Qestion is on the EU MDD requirements and audit non-conformance.

For CE Mark Assessment on a medical device client received a non-conformance
because standards referenced in product spec and technical file docs were
not checked for updates to standards referenced and files were not assessed for update impacts.

Non-conformance was "Assessment of new/revised standards impact on
existing devices needs to be part of Company's process. The review should
be documented and action plans developed."

Applicable Standard Cited: Continued Compliance to MDD Annex 1: Ref Annex I, II

First, does observation make sense to you.
2nd, Does anyone have a checklist, process or flowchart that depicts or guides
the needed review?

Alternately How do you (posting respondents) assure most current standard revision like electronics (IEC 60601-?-? ) called out is updated in technical file and file checked for impacts?

Thanks BILLH007
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
Re: EU MDD Requirement on impact of Rev STDs

Alternately How do you (posting respondents) assure most current standard revision like electronics (IEC 60601-?-? ) called out is updated in technical file and file checked for impacts?
We keep track via management effort. We hear about revised standards here, from other informal notification means, from our NB, from our Authorized Rep, and in some cases from standards organizations. When a revision or update process is underway, we create a project to review and update our compliance stance in a timely manner, generally involving Engineering and Regulatory, plus perhaps Marketing and sometimes Sales.
 

Michael Malis

Quite Involved in Discussions
#3
Qestion is on the EU MDD requirements and audit non-conformance.

For CE Mark Assessment on a medical device client received a non-conformance
because standards referenced in product spec and technical file docs were
not checked for updates to standards referenced and files were not assessed for update impacts.

Non-conformance was "Assessment of new/revised standards impact on
existing devices needs to be part of Company's process. The review should
be documented and action plans developed."

Applicable Standard Cited: Continued Compliance to MDD Annex 1: Ref Annex I, II

First, does observation make sense to you.
2nd, Does anyone have a checklist, process or flowchart that depicts or guides
the needed review?

Alternately How do you (posting respondents) assure most current standard revision like electronics (IEC 60601-?-? ) called out is updated in technical file and file checked for impacts?

Thanks BILLH007
First, observation make sense.
In addition to the above, we also getting updates from our participation with Standards Committees and subscription to e-service...
Second, you also need to identify "the gaps" from most current standard to old revision. This will allow you to address what is required in TF. Don't forget a Risk Assessment for TF while doing this.
 

Peter Selvey

Staff member
Super Moderator
#4
Just to confirm it is a valid non-conformance. The way the MDD works is that if you don't apply a current harmonised standard (even to older designs) you need to provide explanation of alternate solutions. To do this you still need to know what is in the harmonised standard even if you don't want to comply with it, and then document the alternate solution.

But keep in mind this is only for harmonised standards, which means you only need to keep an eye on Europa website for updates. Transition periods of 2-3 years also give plenty of warning. Keeping an eye out for new IEC or ISO documents can give up to 3-5 years warning.
 
B

BILLH007

#5
Thanks Peter,

So if technical file has no risk plan/report as per ISO 14971 now needs added. Or IEC 60601-1-? required safety etc the documentation needs added to technical file prior to superceded date.

Did I get that right?

Bill Hufford
 

Peter Selvey

Staff member
Super Moderator
#6
For ISO 14971:2000, definitely yes since it has been harmonised since April 2004 (replacing EN 1441 on that date). So, nearly 6 years ago, and a high profile standard as well.

Little known fact about risk, though - the MDD specifically states that:

Whereas the essential requirements and other requirements set out in the Annexes to this Directive, including any reference to [FONT=TimesNewRoman+20][FONT=TimesNewRoman+20][/FONT][/FONT]minimizing[FONT=TimesNewRoman+20][FONT=TimesNewRoman+20][/FONT][/FONT]or [FONT=TimesNewRoman+20][FONT=TimesNewRoman+20][/FONT][/FONT]reducing[FONT=TimesNewRoman+20][FONT=TimesNewRoman+20][/FONT][/FONT]risk must be interpreted and applied in such a way as to take account of technology and practice existing at the time of design and of technical and economical considerations compatible with a high level of protection of health and safety;

and


"..the essential requirements should be applied with discretion to take account of the technological level existing at the time of design and of technical and economic considerations compatible with a high level of protection of health and safety"


Both of these statements give you some scope to relax on updating older equipment, but the phrase "... compatible with a high level of protection..." means you still need to keep an eye on things for serious risks.​

As suggested before you should be aware of the content of new standards are and make a decision on what to apply, keep a record of these decisions. In practice this usually means a report (checklist) showing each clause sequentially, and linking to supporting evidence if needed.​
 

Michael Malis

Quite Involved in Discussions
#7
For ISO 14971:2000, definitely yes since it has been harmonised since April 2004 (replacing EN 1441 on that date). So, nearly 6 years ago, and a high profile standard as well.

Little known fact about risk, though - the MDD specifically states that:

Whereas the essential requirements and other requirements set out in the Annexes to this Directive, including any reference to [FONT=TimesNewRoman+20][FONT=TimesNewRoman+20][/FONT][/FONT]minimizing[FONT=TimesNewRoman+20][FONT=TimesNewRoman+20][/FONT][/FONT]or [FONT=TimesNewRoman+20][FONT=TimesNewRoman+20][/FONT][/FONT]reducing[FONT=TimesNewRoman+20][FONT=TimesNewRoman+20][/FONT][/FONT]risk must be interpreted and applied in such a way as to take account of technology and practice existing at the time of design and of technical and economical considerations compatible with a high level of protection of health and safety;

and


"..the essential requirements should be applied with discretion to take account of the technological level existing at the time of design and of technical and economic considerations compatible with a high level of protection of health and safety"


Both of these statements give you some scope to relax on updating older equipment, but the phrase "... compatible with a high level of protection..." means you still need to keep an eye on things for serious risks.​

As suggested before you should be aware of the content of new standards are and make a decision on what to apply, keep a record of these decisions. In practice this usually means a report (checklist) showing each clause sequentially, and linking to supporting evidence if needed.​
14971 latest revision is 2007. So if you plan to do something from scratch, you may as well to comply to the latest standard...
 
