SBS - The best value in QMS software

EU MDR and PSUR (Periodic Safety Update Report)

#1
MDR Article 86 Periodic safety update report reads:

1.Manufacturers of class IIa, class IIb and class III devices shall prepare a periodic safety update report (‘PSUR’) for each device and where relevant for each category or group of devices summarising the results and conclusions of the analyses of the post-market surveillance data gathered as a result of the post-market surveillance plan referred to in Article 84 together with a rationale and description of any preventive and corrective actions taken. ...


Two questions:
1. PSUR is a well known requirement in the pharma world (see attached template from EMA (virus scan before use )). Is it now required for all medical devices in class IIa or higher even if they do not administer medicinal products?
2. Does anyone already have a decent PSUR template and is willing to share it?

Many thanks to the loyal readers (and am I glad that Elsmar is back)
 

Attachments

Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Well, I was hoping someone with expertise would reply but maybe this will help move the discussion.

My understanding is that the PSUR is new to the MDR and it will be required for Class IIa, IIb, and III.

As far as the content of the report goes, just a little further down in Article 86 says:
(a) the conclusions of the benefit-risk determination;
(b) the main findings of the PMCF; and
(c) the volume of sales of the device and an estimate evaluation of the size and other characteristics of the population using the device and, where practicable, the usage frequency of the device.

Then running the thread, the PMCF is addressed in MEDDEV 2.12/2 rev 2. I guess that cover a pretty wide range of things so maybe a template would be difficult?

Hopefully someone has cracked this nut and can share / enlighten. (I'm glad you asked the question!).
 
#4
According my information the European Commission will release a template for the PSUR. Should have happened in December, but as so often they are delayed. Let's hope it will come soon. I personally am waiting with drafting a template.
 
#5
Thank you all for this discussion and input on PSUR. Here are my insights on PSUR being into the EU MDR field.

First of all, Yes, PSUR for medical devices is now a requirement for MDR compliance.

PSUR will be required for all devices classified as IIa, IIb or III. The scope is not limited to just for devices that are intended to administer and/or remove medicinal substances. This requirement will impact all devices (including custom devices) except class I. (As a side note, class I devices will still require PMS report and should be a part of Technical Documentation; Class I PMSR should be updated when necessary and should be made available to the competent authority when requested by a manufacturer).

An Annual update will be needed for PSURs falling under class IIb and III. A bi-annual update will be needed for PSURs falling under class IIa. (for class IIa, the PSUR update can be earlier than 2 years depending on the changes)

All PSURs will be required to be the part of Technical Documentation (Currently, termed as technical File). PSURs for custom devices will be required to be the part of its documentation (not formally termed as technical Documentation for Custom Devices).

PSURs for any class III or any implantable device will be required to be submitted to NB through EUDAMED for review and assessment. It is not clear to me whether a manufacturer or an NB will be required to submit this evaluated PSUR to Competent Authority but I am sure that it should be made available to the Competent Authority after NB's review. (Further reading Art. 86(3) gives a little more clarity that it mostly should be a manufacturer).
PSUR for any class IIa/IIb (non-implantable) devices will be required to be submitted to NB under conformity assessment review and to the competent authority, only when requested.

Hope this is helpful!
Thanks.
 

jiligan

Involved In Discussions
#6
I am currently working on MDR and came across your post since I am looking for a template for the PSUR also. Have you had any luck finding one that will work for you? Thanks!
 
1

14cross

#8
Any update on this? I did a search on the internet but could not find any info on the EU releasing templates for the PSUR.

Thanks!
 
G

GLM12345

#9
There is a template for the PSUR, which is nice because takes the guessing out of how this this thing should look. I cannot figure out how to post a thread here, need to figure that out. My question is given the scope of the new EU MDR, why would a PMS PLan be required for a device that is not even sold/marketed in the Union? EU is concerned about devices in their jurisdiction, versus a China only product, what good is a PM PLan for that if its not even distributed in EU?

