External Calibration Certificate Review
MEDQA
Hello -
I would like to have a simple cover sheet that lists items required on certs from our calibration provider's. What do we need on our external certs in order to be ISO 9000 / 9001 compliant? Please let me know if this is overkill. My goal is to make this simple. and consistent. - thank you
This is what I had in mind.
  • Supplier Name
  • Calibration Tech. Identified
  • Equipment Used to calibrate identified
  • Calibration Dates of Equipment Used present /correct
  • As found is in spec?
    • Yes = no action
    • No = obtain supervisor approval
  • NIST statement present.
  • Serial # of equip. being calibrated present / correct
  • Reviewer Initials and date.
  1. Is NIST statement necessary?
  2. Does NIST statement eliminate the need to provide Calibration Equipment's Calibration dates or any of the other items listed?
ISO/IEC 17025 gives some excellent details of the content of a calibration report. You might want to check into that as a guide.
MEDQA
I was soooo excited to see a reply. - Thank you. I searched for the document to find that I would have to buy it - UGH. I'm at home and thought I could get ahead before Monday.
Any free input is appreciated b/c I might not be permitted to buy the document once I get to work. :(
Big Jim
Hello -
  • Supplier Name
    • Yes = no action
1) Yes
2) No
somashekar
With its calibration status traceable to the national standards.
Claimed accuracy Vs found accuracy of the DUC
The measurement uncertainity calculated
If the calibration provider is ISO/IEC 17025 certified., then the accreditaton mark on the certificate.
The Valid upto date of the certificate.
Hershal
Metrologist-Auditor
MEDQA, a few things you need to make sure it is traceable.
It needs to have the documented path of traceability. That is, what is the specific link to the next level up? BTW, the so-called NIST Numbers are NOT considered valid for traceability, unless you can link them to a specific calibration of YOUR CALIBRATION PROVIDERS' standards at a specific point, and that must agree with the stated recall documented on the cal cert listing the standards, as provided to you. If it is of THEIR calibration providers' standards, then you have a broken link.
Next is uncertainty. You may not need the uncertainty, that is a different question. But the uncertainty of the delivered calibration needs to be stated.
This is consistent with what Metrologists know as VIM, which is the dictionary for calibration.
Ideally the lab is accredited to ISO/IEC 17025:2005 and also provides the logo and certificate number.
The rest of the list you put forward is acceptable.
Hope this helps.
MEDQA
Thanks to all.
I don't expect that any of our providers aren't doing what they say they are doing. I just want internal techs to read the certs and verify their content is accurate. I don't want to ask for more than we need. If any items from my list should be "If Present" please advise.
Ajit Basrur
Hello -
  • Supplier Name
    • Yes = no action
I would like you to refer the Agilent's Calibration Certificate Explained and an attachment from NIST (refer page 2)
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
C External Laboratories and OEM Calibration IATF 16949 - Automotive Quality Systems Standard 0
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
N IATF 16949 7.1.5.3.2 External Laboratory - Calibration of Nikon machine in factory IATF 16949 - Automotive Quality Systems Standard 24
B IATF 16949 7.1.5.3.2 FAQ #7 Audit Finding - External Calibration Laboratory IATF 16949 - Automotive Quality Systems Standard 7
sswaim IATF 16949 7.1.5.3.2 External Calibration Laboratory Requirelents IATF 16949 - Automotive Quality Systems Standard 15
smryan Weight Calibration Standards - External Calibration Frequency General Measurement Device and Calibration Topics 11
S What is the ISO 17025 Certified Laboratory Requirement for External Calibration? ISO 17025 related Discussions 8
G ISO17025 requirement in TS16949 for External Calibration Laboratories ISO 17025 related Discussions 4
N Specifying Measurement Standards on an External Calibration PO? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
B Audit of equipment after certified external calibration... ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Evidence of Contract Review for External Calibration Laboratory Contract Review Process 7