Thread starter Similar threads Forum Replies Date
N Label Placement Requirement in MDD/MDR EU Medical Device Regulations 1
chris1price Is there a regulatory (MDD, MDR, FDA, ISO, etc) requirement to perform a mock recall? CE Marking (Conformité Européene) / CB Scheme 4
K Annual License or Registration Requirement for Europe (MDD 93/42 EEC) EU Medical Device Regulations 1
A Requirement 4 of MDD 93/42/EEC - Definitions of Lifetime, Life Cycle, Shelf Life EU Medical Device Regulations 5
Sam Lazzara MDD Essential Requirement 13.6(q) regarding IFU revision CE Marking (Conformité Européene) / CB Scheme 19
D MDD 2007/ - Essential Requirement Checklist - Physical and Biological Properties EU Medical Device Regulations 4
D Requirement of "State Of The Art" in MDD (Medical Device Directive) EU Medical Device Regulations 42
S MDD Requirement - SOP for on how to translate from one language to another ISO 13485:2016 - Medical Device Quality Management Systems 5
Ed Panek Notified Body Substantial Change MDD EU Medical Device Regulations 1
J How to keep MDD certificate valid when legal manufacturer has liquidity problem EU Medical Device Regulations 0
Ed Panek MDD Cleared Private Labelled Devices into EU EU Medical Device Regulations 0
K Section 13.6 Annex 1 MDD 93/42/EEC for Glove EU Medical Device Regulations 0
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
U MDD Article 12/ MDr Article 22 interpretation EU Medical Device Regulations 2
E Our company is planning to file MDD not MDR next month. Do we require to show chemical characterization report ? CE Marking (Conformité Européene) / CB Scheme 2
S Converting SOP MDD to MDR Noob ISO 13485:2016 - Medical Device Quality Management Systems 4
silentmonkey Are risks in supply chain and development activities within scope of MDD? EU Medical Device Regulations 4
A First CE-mark class III implantable device under MDD EU Medical Device Regulations 8
Q MDD - distributor EU Medical Device Regulations 1
R MDD x PPE Directive - Statement of Non-Applicability EU Medical Device Regulations 3
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 3
M Requirements for an MDD approved device after 26/05/2021 EU Medical Device Regulations 3
H MDD CM, MDR manufacturer. 2020. Is it ok? EU Medical Device Regulations 1
S Technical File for Surgical Mask in Class 1 MDD Elsmar Cove Forum ToS and Forum Policies 1
S MDR Delay - MDD design Change? (before new MDR DOA) EU Medical Device Regulations 8
P What ASTM std is EC 93/42 MDD EQUIVALENT TO CE Marking (Conformité Européene) / CB Scheme 1
R Applicability of new non-harmonized standards (MDD/MDR) EU Medical Device Regulations 14
K Measuring Function MEDDEV 2.1/5 relation with MDD 93/42/EEC CE Marking (Conformité Européene) / CB Scheme 1
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
Z Economic Operators Responsibilities after DOA for legacy MDD devices EU Medical Device Regulations 2
J MDD to MDR transition - Time required for the implementation of the MDR EU Medical Device Regulations 7
F MDR system/procedurepack (article 22) with device acc. to MDD and MDR CE Marking (Conformité Européene) / CB Scheme 2
I Maintaining Technical Documentation under MDD and MDR EU Medical Device Regulations 1
P Adding "new" medical device to existing CER - under MDD CE Marking (Conformité Européene) / CB Scheme 1
Watchcat Duration of MDD Certification EU Medical Device Regulations 4
Z MDD ER 7.5, "phthalate test"? Aspiration tips EU Medical Device Regulations 3
R Notified Body for MDD 1Q20 - Florida Company Registrars and Notified Bodies 4
S Until when can we CE-mark and sell MDD class 1 devices? CE Marking (Conformité Européene) / CB Scheme 17
D CE Marking Requirements MDD & MDR - new product development covered under same scope EU Medical Device Regulations 1
G Harmonized List - MDD 93/42/EEC Other ISO and International Standards and European Regulations 2
B Accredited laboratory or not under MDD 93/42/EEC IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
T Custom made medical device MDD vs MDR definition CE Marking (Conformité Européene) / CB Scheme 12
Watchcat MDD, MDR/IVDR, Working Groups, Expert Panels, MEDDEVs and Other EU Guidances EU Medical Device Regulations 32
D MDD Certification Validity after May 2020 EU Medical Device Regulations 12
A Compatibility between MDD and MDR devices EU Medical Device Regulations 7
M Device CE certified under MDD / MDR > new GTIN? EU Medical Device Regulations 7
N When is "placing on the market"? MDD certified medical devices EU Medical Device Regulations 32
G Essential Requirements Checklist - (MDD 93/42 / EEC) Misc. Quality Assurance and Business Systems Related Topics 2
K EU MDD Class I Device market access / certification after 26th May 2020 EU Medical Device Regulations 9

Similar threads

Top Bottom