Hello,

I am trying to understand MDR requirements for clinical investigations. But it is not clear to me whether a CRO located outside the Union can conduct clinical investigations in the Union (or outside the Union to comply with NB demands).

In Article 62, paragraph (2), there is a statement as follows:

2.Where the sponsor of a clinical investigation is not established in the Union, that sponsor shall ensure that a natural or legal person is established in the Union as its legal representative. Such legal representative shall be responsible for ensuring compliance with the sponsor's obligations pursuant to this Regulation, and shall be the addressee for all communications with the sponsor provided for in this Regulation. Any communication with that legal representative shall be deemed to be a communication with the sponsor. Member States may choose not to apply the first subparagraph to clinical investigations to be conducted solely on their territory, or on their territory and the territory of a third country, provided that they ensure that the sponsor establishes at least a contact person on their territory in respect of that clinical investigation who shall be the addressee for all communications with the sponsor provided for in this Regulation.


It says Sponsor, but does it also apply for CROs?

Thank you in advance :)
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Moderator
#2
Re: EU MDR CRO (Contract Research Organization) requirements for Clinical Investigati

Definition 49 in the MDR's Article 2 reads:

‘sponsor' means any individual, company, institution or organisation which takes responsibility for the initiation, for the management and setting up of the financing of the clinical investigation;
If the CRO meets this definition, it is subject to the requirements called out for Sponsors. If it doesn't, there should be another entity taking on the Sponsor's role, in which case those requirements won't apply to the CRO.
 
Last edited:
Thread starter Similar threads Forum Replies Date
B MDR PSUR versus Canadian Summary report EU Medical Device Regulations 3
C MDR 2017/744 Article 22 EU Medical Device Regulations 3
K Sales of MDD devices with expired CE that are pending MDR certification EU Medical Device Regulations 4
D New Device Post Market Requirement for MDR EU Medical Device Regulations 3
D EU MDR Derogation EU Medical Device Regulations 0
Doninina Risk management file according MDR or ISO 14971:P2019 ? EU Medical Device Regulations 2
M On-demand EU MDR Compliance toolkit with online training courses, templates & document management system Training - Internal, External, Online and Distance Learning 0
M EU MDR (GSPR Checklist) consolidated with TGA EP Checklist EU Medical Device Regulations 5
A PMSR or PSUR for legacy MDD-class I becoming MDR class-IIa EU Medical Device Regulations 7
H MDR clinical evaluation EU Medical Device Regulations 0
M Possible topic for Master thesis topic - MDR and QMS ISO 13485:2016 - Medical Device Quality Management Systems 4
D MDR technical file: does it need to contain duplicates of controlled documents EU Medical Device Regulations 2
D Does the MDR article 10 (14) needs to be documented word for word in a Procedure? EU Medical Device Regulations 1
N Adding New Products during MDR transition period Manufacturing and Related Processes 1
L Internal Audits during the MDR Transition Period EU Medical Device Regulations 5
Z How to keep up with changes on applicable technical standards (EU MDR)? EU Medical Device Regulations 6
S OEM- Virtual Manufacturer contract MDR Quality Management System (QMS) Manuals 0
J TGA CER & EU MDR CER Requirements for Australia Other Medical Device Regulations World-Wide 3
G MDR DRAFT DoC EU Medical Device Regulations 7
F MDR Distributor EU Medical Device Regulations 6
Y EU MDR Technical Documentation - Defined Review EU Medical Device Regulations 1
A Class 2b medical device - best pathway to obtain CE mark? (2022, MDR 2017/745) CE Marking (Conformité Européene) / CB Scheme 2
M Declaration of conformity MDR EU Medical Device Regulations 10
dgrainger Informational Manual on Borderline for MDR/IVDMDR EU Medical Device Regulations 6
G MDR Requirements for Samples EU Medical Device Regulations 5
C Applicability of classification rule 9 under MDR EU Medical Device Regulations 6
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Ed Panek MDR question - Meaning of Stability CE Marking (Conformité Européene) / CB Scheme 6
P Self Certified Class 1 MDR 2017/745 Design Change CE Marking (Conformité Européene) / CB Scheme 2
A EU MDR Annex II and Annex III CE Marking (Conformité Européene) / CB Scheme 6
N MDR requirments EU Medical Device Regulations 0
O MDD products under MDR certificate EU Medical Device Regulations 12
I Where to find MDR compliance device? EU Medical Device Regulations 6
J Conformity assessment route options under the UK MDR 2002 (2002 SI No. 618), as amended UK Medical Device Regulations 1
C Clinical Trial using Prototype devices (EU MDR & ISO 14155) EU Medical Device Regulations 4
M MDR Legal requirements EU Medical Device Regulations 6
Ed Panek MDR Audit Comment EU Medical Device Regulations 9
J Minor / Major findings with MDR EU Medical Device Regulations 4
R MDR Clinical Evaluation Plan EU Medical Device Regulations 6
R EU MDR Article 22 applicability of medical device with components. EU Medical Device Regulations 4
K Qualified MDR internal auditor? EU Medical Device Regulations 8
J Clinical Evaluation For Instruments - Class I, Class Ir, Class Im, and Class IIa - EU MDR EU Medical Device Regulations 4
G MDR Translation Requirements in Member States EU Medical Device Regulations 8
R General Obligations of Importers under EU MDR EU Medical Device Regulations 3
V MDR – Article 65 – Clinical investigations on minors EU Medical Device Regulations 6
I MDR Audit: NC per MDR Annex IX Administrative provisions 8. EU Medical Device Regulations 6
R Device with WiFi communication - EU MDR vs. EU RED EU Medical Device Regulations 2
L Clinical Evaluation MDR EU Medical Device Regulations 15
J Common Specifications per MDR EU Medical Device Regulations 2
L Looking for SOP Template for UDI under MDR EU Medical Device Regulations 4

Similar threads

Top Bottom