Evaluation of Uncertainty of In Vitro Diagnostic Kit Calibrators

K

kuertensun

#1
Dear all,

I am a guy in the in vitro diagnostic medical device industry. As ISO 17511 requires, an uncertainty should accompany any calibrator enclosed in the in vitro diagnostic kit. In my company's product, the value of the calibrator is obtained from 4 parameter data reduction method and is calibrated by our national standard calibrator. In the evaluation of the uncertainty, I get some difficulties in how to calculate the uncertainties of the 4 parameters in the function.

Please anyone can help with this problem?

thanks,
 
Elsmar Forum Sponsor

Stijloor

Staff member
Super Moderator
#2
Dear all,

I am a guy in the in vitro diagnostic medical device industry. As ISO 17511 requires, an uncertainty should accompany any calibrator enclosed in the in vitro diagnostic kit. In my company's product, the value of the calibrator is obtained from 4 parameter data reduction method and is calibrated by our national standard calibrator. In the evaluation of the uncertainty, I get some difficulties in how to calculate the uncertainties of the 4 parameters in the function.

Please anyone can help with this problem?

thanks,
Can someone with this expertise help?

Thank you!

Stijloor.
 
G

George Weiss

#3
It appears there is no one accepted way to determine the uncertainty value for your calibrator/material, in many cases.
This is a problem in your product field, which is under review for future guidance.
Your uncertainty process develops into the various GUM guidance documents.
You describe a 4 point, or a 4 parameter evaluation of a material/calibrator.
It suggests that you are talking about a biological material/calibrator?
There are many IVD tools and materials/calibrators for various tests.
Reviewing competitor values will give you a target, which is worth knowing.
You have a National Standard, which it was compared to, with uncertainties also.
You have given only limited information. I offer an uncertainty guide, which might help:
The Australian Health department has a Uncertainty guide for your type of product @
http://www.health.gov.au/internet/main/publishing.nsf/Content/86A3CE312C612377CA257283007BC92D/$File/dhaeou.pdf
I found some BIPM uncertainty budgets, which might head you in the right direction @
http://www.bipm.org/utils/common/pdf/final_reports/QM/K6/CCQM-K6.pdf
The uncertainty budget in pharmaceutical industry @
http://www.pharmchem.tu-bs.de/forschung/waetzig/dokumente/DPhG_2005_Heydorn.pdf
NIST Special Publication 260-162 @
http://ts.nist.gov/MeasurementServices/ReferenceMaterials/upload/SP260-162.pdf
A common comment is that the question poster offer up more information. Sometimes within these added points is enough to plot a course for you. Some uncertainty wiz-guys want the numbers to work with.
SO, what kind of calibrator.material is it? A complex ASR?
I am not the expert, just passing on some general help.
 
Last edited by a moderator:

BradM

Staff member
Admin
#4
Dear all,

I am a guy in the in vitro diagnostic medical device industry. As ISO 17511 requires, an uncertainty should accompany any calibrator enclosed in the in vitro diagnostic kit. In my company's product, the value of the calibrator is obtained from 4 parameter data reduction method and is calibrated by our national standard calibrator. In the evaluation of the uncertainty, I get some difficulties in how to calculate the uncertainties of the 4 parameters in the function.

Please anyone can help with this problem?

thanks,
OK, first are the different components of the total uncertainty:
1. The uncertainty attributed by the national standard calibrator
2. Uncertainty associated with the error of someone using the calibrator enclosed with the product. I would think there could be some potential error there. Given human nature as it is:D, there is no task so simple that people can't mess up.:) So I would think that would be a contributor.
3. Are there any more contributors? Shelf life? Environmental factors?

To your question.... I realize you probably need to be vague about these four parameters, as you certainly don't want to divulge anything too confidential and such. However.... it's awfully generic, and difficult to get a handle on what it is.

With this data reduction method.... is it an identical mathematical calculation, or is there some statistical inferences being made? Out of say 100 trials, what kind of error would be generated using the four parameters simply given the nature of the parameters?

Anyway, I'm pretty sure there is nothing helpful there.:D But possibly if you could add a little more explanation of the parameters, that might help.:)
 
K

kuertensun

#5
OK, first are the different components of the total uncertainty:
1. The uncertainty attributed by the national standard calibrator
2. Uncertainty associated with the error of someone using the calibrator enclosed with the product. I would think there could be some potential error there. Given human nature as it is:D, there is no task so simple that people can't mess up.:) So I would think that would be a contributor.
3. Are there any more contributors? Shelf life? Environmental factors?