:bonk:
 
1

14cross

#10
Can you provide a link to this template you reference? I have searched and only find templates for drugs, not medical devices.
 
Thread starter Similar threads Forum Replies Date
S PSUR - MDR First Submission Requirement? EU Medical Device Regulations 2
L EU MDR Article 86 PSUR EU Medical Device Regulations 1
A MDR ANNEX XI part A or Part B EU Medical Device Regulations 0
N OBL under the MDR. Which are the obligations of the OEM? EU Medical Device Regulations 0
U Distributor obligations under MDR EU Medical Device Regulations 1
Raisin picker CECP, consultation according to Art 54 MDR EU Medical Device Regulations 1
K Question on MDR classification EU Medical Device Regulations 4
M STED vs MDR EU Medical Device Regulations 0
F UDI-PI required on packaging (MDR) EU Medical Device Regulations 2
M Extension to shelf life of new product under MDR EU Medical Device Regulations 1
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
K Importer in EU, but not distribute in the EU - MDR obligations EU Medical Device Regulations 0
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
M Responsibilities of Importers / Distributors - MDR EU Medical Device Regulations 3
R EU-MDR Article 9-common specifications Other ISO and International Standards and European Regulations 2
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
JoCam MDR Gap Analysis - Consultant fees EU Medical Device Regulations 10
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
M Examples of Combination Products - MDR Article 1 (8) and MDR Article 1(9) Medical Device and FDA Regulations and Standards News 3
P EU MDR Translation of Product Name EU Medical Device Regulations 0
N Is this a single integral drug device combination product EU MDR CE Marking (Conformité Européene) / CB Scheme 1
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
K Quality Agreement MDR - The manufacturer is outside of EU Other Medical Device Related Standards 4
M MDR - eIFU - When you can supply an electronic IFU instead of a paper IFU EU Medical Device Regulations 5
M MDR harmonized standards EU Medical Device Regulations 3
cscalise RAPS/Med Dev EU MDR Guidebook - Reviews or Opinions? EU Medical Device Regulations 0
V New Product Model CE Mark Certification under MDR EU Medical Device Regulations 0
E MDR PRRC "jointly responsible" EU Medical Device Regulations 1
J Strategy for MDR Regulatory Compliance Procedure ISO 13485:2016 - Medical Device Quality Management Systems 4
S EU MDR Annex I Chapter III 23.4 z EU Medical Device Regulations 2
R MDR & depth of technical files review by NB's EU Medical Device Regulations 18
U EU-MDR: Complicated supply chain EU Medical Device Regulations 4
C EU MDR - Annex II 6.1 Pre-clinical and clinical data EU Medical Device Regulations 4
A IVD in MDR procedure pack EU Medical Device Regulations 2
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
S Are EC type examinations still being conducted under MDR? EU Medical Device Regulations 5
U MDD Article 12/ MDr Article 22 interpretation EU Medical Device Regulations 2
E MDR internal audit Internal Auditing 1
R The term "Benefit Risk Ratio" in EU MDR, do I need to present benefit risk analysis as a RATIO Risk Management Principles and Generic Guidelines 4
R MDR, Annex I, 23.1 Interpretation - IFU on the website EU Medical Device Regulations 5
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
S EU MDR CAPA - GAP Assessment on CAPA SOPs EU Medical Device Regulations 1
M MDR Article 22 Interpretation EU Medical Device Regulations 3
A User interface Language Requirements - EU MDR EU Medical Device Regulations 1
S How is the service life for class I reusable medical devices described in MDR? EU Medical Device Regulations 7
S How is the service life described for class I reusable medical devices in MDR? Elsmar Xenforo Forum Software Instructions and Help 1
P MDR/IVDR Annex III - Technical documentation on PMS EU Medical Device Regulations 1
E Our company is planning to file MDD not MDR next month. Do we require to show chemical characterization report ? CE Marking (Conformité Européene) / CB Scheme 2
S Classification of a product according to MDR EU Medical Device Regulations 3
L Classification under MDR EU Medical Device Regulations 1

Similar threads

Top Bottom