A Need to Audit an External Calibration Laboratory? - Customer Audit Finding Other Medical Device Regulations World-Wide 5
N ISO 17025 External Calibration Laboratory Requirements within ISO/TS 16949 ISO 17025 related Discussions 28
R Acceptance Criteria for Calibration by an External Calibration Laboratory General Measurement Device and Calibration Topics 9
F Process Monitoring Gage Calibration Requirements - 7.6.3.2 External laboratory IATF 16949 - Automotive Quality Systems Standard 1
V ESD testers and providing calibration service - 7.6.3.2 External Laboratory IATF 16949 - Automotive Quality Systems Standard 3
Q Can a waivers for external calibration labs be implemented via the PPAP process? IATF 16949 - Automotive Quality Systems Standard 3
Z Ring Gage Calibration Sizes Shrunk with New External Vendor (supplier) General Measurement Device and Calibration Topics 16
R TS16949 External Calibration Laboratory Requirements - Thermocouples IATF 16949 - Automotive Quality Systems Standard 7
M TS Clause 7.6.3.2 - Can Non-Accreditated External Laboratory Carry out Calibration? IATF 16949 - Automotive Quality Systems Standard 14
S Calibration System was In-House now Outside (External Service Provider) General Measurement Device and Calibration Topics 14
F Calibration Tags - External company applied their tags - Do I need to apply our tags Document Control Systems, Procedures, Forms and Templates 6
C External calibration provider does not issue uncertainty of measurement Measurement Uncertainty (MU) 2
P About 7.6.3.2 External Laboratory - Calibration by Manufacturer General Measurement Device and Calibration Topics 2
P TS 16949 Clause 7.6.3.2 - Use of External Calibration Laboratories General Measurement Device and Calibration Topics 7
R 17025 Calibration Requirements - External calibration service cannot calibrate ISO 17025 related Discussions 4
T IM&TE - Outside calibration companies - 7.6.3.2 External Laboratory IATF 16949 - Automotive Quality Systems Standard 2
I Defined External Calibration Acceptance Criteria General Measurement Device and Calibration Topics 10
K External calibration laboratories - IS0 17025 Equivalents? ISO 17025 related Discussions 19
G What is your expenses for external calibration per year? General Measurement Device and Calibration Topics 7
J External third party audits Registrars and Notified Bodies 1
J Requirement for Retention of Records of Withdrawn Documents of External Origin Document Control Systems, Procedures, Forms and Templates 3
L IATF external audit virtual (remote) IATF 16949 - Automotive Quality Systems Standard 13
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
R External Audit and Certificate prorogation due to the pandemic General Auditing Discussions 10
N Audit non-compliance API Q1 - Use of External Documents 4.4.4 in Product Realization Oil and Gas Industry Standards and Regulations 8
J 510(k) for a control kit for an external IVD test kit 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R Is it required to have an SOP for external audits? Medical Device and FDA Regulations and Standards News 7
Pmarszal External Standards and Regulations Management Process Document Control Systems, Procedures, Forms and Templates 10
I What kind of wine best complements the Friday that you close out your external audit findings? Opinions are welcome. Coffee Break and Water Cooler Discussions 12
B How to apply external voltage to SIP/SOP IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Qualifying commercial off the shelf (COTS) external suppliers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
B Internal and external auditor competency to CSR's IATF 16949 - Automotive Quality Systems Standard 20
N IATF 16949:2016 7.1.5.3.2 External Laboratory - How to approve the Testing Laboratory without accreditation scope IATF 16949 - Automotive Quality Systems Standard 2
A OHSAS 18001 external auditor finding personal interpretation? Occupational Health & Safety Management Standards 5
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Reduction of software class based on multiple external risk controls IEC 62304 - Medical Device Software Life Cycle Processes 5
C Identifying and Controlling External Documents Document Control Systems, Procedures, Forms and Templates 3