SBS - The Best Value in QMS software

EU MDR and PSUR (Periodic Safety Update Report)

#1
MDR Article 86 Periodic safety update report reads:

1.Manufacturers of class IIa, class IIb and class III devices shall prepare a periodic safety update report (‘PSUR’) for each device and where relevant for each category or group of devices summarising the results and conclusions of the analyses of the post-market surveillance data gathered as a result of the post-market surveillance plan referred to in Article 84 together with a rationale and description of any preventive and corrective actions taken. ...


Two questions:
1. PSUR is a well known requirement in the pharma world (see attached template from EMA (virus scan before use )). Is it now required for all medical devices in class IIa or higher even if they do not administer medicinal products?
2. Does anyone already have a decent PSUR template and is willing to share it?

Many thanks to the loyal readers (and am I glad that Elsmar is back)
 

Attachments

Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Well, I was hoping someone with expertise would reply but maybe this will help move the discussion.

My understanding is that the PSUR is new to the MDR and it will be required for Class IIa, IIb, and III.

As far as the content of the report goes, just a little further down in Article 86 says:
(a) the conclusions of the benefit-risk determination;
(b) the main findings of the PMCF; and
(c) the volume of sales of the device and an estimate evaluation of the size and other characteristics of the population using the device and, where practicable, the usage frequency of the device.

Then running the thread, the PMCF is addressed in MEDDEV 2.12/2 rev 2. I guess that cover a pretty wide range of things so maybe a template would be difficult?

Hopefully someone has cracked this nut and can share / enlighten. (I'm glad you asked the question!).
 
#4
According my information the European Commission will release a template for the PSUR. Should have happened in December, but as so often they are delayed. Let's hope it will come soon. I personally am waiting with drafting a template.
 
#5
Thank you all for this discussion and input on PSUR. Here are my insights on PSUR being into the EU MDR field.

First of all, Yes, PSUR for medical devices is now a requirement for MDR compliance.

PSUR will be required for all devices classified as IIa, IIb or III. The scope is not limited to just for devices that are intended to administer and/or remove medicinal substances. This requirement will impact all devices (including custom devices) except class I. (As a side note, class I devices will still require PMS report and should be a part of Technical Documentation; Class I PMSR should be updated when necessary and should be made available to the competent authority when requested by a manufacturer).

An Annual update will be needed for PSURs falling under class IIb and III. A bi-annual update will be needed for PSURs falling under class IIa. (for class IIa, the PSUR update can be earlier than 2 years depending on the changes)

All PSURs will be required to be the part of Technical Documentation (Currently, termed as technical File). PSURs for custom devices will be required to be the part of its documentation (not formally termed as technical Documentation for Custom Devices).

PSURs for any class III or any implantable device will be required to be submitted to NB through EUDAMED for review and assessment. It is not clear to me whether a manufacturer or an NB will be required to submit this evaluated PSUR to Competent Authority but I am sure that it should be made available to the Competent Authority after NB's review. (Further reading Art. 86(3) gives a little more clarity that it mostly should be a manufacturer).
PSUR for any class IIa/IIb (non-implantable) devices will be required to be submitted to NB under conformity assessment review and to the competent authority, only when requested.

Hope this is helpful!
Thanks.
 

jiligan

Involved In Discussions
#6
I am currently working on MDR and came across your post since I am looking for a template for the PSUR also. Have you had any luck finding one that will work for you? Thanks!
 
1

14cross

#8
Any update on this? I did a search on the internet but could not find any info on the EU releasing templates for the PSUR.

Thanks!
 
G

GLM12345

#9
There is a template for the PSUR, which is nice because takes the guessing out of how this this thing should look. I cannot figure out how to post a thread here, need to figure that out. My question is given the scope of the new EU MDR, why would a PMS PLan be required for a device that is not even sold/marketed in the Union? EU is concerned about devices in their jurisdiction, versus a China only product, what good is a PM PLan for that if its not even distributed in EU?

:bonk:
 
1

14cross

#10
Can you provide a link to this template you reference? I have searched and only find templates for drugs, not medical devices.
 
