harmonized in the EU?
shimonv
Hi DCONN1,
This is the official source for harmonized standards. I can tell specifically that ISO 15223-1:2012 became harmonized at one point but later on was withdrawn and EN 980 was reinstate. So, in effect it is up to date.
Shimon
Mark Meer
Yes, this is the list I reference...
If a standard has been updated but is not yet on the list, it is a pretty simple matter to justify applying the updated one as a better reflection of "state-of-the-art" in your technical file... This approach has worked fine for us (and accepted by our NB) in the past...
Also, some standards are simply referenced from ones on the list (even though they do not appear themselves on the list). For example, EN ISO 10993-10 is not on the list, but is referenced by 10993-1 (which is on the list).
Ronen E
Problem Solver
Moderator
The website disclaimer is mostly a formality, but if you want to be 100% sure you can check the actual OJEU record. Whatever is in the latest OJEU record is what’s harmonized.
Ronen E
Problem Solver
Moderator
ISO 10993-1 suggests considering tests specified in ISO 10993-10, so that reference does not confer any “implied harmonized status” on the latter.
Let’s not forget what harmonizing standards is all about in the MDD context. It is about equating a published standard compliance with some (specified) MDD requirements compliance, via the “presumption of conformity”. Complying with ISO 10993-1 (and this can sometimes be done without any testing) grants presumption of conformity with some ER; complying with ISO 10993-10, in itself, doesn’t grant any such presumption.
Mark Meer
It is true that ISO 10993-1 suggests tests, but if assessed as necessary, the standard then points to an associated 10993-x standard.
If, for example, you've assessed that a cytotoxicity test is necessary, and then did NOT do the cytotoxicity test as per 10993-5, then presumably you can't claim compliance to 10993-1, no?
Also, to use another example, IEC 62133 (Safety requirement for secondary cells) is not a harmonized standard, but is referenced by EN 60601-1. Therefore, wouldn't it be safe to assume that compliance to IEC 62133, even though it is not on the list of harmonized standards, bears some evidence of conformity with respect to electrical safety vis-a-vis its relationship to EN 60601-1?
I've mentioned this in another thread already, but I have to admit to being still confused in this case. EN ISO 10993-10:2013, although not presently on the list of harmonised standards, does have an Annex ZA listing the presumptions of conformity with ERs. Granted, the annex does state "Once this Standard is cited in the Official Journal of the European Union...", so I presume that the Z-Annex is not "official" until officially harmonized.
That being said, I think that it is very unlikely that an NB would reject or question the use of such a standard to evidence ER conformity as (a) it is tied to (harmonized) 10993-1; (b) it has a Z annex.
Ronen E
Problem Solver
Moderator
No. ISO 10993-1:2009 refers to parts like -10 and -5 in a rather neutral way (“see ISO 10993-X”). I was never in a situation where an alternative test method was used so I can’t say what would have been the outcome, but I wouldn’t be surprised if the NB would have insisted that it’s a NC because the ISO 10993 part method wasn’t used, despite no such clear requirement in part 1. As mentioned in some threads here, ISO 10993-1 is routinely misread and misapplied. :(
I don’t know these standards well enough to comment. The formal relationship should be determined based on the actual language in the standards. Again, I wouldn’t be surprised if this is practiced (by NB) in contrast with the actual language or intent, though...
If I’m not mistaken, this standard was once harmonized, then taken off the list. I can’t remember the reason (not sure I ever even fully knew it); there might have been some discussion about it in Elsmar.
The NB would probably not reject because it can be conveniently considered “state of the art” (unless proved otherwise). As I stated above, part 10 is not “tied” to part 1, only loosely associated. The existence of an annex Z in a standard has no binding meaning, it’s just a guidance (and a hypothetical one if the standard is not officially harmonized).
Ronen E
Problem Solver
Moderator
You can find/follow harmonised standards from only following link; CEN - Advanced search - Publications and Work in Progress
Journal is not updated. EN 980 is withdrawn ages ago they forget to remove from journal.
These statements are strange to me.
CEN prepare European standards but I don’t think they have the authority to actually harmonize them.
The OJEU publication is what gives force to the harmonisation. It is regrettable if OJEU publications lag behind or are erroneous (as sometimes happens), but they are legally binding nevertheless, at least as far as I know.
The EN 980 / ISO 15223 ordeal has been discussed here when it happened, and again later. Apparently it was a collection of mistakes, some of which have been rectified since.
Remus
CEN is preparing the Annex ZA/ZB of standards then send to EC, if they accept it is harmonised and updated on CEN site. However, EC guys are too lazy to update journal...
So don't bother journal...
