External calibration laboratories - IS0 17025 Equivalents?

K

Knowledge seeker

#1
IS0 17025 equivalent

My company is going for QS9000 certification,
Can you guy help me by providing some answer to the following question?

1)QS9000 required external calibration laboratories to be Iso 17025 accredited or equivalent. May i know what standard used in japan, UK and USA (eg: NIST/ ANSI)are equivalent to Iso 17025 and is accepted by QS9000 standard?

2)i had this calibration supplier who claim that for calibration to be perform onsite, ISO 9001/ 9002 will be approriate. ISO 17025 is meant for their laboratory (This statement mean that QS9000 does not required external calibration laboratories who perform onsite calibration to be ISO 17025 accredited, Which i don't agree with..) Comments???:confused:

3) Does QS9000 required external calibration lab for equipments such as noise Dosimeter, Gas detector, to be ISO 17025 accredited? (They have no impact on product quality)

4) Is it necessary to place a status label (Reference only) on every instruments which are part of the machine(such as pressure gauge, Voltage/current indicator)? They are part of the machine to produce the product, not product inspection equipment.

That's all for the time being. Thanks
 
Elsmar Forum Sponsor
D

D.Scott

#2
As far as I can help - IMHO

In the US - and under QS-9000 - If you use an outside calibration lab they must be:
Accredited to ISO 17025 (the standard says "Guide 25" -17025 is the new version of that)
or
Approved via the 2pCP program (AIAG approved auditors using ISO Guide 25 as the base)
or
If there is no accredited lab, the OEM can perform the calibration.

The second question is confusing - An in-house lab doing it's own testing for their process does not have to be accredited to 17025. If they are testing their product and sending you the results (certs) their QS-9000 is good enough. If they sent something to another lab for testing, the lab they sent it to would have to be accredited to 17025. If the in-house lab calibrated their own equipment they do not have to be accredited. They must be able to trace the calibration to a national standard. This would mean that their "master" would have to be calibrated and a certificate issued by an accredited calibration lab.

Third - any calibration done by an outside lab must be done by an accredited lab. The type of equipment is not important (it is possible for a non-accredited OEM to calibrate their own equipment IF an accredited lab DOES NOT EXIST.

Fourth - all measurement devices listed on your control plan or used to determine a state of quality should be in your calibration system and labeled. You determine what goes in the system and therefore determine what needs to be calibrated. If the items you list affect the outcome of the product and someone has to rely on the gage being accurate, then I would say calibrate them. If you decide they don't need calibration then a label shows you have addressed it.

Hope this gets you started.

Dave
 
R

Ryan Wilde

#3
Re: IS0 17025 equivalent

Originally posted by Knowledge seeker
1)QS9000 required external calibration laboratories to be Iso 17025 accredited or equivalent. May i know what standard used in japan, UK and USA (eg: NIST/ ANSI)are equivalent to Iso 17025 and is accepted by QS9000 standard?
ISO 17025 is the International Standard, followed by each of these nations. There is no equivalent recognized by QS9000.

Japan Accreditation Organizations that you may wish to contact are:

Japan Quality Assurance Organization-(JQA)
Japan Accreditation Board-(JAB)
Japan Calibration Service System-(JCSS) [Sorry, I have no link for them]

UK Accreditation Organizations that you may wish to contact are:

United Kingdom Accreditation Service-(UKAS)
National Measurement Partnership-(NMP)
National Physics Laboratory-(NPL)

USA Accreditation Organizations that you may wish to contact are:

National Voluntary Laboratory Accreditation Program-(NVLAP)
American Association for Laboratory Accreditation-(A2LA)
Laboratory Accreditation Bureau-(LAB)

And since your profile says that you are from Singapore, your Accreditation Organizations are:
Dead link: http://www.psb.gov.sg - Productivity and Standards Board-(PSB)
Dead link: http://www.sac-sci.org.sg - Singapore Accreditation Council
Dead Link: http://www.np.edu.sg/~tankc/qrc/sislab.htm - Singapore Laboratory Accreditation Scheme-(SINGLAS)

Any of these organizations should be able to put you in contact with an accredited laboratory to do your outside calibrations, therefore meeting the QS9000 requirement.
2)i had this calibration supplier who claim that for calibration to be perform onsite, ISO 9001/ 9002 will be approriate. ISO 17025 is meant for their laboratory (This statement mean that QS9000 does not required external calibration laboratories who perform onsite calibration to be ISO 17025 accredited, Which i don't agree with..) Comments???:confused:
Your calibration supplier is wrong. In the USA, adding site calibration capability to any portion of your scope actually must be specified on the scope. Any function being calibrated by an outside vendor must be accredited. I am not familiar with the methodology employed by other nations, but I do know that the 17025 Accreditation is required, whether on-site or in the vendor laboratory.
3) Does QS9000 required external calibration lab for equipments such as noise Dosimeter, Gas detector, to be ISO 17025 accredited? (They have no impact on product quality)
It is my understanding that equipment that does impact product quality is not covered under the accredited calibration clause.
4) Is it necessary to place a status label (Reference only) on every instruments which are part of the machine(such as pressure gauge, Voltage/current indicator)? They are part of the machine to produce the product, not product inspection equipment.
Are any of the indications provided by the machine instruments used in the final inspection report? If not, then they do not have to be calibrated, although I personally would have some system of calibration set up for them. As far as the label goes, I would say that it would vary from auditor to auditor, as long as there is documentation in place that states that machine instrumentation is understood to be Reference Only.