Thread starter Similar threads Forum Replies Date
S PSUR - MDR First Submission Requirement? EU Medical Device Regulations 2
L EU MDR Article 86 PSUR EU Medical Device Regulations 1
E Article 17 MDR - (Re-)Processing EU Medical Device Regulations 0
H Article 22 MDR System EU Medical Device Regulations 15
L REACH compliance with MDR EU Medical Device Regulations 1
L EU MDR Ramifications for no expiration date on labeling EU Medical Device Regulations 1
R How to obtain OBL licence under Indian MDR, 2017 Other Medical Device Related Standards 2
F MDR – Article 120 – Transitional provisions EU Medical Device Regulations 2
H MDR Article 13(c) EU Medical Device Regulations 6
Ed Panek Auditor MDR (Presub audit) finding EU Medical Device Regulations 2
M Classification of Instruments under EU MDR EU Medical Device Regulations 1
M Procedure for vigilance system according to new MDR EU Medical Device Regulations 0
M Procedure for clinical evaluation according to new MDR EU Medical Device Regulations 1
R MEDDEV 2.12-1 rev 8 (Vigilance guidelines) still applicable with the MDR implementation? EU Medical Device Regulations 1
S Determining a device category according to the MDR EU Medical Device Regulations 3
A Transactions under MDR Medical Device and FDA Regulations and Standards News 3
H Has anyone undergone MDR FQA review yet? EU Medical Device Regulations 10
S UK MDR + EU MDR Declaration of Conformity UK Medical Device Regulations 0
D Does Risk Management apply to re-labeler (MDR) EU Medical Device Regulations 1
H MDD VS MDR 2002-218 UK Medical Device Regulations 6
Ed Panek MDR Liability Insurance EU Medical Device Regulations 1
S MDD to MDR - Tallow Derivatives Impact EU Medical Device Regulations 1
P Update on NBOG 2014-3 to address MDR/IVDR, or any plan to do that? EU Medical Device Regulations 0
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
S MDR GSPR Standards EU Medical Device Regulations 1
A MDR - Legacy Device Review Timeframe and Requirements EU Medical Device Regulations 3
J Translation requirements for the statement referred to in ANNEX XIII of MDR EU Medical Device Regulations 0
J EU MDR GSPR 10.4.3 and 10.4.4 EU Medical Device Regulations 2
F Labelling to comply with both FDA and MDR US Food and Drug Administration (FDA) 6
N EU MDR and its impact on existing Registrations EU Medical Device Regulations 1
M Post Market Safety Update Report for devices that will be up-classified under MDR but are currently under MDD EU Medical Device Regulations 3
Y MDR Transition - make "Available on the market" after CE certificate expiration EU Medical Device Regulations 3
C MDR Classification Rule 10 EU Medical Device Regulations 14
C MDR software classification EU Medical Device Regulations 12
Y MDR requirements for Class I accessories EU Medical Device Regulations 2
M MDR legal actions - manufacturers CE Marking (Conformité Européene) / CB Scheme 8
R Products not within the MDR grace period EU Medical Device Regulations 1
shimonv MDR transition checklist EU Medical Device Regulations 0
K Re-packaging under MDR EU Medical Device Regulations 3
Ed Panek Upcoming NB MDD ---> MDR Crunch EU Medical Device Regulations 0
S MDR - System and procedure pack article 22 and all sub processes that apply ISO 13485:2016 - Medical Device Quality Management Systems 0
C Requirements for distributors under MDR: translation EU Medical Device Regulations 0
A Should we assign the PRRC before the date of application of MDR (26 May 2021)? EU Medical Device Regulations 0
N Looking for a recommendation for an EU MDR Importer EU Medical Device Regulations 1
JoCam Rental and MDD versus MDR EU Medical Device Regulations 1
A Can anyone share a Distribution Agreement template under MDR 2017/745? EU Medical Device Regulations 0
A MDR ANNEX XI part A or Part B EU Medical Device Regulations 0
N OBL under the MDR. Which are the obligations of the OEM? EU Medical Device Regulations 0
U Distributor obligations under MDR EU Medical Device Regulations 5
Raisin picker CECP, consultation according to Art 54 MDR EU Medical Device Regulations 1

Similar threads

Top Bottom