M Subscribe to harmonized standards updates EU Medical Device Regulations 5
A Common Specifications versus Harmonized Standards EU Medical Device Regulations 3
T Justification of use of non-harmonized standards for MDR conformity Other Medical Device Related Standards 12
C By when should harmonized standards be complied with? EU Medical Device Regulations 5
C Deadline for compliance with added MDR harmonized standards CE Marking (Conformité Européene) / CB Scheme 2
Ed Panek MDR Harmonized Standards Updated EU Medical Device Regulations 4
M Why are EU harmonized particular standards so old & out-of-date? IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
M MDR harmonized standards EU Medical Device Regulations 7
N Harmonized Standards for MDR Certification EU Medical Device Regulations 7
R Applicability of new non-harmonized standards (MDD/MDR) EU Medical Device Regulations 14
M EU Harmonized Standards - Withdrawals and Replacements EU Medical Device Regulations 14
Q Compliance to EU Official Journal Harmonized Standards (MD/IVD products in EU) EU Medical Device Regulations 7
A 93/42/EEC Declaration of Conformity and Harmonized Standards EU Medical Device Regulations 12
M Harmonized standards vs. EN ISO References EU Medical Device Regulations 5
Marc MDD Harmonized List of Standards Website Moved Other Medical Device and Orthopedic Related Topics 1
W New list of MDD Harmonized Standards Other Medical Device Related Standards 2
D Harmonized Standards EN vs. IEC CE Marking (Conformité Européene) / CB Scheme 11
R No Stability Standard on latest List of Harmonized Standards IVDD ISO 13485:2016 - Medical Device Quality Management Systems 2
F ISO 10993-10 - Removed from List of Harmonized Standards Other Medical Device Related Standards 4
S Annex Z of Harmonized Standards EU Medical Device Regulations 3
P Retrospective Application of EN 60601-1:2006 - Harmonized Standards IEC 60601 - Medical Electrical Equipment Safety Standards Series 27
O Harmonized Standards on Medical Device IFU (Instructions for Use) ISO 13485:2016 - Medical Device Quality Management Systems 7
xcanals_tecno-med.es Revised List of Harmonized Standards for Medical Device Directives (MDD) in the EU EU Medical Device Regulations 48
E Types of Standards and Harmonized Standards Other ISO and International Standards and European Regulations 6
xcanals_tecno-med.es Revised list of harmonized standards for medical device directives in the EU (7/7/10) EU Medical Device Regulations 0
Sidney Vianna The Future Structure of ISO Management System Standards - from HLS to HS (Harmonized Structure) June 2021 Other ISO and International Standards and European Regulations 70
R Medical Device: New List of Harmonized Standards in EU under MDD93/42/EEC EU Medical Device Regulations 6
M Revised list of harmonized standards for medical device directives in the EU EU Medical Device Regulations 1
W New Harmonized Standards MDD and IVDD EU Medical Device Regulations 5
S EC Official Journal Harmonized Standards - Additional requirements? ISO 13485:2016 - Medical Device Quality Management Systems 3
W New List of Medical Device Harmonized Standards - MDD CE Marking (Conformité Européene) / CB Scheme 5
Al Rosen List of Harmonized Standards for the MDD, IVD and IMDD - June 2006 ISO 13485:2016 - Medical Device Quality Management Systems 5
amjadrana CE Mark & Harmonized Standards EU Medical Device Regulations 8
Y Retroactive application of harmonized standards US Food and Drug Administration (FDA) 2
Y CE Mark & Harmonized Standards - Article 5.1, MDD - Has anyone challanged this? EU Medical Device Regulations 5
H Harmonized Standard for the refrigerated lab device safety CE Marking (Conformité Européene) / CB Scheme 2
R ISO 14971 not harmonized ISO 14971 - Medical Device Risk Management 5
B European Harmonized vs IEC standard IEC 62366 - Medical Device Usability Engineering 2
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
G Harmonized List - MDD 93/42/EEC Other ISO and International Standards and European Regulations 2
P ISO 10993-1:2018 - When will the European equivalent become a Harmonized Standard? Other Medical Device Related Standards 13
R Information on the roll out of the new harmonized DRAFT FMEA AIAG/VDA manual IATF 16949 - Automotive Quality Systems Standard 4
M EN Standard (ETSI EN 301 489-3 v1.6.1) is Harmonized but not on Europa List Other ISO and International Standards and European Regulations 3
M Harmonized EN 60601 Z Annexes are Not Useful IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Is ISO 14644 harmonized for US? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
O Harmonized standard of EU---EN 60601-2-10:2000/A1:2001 EU Medical Device Regulations 1
M Labelling and the harmonized EN ISO 14971:2012 ISO 14971 - Medical Device Risk Management 9