Hi all,
We're developing a medical humidifier - to be used in hospitals (no home use). In case we'll plan the device to be with an external power supply unit, will it change any applicable standard requirements (as compared to a device with an internal power unit) ? I guess it will be relevant for some requirements in the 60601-1 or 60601-1-2?
Thanks and best,
Elsmar Forum Sponsor
Marc
Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#2
This is a "specialty" question. A quick "Bump". Thanks in advance to anyone who can help with this.
#4
I think that looking at 62368 is not necessary, or relevant, as it relates to IT and AV equipment.
The external PSU may be an IT type, but if used with a humidifier it has to meet the requirements for Medical Electrical Equipment - that is the 60601 series.
Being external can cause a little difficulty in testing, but the standards are now written in such a way as to allow understanding of this situation.
Is there any particular part of the standards that is causing difficulties?
Jaydub
Involved In Discussions
#5
If it is a dedicated power supply for your system, some EMC tests (Conducted Emissions and Immunity, EFT, Surge) will probably be done on the AC side of the power supply. If it is something like a generic 48V power supply, some of the tests may be done on the DC side.
Thread starter Similar threads Forum Replies Date
A Medical Device Leakage Current Measurement with External Power Supply IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
V External power supply with reinforced insulation as a simple secondary circuit? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
O LVD Requirements with Use of External Power Supply CE Marking (Conformité Européene) / CB Scheme 1
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Can I use CE marked Li-Ion external power bank for medical device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J External third party audits Registrars and Notified Bodies 1
J Help settle a disagreement: Should external providers of preventive maintenance be on your ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
J Requirement for Retention of Records of Withdrawn Documents of External Origin Document Control Systems, Procedures, Forms and Templates 3
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
L IATF external audit virtual (remote) IATF 16949 - Automotive Quality Systems Standard 13
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
R External Audit and Certificate prorogation due to the pandemic General Auditing Discussions 10
N Audit non-compliance API Q1 - Use of External Documents 4.4.4 in Product Realization Oil and Gas Industry Standards and Regulations 8
J 510(k) for a control kit for an external IVD test kit 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R Is it required to have an SOP for external audits? Medical Device and FDA Regulations and Standards News 7
Pmarszal External Standards and Regulations Management Process Document Control Systems, Procedures, Forms and Templates 10
I What kind of wine best complements the Friday that you close out your external audit findings? Opinions are welcome. Coffee Break and Water Cooler Discussions 12
B How to apply external voltage to SIP/SOP IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Qualifying commercial off the shelf (COTS) external suppliers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
B Internal and external auditor competency to CSR's IATF 16949 - Automotive Quality Systems Standard 20
N IATF 16949:2016 7.1.5.3.2 External Laboratory - How to approve the Testing Laboratory without accreditation scope IATF 16949 - Automotive Quality Systems Standard 2
A OHSAS 18001 external auditor finding personal interpretation? Occupational Health & Safety Management Standards 5
D Reduction of software class based on multiple external risk controls IEC 62304 - Medical Device Software Life Cycle Processes 5
C Identifying and Controlling External Documents Document Control Systems, Procedures, Forms and Templates 3
C External Laboratories and OEM Calibration IATF 16949 - Automotive Quality Systems Standard 0
Q AS9120B flow down to external providers: Records Retention AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
B Using external FDA and ISO 13485 audit as internal audit Internal Auditing 6
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
C Determining if Maintenance Contractor is an External Service subject to ISO 9001 Clause 8.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
qualprod The Perfect audit? External Audit causes a significant negative impact in a company General Auditing Discussions 9
K AS9100D 8.4.1.1 external providers question - Walmart, Home Depot, our lawn care team. etc. AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 20
A External Auditor issue with Internal Audits Internal Auditing 7
R Do we need to treat local law as external origin documents? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
D ISO 9001:2015 Clause 8.4.3 "Information for External Providers" buying from online retailers. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
M Applicability of Means of Protection, working voltage in an Automated External Defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D Necessity of external watchdog next to internal watchdog ISO 14971 - Medical Device Risk Management 1
B How to reply NCR on ineffectiveness of corrective action during IATF external audit? This is repeated issue whereby some mistake was done. IATF 16949 - Automotive Quality Systems Standard 7
I AS9100 8.4.2 Type and Extent of Control - External provider test reports Manufacturing and Related Processes 24
P AS9120B Control of External Providers for Franchised Distribution AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S Tools and equipment provided by customer - Considered as external provider? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
T Transportation - External Service Provider ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S ISO9001:2015 Clause 9.1 - What the external auditor will look at? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
N IATF 16949 7.1.5.3.2 External Laboratory - Calibration of Nikon machine in factory IATF 16949 - Automotive Quality Systems Standard 24
Ashok sunder How to determine the internal and external issues in ISO 45001 standard? Occupational Health & Safety Management Standards 2
Q How to efficiently and compliantly reference external standards/regulations Other Medical Device Related Standards 2
B IATF 16949 7.1.5.3.2 FAQ #7 Audit Finding - External Calibration Laboratory IATF 16949 - Automotive Quality Systems Standard 7
P IATF checklist for External Supplier Audit IATF 16949 - Automotive Quality Systems Standard 9
J Ignoring a Failed Gage R&R and upcoming external audit IATF 16949 - Automotive Quality Systems Standard 22
sswaim IATF 16949 7.1.5.3.2 External Calibration Laboratory Requirelents IATF 16949 - Automotive Quality Systems Standard 15
Similar threads
Top Bottom