Extending the Calibration Frequency of a Measurement Device
C
Corsair
#1
Gentleman,
Increasing the frequency would be going from 6 months to 3months, an increase in how many times it is done. Whereas decreasing the frequency is going from 6months to 12months, a decrease in the number of times in a period.
I need your toughs on the above statement.Any body agree to that?
:thanx:
Elsmar Forum Sponsor
J
Jason PCSwitches
#2
Re: Calibration Frecuency Extension.
Depending on the type of equipment and its usage (criticality & sometimes contract requirements, though rarely the latter) as long as you have established reliability at the 6 month interval (history) you can increase.
Decreasing is never a problem, cost should be considered though.
C
Corsair
#3
May be I was not clear enough, on this one.
But to me, when we increase the frequency of an instrument is to "prolong"
the interval of calibration and to decrese is to shortened the interval of cal-.
Is this correct?
D
dkusleika
#5
Agreed. Increasing the cycle or the term would mean calibrating less often. Increasing the frequency would be calibrating more often.
Sidney Vianna
Post Responsibly
Staff member
Admin
#6
Is this correct?
No, it is not. Increasing the frequency of calibration means shortening the interval between calibration cycles.
E.g., calibrating a gage every 3 months is twice as frequent as calibrating it every six months. The longer the interval between calibrations, the lower the frequency.
Marc
Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#7
Sidney is right.
Depending on the type of equipment and its usage (criticality & sometimes contract requirements, though rarely the latter) as long as you have established reliability at the 6 month interval (history) you can increase decrease the calibration frequency.
Decreasing Increasing the calibration frequency is never a problem, cost should be considered though.
Fixed that for you.
C
Corsair
#8
The words that we usually use in this business (frequency & interval) have not the same meaning then.
Frequency, meaning rate of occurrence and interval meaning cycle (the amount of time between a specified events).
Correct?
Hershal
Metrologist-Auditor
Staff member
Super Moderator
#9
Gentleman,
:thanx:
The statement is correct.
Hershal
Metrologist-Auditor
Staff member
Super Moderator
#10
Correct?
The definitions are also correct.
More calibrations in a given time frame is an increase in frequency and a decrease in cycle.
There is an inverse relationship.
Thread starter Similar threads Forum Replies Date
M Extending Surface Plate Calibration Frequency General Measurement Device and Calibration Topics 3
M Illuminance Meter - Extending Manufacturer's Recommended Calibration Frequency Calibration Frequency (Interval) 3
M Procedure or Specification For Extending Gages Calibration Calibration Frequency (Interval) 2
B Extending a Shielded Cable into an AP IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
E Extending current ISO 9001 registration scope to cover SW development Software Quality Assurance 2
G Validity of Extending Temperature Cycling Various Other Specifications, Standards, and related Requirements 3
J Extending QMS Scope to External Company Site (Outsourcing) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
J Extending the Scope during Transition Audit to AS9100 Rev. C AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
hogheavenfarm Extending control of measuring equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
B Calculation for extending control limits Statistical Analysis Tools, Techniques and SPC 7
RoxaneB ISO 9001:1994 Certificates of Compliance - Are any registrars extending these? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Similar threads
Top Bottom