Hello everyone,
I have a question on Procedure Packs and the TD / classification / route to conformity.
Under MDD we have always been claiming compliance to Art. 12 on the TD for procedure packs, Class Is.
During our MDR Audit the auditor wrote a NC on my TD / DoC as i had stated
"procedure packs are classified through EU MDR 2017/745 rule 1, Art.22.3 and Annex IX". I also called Art 22.3 on the DoC.
The auditor raised the point " but there is no statement that for those procedure packs, sterilization has been carried out in accordance with the manufacturer's instructions."
I do not have this statement and would be shocked if every procedure pack producer had it. I have requested it from each manufacturer of the CE marked devices inside the procedure pack but no manufacturer will provide.
I have an audit in 2 weeks to close the NCs and have become stuck on this. I tried to justify it during the audit saying we check the labelling of each device to ensure they do not state do not resterilise, and the fact that we have done this for 20+ years with no complaints, recalls etc. but the auditor was having none of it.
Can anyone help me here please?
I have a question on Procedure Packs and the TD / classification / route to conformity.
Under MDD we have always been claiming compliance to Art. 12 on the TD for procedure packs, Class Is.
During our MDR Audit the auditor wrote a NC on my TD / DoC as i had stated
"procedure packs are classified through EU MDR 2017/745 rule 1, Art.22.3 and Annex IX". I also called Art 22.3 on the DoC.
The auditor raised the point " but there is no statement that for those procedure packs, sterilization has been carried out in accordance with the manufacturer's instructions."
I do not have this statement and would be shocked if every procedure pack producer had it. I have requested it from each manufacturer of the CE marked devices inside the procedure pack but no manufacturer will provide.
I have an audit in 2 weeks to close the NCs and have become stuck on this. I tried to justify it during the audit saying we check the labelling of each device to ensure they do not state do not resterilise, and the fact that we have done this for 20+ years with no complaints, recalls etc. but the auditor was having none of it.
Can anyone help me here please?