To your question.... I realize you probably need to be vague about these four parameters, as you certainly don't want to divulge anything too confidential and such. However.... it's awfully generic, and difficult to get a handle on what it is.

With this data reduction method.... is it an identical mathematical calculation, or is there some statistical inferences being made? Out of say 100 trials, what kind of error would be generated using the four parameters simply given the nature of the parameters?

Anyway, I'm pretty sure there is nothing helpful there.:D But possibly if you could add a little more explanation of the parameters, that might help.:)
Dear BradM,

Thanks for your reply,

Actually, 4 parameter data reduction method is a commonly used regression method in the in vitro diagnostic industry, people in this field should know this model. There are 4 parameters in the function used to calculate concentration from the output signal, say, an absorbance. I read GUM, and it includes an example similar to my case, however, in the example, only 2 parameters need to be determined, and the example says the uncertainty of these 2 parameters are evaluated by well established equations, so I am thinking if there is also a well established theory for my case.

Hope this makes the question clearer.

Thanks.
 
Last edited by a moderator:

BradM

Staff member
Admin
#6
Dear BradM,

Thanks for your reply,

Actually, 4 parameter data reduction method is a commonly used regression method in the in vitro diagnostic industry, people in this field should know this model. There are 4 parameters in the function used to calculate concentration from the output signal, say, an absorbance. I read GUM, and it includes an example similar to my case, however, in the example, only 2 parameters need to be determined, and the example says the uncertainty of these 2 parameters are evaluated by well established equations, so I am thinking if there is also a well established theory for my case.

Hope this makes the question clearer.

Thanks.
:agree1: And I'm glad you mentioned

people in this field should know this model
I should have mentioned earlier that I am not in this industry, and am totally unfamiliar with the formula you mentioned. :) Of course, I think you figured that out fairly quickly.:tg::lol:

I am just trying to pitch in some on the calculation of the uncertainty. Now if you are performing regression, would the R Squared not be useful in estimating uncertainty? Maybe the lower R Squared would contribute more uncertainty than a higher R Squared?

I would look for some information at the N.I.S.T. site for you, if I understand better what I was talking about.:D

Here is a list of the publications that might help:

http://www.nist.gov/srm/publications.cfm


Here is their contact list:

http://www.nist.gov/public_affairs/contact.cfm

Maybe one of them can guide you to the perfect document for guidance, or, might be able to help you in an e-mail or something.
 
G

George Weiss

#8
Hello again,
Your comment, "There are 4 parameters in the function used to calculate concentration from the output signal, say, an absorbance." Was helpful.
you might be speaking of calibration points "A", "B", "C", "D", "E" of material/calibrator dilutions.
Calibration point "A" is the reference without material.
You might also be talking about standard reporting outputs of: "four parameters", or polygonal regression curve in linear or logarithmic scale.
There is a possibility that your 4-parameter term is one of the common IVD analyzer outputs, Curve-fitting options: linear, 4-P, 2-P (logit/log), cubic, quadratic, cubic spline, point-to-point. In the end you have a value or values. These values should compare with the national standard calibrator. The significance of these values, is their variance from the national standard calibrator. This variance/error is a combination of some of the uncertainties, which you are trying to describe.
The significant absolute uncertainty values are:
  1. national standard
  2. process & bias errors (pipette volume, time, temperature, sample-number, and others)
all of these combined via GUM RSS method will give you your IVD kit calibrator uncertainty.

.
Absorbency calibration @
http://www.cisbio-ivd.com/documents/chromoa-package-insert.pdf
A nice IVD calibration uncertainty method description @
http://www.edma-ivd.be/fileadmin/upl_documents/Position_Papers/Estimation_of_uncertainty.pdf
.
The extremely wide variety of in vitro diagnostic devices currently available made it apparent that a single experimental design would not be appropriate for all devices.
.
For complex-calibrators, there is a defined assigned uncertainty process @
http://www.clinchem.org/cgi/reprint/53/4/735.pdf
.
I am just rewording the thread comments to assist in the next poster developing an answer to the question:
"what is the formula for determining the uncertainty of my IVD kit calibrator/reference-material from my 4-parameter data"
 
Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
A Applicability of Photobiological Safety Evaluation for LED used in medical devices Reliability Analysis - Predictions, Testing and Standards 1
P GSPRs / Clinical Evaluation EU Medical Device Regulations 3
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
M NICE Medical Technology Evaluation Programme - Recommendations Service Industry Specific Topics 0
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 6
S Regular updates of clinical evaluation report EU Medical Device Regulations 6
K Surface finish (Evaluation Length) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
M Clinical Evaluation Benchmark vs. Equivalent EU Medical Device Regulations 2
M IVDR and Performance Evaluation Plan CE Marking (Conformité Européene) / CB Scheme 2
S User evaluation for self monitoring blood glucose test systems US Food and Drug Administration (FDA) 4
S Australia TGA Clinical Evaluation Report (CER) Other Medical Device Regulations World-Wide 0
S API Spec. Q1 clause 5.6.1.2 On site evaluation Oil and Gas Industry Standards and Regulations 10
B Biological evaluation plan and report Other Medical Device Related Standards 5
F Biocompatibility evaluation for Hardware/Interface Components? Other Medical Device Related Standards 12
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
T Clinical evaluation of a new medical device EU Medical Device Regulations 0
dgrainger Informational MDCG 2020:13 - Clinical evaluation assessment report template EU Medical Device Regulations 0
S A clinical performance evaluation study with an IVD product as Investagional Use product - Clinical Monitor requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B AS9100 8.4.1 Supplier Selection/Evaluation criteria and reevaluations AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 13
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 8
M Clinical Evaluation Plan vs. PMCF Plan EU Medical Device Regulations 24
K Biocompatibility evaluation of gas pathways Medical Device and FDA Regulations and Standards News 5
T HF testing / Summative evaluation for MDDS class I necessary? Human Factors and Ergonomics in Engineering 2
N Evaluation of service providers Supplier Quality Assurance and other Supplier Issues 2
M Can someone share a scrubbed version of Clinical Evaluation Plan (CEP) EU Medical Device Regulations 8
S EU MDR Annex XIV - Clinical Evaluation Plan - What do these methods mean? EU Medical Device Regulations 16
C Material from outside CER evaluation period CE Marking (Conformité Européene) / CB Scheme 8
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Clinical Evaluation - Is this an ISO 13485:2016 requirement? ISO 13485:2016 - Medical Device Quality Management Systems 4
P SOUP anomaly evaluation for MMA (Mobile Medical Application) IEC 62304 clause 7.1.3 IEC 62304 - Medical Device Software Life Cycle Processes 6
M Informational How to perform a clinical evaluation of medical devices – Part 2 – Level of clinical evidence and what sufficient clinical evidence means Medical Device and FDA Regulations and Standards News 9
P IEC 62304 - evaluation of integration and system testing IEC 62304 - Medical Device Software Life Cycle Processes 4
R Supplier evaluation and business needs in the context of ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 2
U Changes to Internal Processes and Risk Evaluation - Mitigations Risk Management Principles and Generic Guidelines 10
M Informational Work in progress at the FDA for biological evaluation – Color Hazard and RISk calculator (CHRIS) Medical Device and FDA Regulations and Standards News 0
M Informational How to perform a clinical evaluation of medical devices – Part 1 – Overview and sample of activities Medical Device and FDA Regulations and Standards News 0
T EU MDR Article 61- Clinical Evaluation EU Medical Device Regulations 2
pashah Looking for Clinical Evaluation SOP acc. MEDDEV and EU MDR Other Medical Device Related Standards 1
K Supplier re-evaluation (API Q1) Manufacturing and Related Processes 8
B Evaluation of Basic Safety during EMC Immunity or Climate Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S ISO 10993 Biocompatibility Evaluation - Electronic thermometer Other Medical Device Related Standards 3
D Clinical Evaluation Report - Consultant Recommendations Consultants and Consulting 9
qualprod ISO 9001 Clause 9.1 - Monitoring measurement analysis and evaluation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
M CER (Clinical Evaluation Reports) updates - Product Codes EU Medical Device Regulations 2
A Medical device CER (clinical evaluation report) training/seminar services EU Medical Device Regulations 2
M Informational US – National Evaluation System for Health Technology Coordinating Center (NESTcc) Solicits Public Comments for Data Quality and Methods Frameworks Medical Device and FDA Regulations and Standards News 0
M Informational IMDRF – Proposed update to Clinical Evaluation documents Medical Device and FDA Regulations and Standards News 0
M Informational MDCG 2019-3 Interpretation of Article 54(2)b – Pre- market clinical evaluation consultation procedure with the involvement of expert panels Medical Device and FDA Regulations and Standards News 0
TheMightyWife Clinical Evaluation MEDDEV 2.7.1 Rev 4 demonstration of equivalence EU Medical Device Regulations 6