MDR Article 86 Periodic safety update report reads:

1.Manufacturers of class IIa, class IIb and class III devices shall prepare a periodic safety update report (‘PSUR’) for each device and where relevant for each category or group of devices summarising the results and conclusions of the analyses of the post-market surveillance data gathered as a result of the post-market surveillance plan referred to in Article 84 together with a rationale and description of any preventive and corrective actions taken. ...


Two questions:
1. PSUR is a well known requirement in the pharma world (see attached template from EMA (virus scan before use )). Is it now required for all medical devices in class IIa or higher even if they do not administer medicinal products?
2. Does anyone already have a decent PSUR template and is willing to share it?

Many thanks to the loyal readers (and am I glad that Elsmar is back)
 

Attachments

Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Well, I was hoping someone with expertise would reply but maybe this will help move the discussion.

My understanding is that the PSUR is new to the MDR and it will be required for Class IIa, IIb, and III.

As far as the content of the report goes, just a little further down in Article 86 says:
(a) the conclusions of the benefit-risk determination;
(b) the main findings of the PMCF; and
(c) the volume of sales of the device and an estimate evaluation of the size and other characteristics of the population using the device and, where practicable, the usage frequency of the device.

Then running the thread, the PMCF is addressed in MEDDEV 2.12/2 rev 2. I guess that cover a pretty wide range of things so maybe a template would be difficult?

Hopefully someone has cracked this nut and can share / enlighten. (I'm glad you asked the question!).
 
#4
According my information the European Commission will release a template for the PSUR. Should have happened in December, but as so often they are delayed. Let's hope it will come soon. I personally am waiting with drafting a template.
 
#5
Thank you all for this discussion and input on PSUR. Here are my insights on PSUR being into the EU MDR field.

First of all, Yes, PSUR for medical devices is now a requirement for MDR compliance.

PSUR will be required for all devices classified as IIa, IIb or III. The scope is not limited to just for devices that are intended to administer and/or remove medicinal substances. This requirement will impact all devices (including custom devices) except class I. (As a side note, class I devices will still require PMS report and should be a part of Technical Documentation; Class I PMSR should be updated when necessary and should be made available to the competent authority when requested by a manufacturer).

An Annual update will be needed for PSURs falling under class IIb and III. A bi-annual update will be needed for PSURs falling under class IIa. (for class IIa, the PSUR update can be earlier than 2 years depending on the changes)

All PSURs will be required to be the part of Technical Documentation (Currently, termed as technical File). PSURs for custom devices will be required to be the part of its documentation (not formally termed as technical Documentation for Custom Devices).

PSURs for any class III or any implantable device will be required to be submitted to NB through EUDAMED for review and assessment. It is not clear to me whether a manufacturer or an NB will be required to submit this evaluated PSUR to Competent Authority but I am sure that it should be made available to the Competent Authority after NB's review. (Further reading Art. 86(3) gives a little more clarity that it mostly should be a manufacturer).
PSUR for any class IIa/IIb (non-implantable) devices will be required to be submitted to NB under conformity assessment review and to the competent authority, only when requested.

Hope this is helpful!
Thanks.
 

jiligan

Involved In Discussions
#6
I am currently working on MDR and came across your post since I am looking for a template for the PSUR also. Have you had any luck finding one that will work for you? Thanks!
 
#9
There is a template for the PSUR, which is nice because takes the guessing out of how this this thing should look. I cannot figure out how to post a thread here, need to figure that out. My question is given the scope of the new EU MDR, why would a PMS PLan be required for a device that is not even sold/marketed in the Union? EU is concerned about devices in their jurisdiction, versus a China only product, what good is a PM PLan for that if its not even distributed in EU?