Ryan
 
K

Knowledge seeker

#4
Hi D Scott, i am sorry to confuse by my question What i really mean is that if a commercial Calibration lab is ISO 17025 accredited, its scope of accrediation should have includ onsite calibration perform for its customer..(somehow they have to provide me with calibration uncertainty at the end of the day)

Ryan, i am perfectly agree with you...
i think it is perfectly alright to have a documentation in place which state that machine instrumentation whose indications is not used in the final inspection report is understood to be Reference Only. That could save the hassle of putting those " reference" label.

Thanks alot...:D
 
K

Ken K

#5
An in-house lab doing it's own testing for their process does not have to be accredited to 17025. If they are testing their product and sending you the results (certs) their QS-9000 is good enough.
Dave, until yesterday I would have agreed with you but...

After rejection of a certification packet to GM and a heated discussion with those involved, that is no longer the case. As far as GM is concerned, " Registration to QS9000 is not a substitute for lab accreditation." As of yesterday, unless our internal lab gets accredited to ISO 17025, they will not accept test data from us.

Go figure! :mad: :mad: :mad: :mad:
 
R

Ryan Wilde

#6
Originally posted by Ken K


Dave, until yesterday I would have agreed with you but...

After rejection of a certification packet to GM and a heated discussion with those involved, that is no longer the case. As far as GM is concerned, " Registration to QS9000 is not a substitute for lab accreditation." As of yesterday, unless our internal lab gets accredited to ISO 17025, they will not accept test data from us.

Go figure!
Two things:

A) That really sucks! Accreditation is no walk in the park, and if they require manufacturers to have their internal labs accredited, a Cavalier would cost in the region of $50k. OUCH!

B) Did they show you, in writing, the requirement for manufacturer's internal labs to be accredited? Either in QS9000 (it's not there) or an honest to goodness GM document? Could this be a case of an overzealous assessor?

If this is genuinely the case, we'll all be buying Hondas and Toyotas (I already do, but that's not the point.)!

Ryan
 
K

Knowledge seeker

#7
Well, As far as i know it is not a QS9000 requirement for internal lab to be accredited to ISO 17025. The Standard state that internal lab have to comply with ISO 17025 while external commercial lab used by the company have to eaccredited to the standard. If accredited commercial lab is not available, calibration can be done by OEM.
The only exception is when the testing / calibration lab is being appointed by your customer.

I think for internal lab to be accredited to ISO 17025 is more of a customer's Requirement than the QS 9000 requirement.
 
K

Ken K

#8
This could get quite interesting. It seems this is being driven by the Interior Systems Group. It all started with one of GM's recently updated specifications which states;

Testing is to be performed by a laboratory accredited
or audited, by a third party organization recognized by
GM, to ISO17025/IEC Guide 25 or GM GP-10. Include a
Scope of Accreditation or a Scope of Activity with the test
data. Registration to QS9000 is not a substitute for lab
accreditation.

According to the person I talked with, GP-10 was also not a substitute for accreditation. But, we have a signed document from GM ( our GP-10 auditor) dated August 1997 that states that since we have acheived QS9000 compliance (1996), GP-10 would no longer be required. And our GP-10 accreditation was allowed to expired in December 1997.
She didn't want anything to do with the letter and said if we are not ISO17025 accredited, she would not recognize any test data we supply.

So, in my opinion, we have danced with QS for six years for nothing. 10 audits with absolutely nothing to show for it. When we asked our auditor about this last year, we were told it did not pertain to internal labs and we have nothing to worry about. I wonder who's the one that's confused.
 
D

D.Scott

#9
Here we go again ....

Just when I thought it was safe to go back in the water -

Ken, could you give me a little more info on this? - Was this a test you were doing internally on your oun product? Was it part of an original material approval? Was any supporting portion of the test done on the outside?

This is directly contrary to the IASG interpretation and comes as a complete shock. I have not heard it from other sources yet.

If anyone gets any kind of written information on this please post it for us. I would love to see the whole thing refuted but confirmed will do just as well.

In the meantime, I will present a request for clarification to IASG and see if I can get an answer.

Thanks for the "heads-up" Ken.

Dave
 
K

Ken K

#10
Ken, could you give me a little more info on this? - Was this a test you were doing internally on your oun product? Was it part of an original material approval? Was any supporting portion of the test done on the outside?
We developed a new material to replace existing material for which we submitted lab packets to GM, Ford and Chrysler. The material was approved by Ford and Chrysler with no problems. GM balked. Don't know why yet, but we are attempting to get this straightened out. All we were really told is this is GM policy which was likely never conveyed to suppliers. I know we never received anything from GM concerning this.

All testing was done in-house. It consisted of a complete material certification. The test samples were included in the packets along with our Lab Scope, QS Certificate and calibration reports on the equipment used for testing.

Will keep you posted on any further developments.
 
Thread starter Similar threads Forum Replies Date
C External Laboratories and OEM Calibration IATF 16949 - Automotive Quality Systems Standard 0
G ISO17025 requirement in TS16949 for External Calibration Laboratories ISO 17025 related Discussions 4
P TS 16949 Clause 7.6.3.2 - Use of External Calibration Laboratories General Measurement Device and Calibration Topics 7
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
N IATF 16949 7.1.5.3.2 External Laboratory - Calibration of Nikon machine in factory IATF 16949 - Automotive Quality Systems Standard 24
B IATF 16949 7.1.5.3.2 FAQ #7 Audit Finding - External Calibration Laboratory IATF 16949 - Automotive Quality Systems Standard 7
sswaim IATF 16949 7.1.5.3.2 External Calibration Laboratory Requirelents IATF 16949 - Automotive Quality Systems Standard 15
smryan Weight Calibration Standards - External Calibration Frequency General Measurement Device and Calibration Topics 11
Sean Kelley What is the ISO 17025 Certified Laboratory Requirement for External Calibration? ISO 17025 related Discussions 8
N Specifying Measurement Standards on an External Calibration PO? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
B Audit of equipment after certified external calibration... ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Evidence of Contract Review for External Calibration Laboratory Contract Review Process 7
A Need to Audit an External Calibration Laboratory? - Customer Audit Finding Other Medical Device Regulations World-Wide 5
M External Calibration Certificate Review ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
N ISO 17025 External Calibration Laboratory Requirements within ISO/TS 16949 ISO 17025 related Discussions 28
R Acceptance Criteria for Calibration by an External Calibration Laboratory General Measurement Device and Calibration Topics 9
F Process Monitoring Gage Calibration Requirements - 7.6.3.2 External laboratory IATF 16949 - Automotive Quality Systems Standard 1
V ESD testers and providing calibration service - 7.6.3.2 External Laboratory IATF 16949 - Automotive Quality Systems Standard 3
Q Can a waivers for external calibration labs be implemented via the PPAP process? IATF 16949 - Automotive Quality Systems Standard 3
Z Ring Gage Calibration Sizes Shrunk with New External Vendor (supplier) General Measurement Device and Calibration Topics 16
R TS16949 External Calibration Laboratory Requirements - Thermocouples IATF 16949 - Automotive Quality Systems Standard 7
M TS Clause 7.6.3.2 - Can Non-Accreditated External Laboratory Carry out Calibration? IATF 16949 - Automotive Quality Systems Standard 14
S Calibration System was In-House now Outside (External Service Provider) General Measurement Device and Calibration Topics 14
F Calibration Tags - External company applied their tags - Do I need to apply our tags Document Control Systems, Procedures, Forms and Templates 6
C External calibration provider does not issue uncertainty of measurement Measurement Uncertainty (MU) 2
P About 7.6.3.2 External Laboratory - Calibration by Manufacturer General Measurement Device and Calibration Topics 2
rvin59 17025 Calibration Requirements - External calibration service cannot calibrate ISO 17025 related Discussions 4
Tom W IM&TE - Outside calibration companies - 7.6.3.2 External Laboratory IATF 16949 - Automotive Quality Systems Standard 2
I Defined External Calibration Acceptance Criteria General Measurement Device and Calibration Topics 10
G What is your expenses for external calibration per year? General Measurement Device and Calibration Topics 7
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
R External Audit and Certificate prorogation due to the pandemic General Auditing Discussions 10
N Audit non-compliance API Q1 - Use of External Documents 4.4.4 in Product Realization Oil and Gas Industry Standards and Regulations 5
J 510(k) for a control kit for an external IVD test kit 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R Is it required to have an SOP for external audits? Medical Device and FDA Regulations and Standards News 7
Pmarszal External Standards and Regulations Management Process Document Control Systems, Procedures, Forms and Templates 10
I What kind of wine best complements the Friday that you close out your external audit findings? Opinions are welcome. Coffee Break and Water Cooler Discussions 12
B How to apply external voltage to SIP/SOP IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Qualifying commercial off the shelf (COTS) external suppliers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
B Internal and external auditor competency to CSR's IATF 16949 - Automotive Quality Systems Standard 20
N IATF 16949:2016 7.1.5.3.2 External Laboratory - How to approve the Testing Laboratory without accreditation scope IATF 16949 - Automotive Quality Systems Standard 2
A OHSAS 18001 external auditor finding personal interpretation? Occupational Health & Safety Management Standards 5
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Reduction of software class based on multiple external risk controls IEC 62304 - Medical Device Software Life Cycle Processes 5
C Identifying and Controlling External Documents Document Control Systems, Procedures, Forms and Templates 3
Q AS9120B flow down to external providers: Records Retention AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
B Using external FDA and ISO 13485 audit as internal audit Internal Auditing 6
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
C Determining if Maintenance Contractor is an External Service subject to ISO 9001 Clause 8.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
qualprod The Perfect audit? External Audit causes a significant negative impact in a company General Auditing Discussions 9