:bonk:
 
Thread starter Similar threads Forum Replies Date
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
M EU MDR - Labilities when a device is discontinued from the Market EU Medical Device Regulations 1
S National registrations for MDR certified devices in absence of EUDAMED EU Medical Device Regulations 1
Z MDR Class Ir deadline (Reusable surgical instruments) EU Medical Device Regulations 27
B MDR equvalent device - can you use a medicine as an equivalent 'device' EU Medical Device Regulations 8
N Harmonized Standards for MDR Certification EU Medical Device Regulations 6
K Is UDI (Unique Device Identifier) symbol mandatory under EU MDR? EU Medical Device Regulations 1
K EU MDR Annex 1 Chapter III: Information in the Instructions for Use-23.4 (e) the performance characteristics of the device; EU Medical Device Regulations 1
R Requirements for MDR Article 22 Systems EU Medical Device Regulations 1
H EU MDR Economic Operator notifications? EU Medical Device Regulations 1
M MDR EU Distributor for our CE Mark product in the for Netherlands EU Medical Device Regulations 6
R Declaration of conformity as per EU MDR (2017/745) CE Marking (Conformité Européene) / CB Scheme 5
K Class III Kit Box - Are translations mandatory under EU MDR? EU Medical Device Regulations 5
B Herbal ingredients MDR EU Medical Device Regulations 0
S MDR GSPR Clause 17 - Software Requirements EU Medical Device Regulations 2
N Which EN ISO 17664 version compliance to EU MDR? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 3
D Hospital Decontamination and the MDR - Technical Files, UDI and PMS EU Medical Device Regulations 3
Kuldeep Singh Submission of MDR File - Partner Companies - Computed Tomography (CT) CE Marking (Conformité Européene) / CB Scheme 10
B How to classify a medical device based on MDR? EU Medical Device Regulations 3
A MDR GSPR Checklist wanted EU Medical Device Regulations 3
Rincewind MDR - GSPR - Annex I - Chapter II EU Medical Device Regulations 2
N EU MDR - Applicability Article 22 Systems and Procedure Packs CE Marking (Conformité Européene) / CB Scheme 4
B EU MDR Perscription product guidance EU Medical Device Regulations 1
cscalise Suggestions for MDR Auditing tools EU Medical Device Regulations 1
M Example of statement for procedure pack MDR Article 22? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
H MDD CM, MDR manufacturer. 2020. Is it ok? EU Medical Device Regulations 1
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 2
E UK MDR brexit out of sync with EU MDR deadline? Other Medical Device Regulations World-Wide 4
S MDR Delay - MDD design Change? (before new MDR DOA) EU Medical Device Regulations 8
S Class I MDR Article 10 – QMS EU Medical Device Regulations 1
dgrainger Medical Device News MDR postponed - link to Official Journal of the European Union (24-4-2020) EU Medical Device Regulations 2
K Supply Agreement Regarding MDR 2017/745 - Manufacturer-supplier cooperation Other Medical Device Related Standards 0
Ed Panek Medical Device News MDR Delayed 1 year (Posted 4/2020) EU Medical Device Regulations 9
dinaroxentool Former Class I device that is upscaled to IIa if the MDR is delayed EU Medical Device Regulations 2
B FDA-Medical Device Reporting (MDR )procedure compliant with 21CFR section 803 US Food and Drug Administration (FDA) 0
J PPE, the MDR and EU 2016/425 EU Medical Device Regulations 4
R An indication that the device is a medical device (MDR, Annex 23.2q) - applicable for accessories? EU Medical Device Regulations 5
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 6
dgrainger Informational Postponement of MDR - adopted by the Commission - April 2020 Medical Device and FDA Regulations and Standards News 3
R Applicability of new non-harmonized standards (MDD/MDR) EU Medical Device Regulations 8
F The EU Drafts Bill to Postpone MDR by a Year (3/2020) Medical Device and FDA Regulations and Standards News 2
T Commission proposal being drawn up for postponement for date of application of MDR (2017/745) EU Medical Device Regulations 7
A MDR Article 22 applicability - Legal manufacturer EU Medical Device Regulations 6
M Estimating the benefit-risk ration under MDR EU Medical Device Regulations 1
F Will the EU postpone 2017/745 MDR Implementation - COVID-19 Impact Medical Device and FDA Regulations and Standards News 11
Ajit Basrur Physical Address on Product Labels (EU MDR perspective) EU Medical Device Regulations 2
B Notified Bodies for Software (MDR) EU Medical Device Regulations 1
P MDR Rule 10 interpretation - Active Device EU Medical Device Regulations 3
JoCam Certified Body Audit of MDR requirements EU Medical Device Regulations 4
N Label Placement Requirement in MDD/MDR EU Medical Device Regulations 1

Similar threads

Top Bottom