Are two questions (about 60601-1).
It is possible to consider an external power supply with reinforced insulation (Fulfils Class II SELV for the following applications: IEC 60601-1, UL 2601, VDE, CE label, fulfils medical application class B / BF / CF) as a simple secondary circuit (which would have a transient voltage network of 1500V, operating voltage in primary circuit is 240 ac)?
This external power supply have an insulation of 4000V (2 MOPP), <10uA of leakage current, and 5 V cc in the ouput.
If I use this power supply (only this one), I need to protect with one aditional MOPP (of 1500V test voltage) applicable parts which would connect to it?
I think that this MOPP must be of 1000V test voltage (working voltage of 5 Vcc in the output of my external voltage supply.
Please help me.
BR
Velarde
Peter Selvey
The main clause here is 8.4.2 c) of IEC 60601-1:2005.
For the operator, it is clear that a 5V circuit, fully insulated from mains parts and relatively low power (<240VA) is safe to contact directly. No MOOP is required.
For the patient, the need is based on the risk of electric shock, taking into account the probability of contact with the patient (as stated in the standard).
Note that in the end, risk should be based on the probability of harm, not the probability of contact.
For 5V, examples of low impedance contacts are ECG electrodes, catheters, large metal surfaces or use in a wet environment. In order to current to flow, two low impedance contact points are required (a return path back the GND of the 5V circuit). If the current path includes the operator, the operator would need to have two low impedance contacts as well.
Overall, the probability of harm is usually negligible.
If the probability of harm is negligible, no MOPP is required (which is normally the case).
V
Velarde
Following this thread...
My medical device is an internally powered equipment, but has a connecting mean (connection plug) to Main network (for recharging the battery), therefore it must comply with Class I or Class II requirements when connected to the network.
The envelope of my medical device is not a patient's applicable part, as there no needs to be touched by the patient for normal operation (section 3.8 (60601-1)). But, I use with my medical device 3 channel electrode cables (no more than 30 cm length). In my opinion, is necessary to consider the envelope as patient's applicable part, because the risk that exists due to the proximity to the patient (30cm electrodes cables) (section 4.6 (60601-1)). This situation will be included in the Risk Management Process (4.6). Then, the envelope of my medical device must comply with patient's applicable parts requirements.
In order to comply those requirements, I use a 60601-1 certified external power supply (reinforced insulation 4000V, CF, <10uA leakage). The output of my external power supply is 5V cc. I think that this power supply must have 2 MOPP (230 Vac of operation voltage) on its output, because it will be connected into the envelope connection plug and it must be tested with 4000V.
Certain Certification Laboratory suggests apply an (1500V, for 230V ac) insulation test between the output of my power supply and patient's applicable parts. I think that if the envelope is treated as a patient's applicable part, it is not necessary this test, because the patient is protected with two MOPP.
What do you guys think about my analysis, including considering the envelope as patient's applicable part?
Best Regards
Peter Selvey
For the enclosure (envelope?) under Clause 4.6, keep in mind it is risk of harm to consider, not the risk of contact. It may be that in a particular situation the probability of contact with the patient is high (e.g. due to short 30cm cables), but still chance of electric shock is negligible.
It is confusing because the circuits inside the enclosure are already safe in general, so the risk management is really addressing the question of whether the special (more strict) requirements for patient contact are necessary. If the enclosure is plastic, I don't see strong reason to apply Clause 4.6.
For the electrodes, under Clause 8.3 b) the circuit must be Type BF or Type CF. In this case you will need an extra 1.5kV / 4mm isolation barrier between the electrode circuit and all other parts, including the 5V power supply. Such a barrier would normally be implemented with DC/DC converters, optocouplers and the like.
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Medical Device Leakage Current Measurement with External Power Supply IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O LVD Requirements with Use of External Power Supply CE Marking (Conformité Européene) / CB Scheme 1
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Can I use CE marked Li-Ion external power bank for medical device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J External third party audits Registrars and Notified Bodies 1
J Requirement for Retention of Records of Withdrawn Documents of External Origin Document Control Systems, Procedures, Forms and Templates 3
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
L IATF external audit virtual (remote) IATF 16949 - Automotive Quality Systems Standard 13
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
R External Audit and Certificate prorogation due to the pandemic General Auditing Discussions 10
N Audit non-compliance API Q1 - Use of External Documents 4.4.4 in Product Realization Oil and Gas Industry Standards and Regulations 8
J 510(k) for a control kit for an external IVD test kit 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R Is it required to have an SOP for external audits? Medical Device and FDA Regulations and Standards News 7
Pmarszal External Standards and Regulations Management Process Document Control Systems, Procedures, Forms and Templates 10
I What kind of wine best complements the Friday that you close out your external audit findings? Opinions are welcome. Coffee Break and Water Cooler Discussions 12
B How to apply external voltage to SIP/SOP IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Qualifying commercial off the shelf (COTS) external suppliers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
B Internal and external auditor competency to CSR's IATF 16949 - Automotive Quality Systems Standard 20
N IATF 16949:2016 7.1.5.3.2 External Laboratory - How to approve the Testing Laboratory without accreditation scope IATF 16949 - Automotive Quality Systems Standard 2
A OHSAS 18001 external auditor finding personal interpretation? Occupational Health & Safety Management Standards 5
D Reduction of software class based on multiple external risk controls IEC 62304 - Medical Device Software Life Cycle Processes 5
C Identifying and Controlling External Documents Document Control Systems, Procedures, Forms and Templates 3
C External Laboratories and OEM Calibration IATF 16949 - Automotive Quality Systems Standard 0
Q AS9120B flow down to external providers: Records Retention AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
B Using external FDA and ISO 13485 audit as internal audit Internal Auditing 6
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
qualprod The Perfect audit? External Audit causes a significant negative impact in a company General Auditing Discussions 9
K AS9100D 8.4.1.1 external providers question - Walmart, Home Depot, our lawn care team. etc. AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 20
A External Auditor issue with Internal Audits Internal Auditing 7
R Do we need to treat local law as external origin documents? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
D ISO 9001:2015 Clause 8.4.3 "Information for External Providers" buying from online retailers. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
M Applicability of Means of Protection, working voltage in an Automated External Defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D Necessity of external watchdog next to internal watchdog ISO 14971 - Medical Device Risk Management 1
B How to reply NCR on ineffectiveness of corrective action during IATF external audit? This is repeated issue whereby some mistake was done. IATF 16949 - Automotive Quality Systems Standard 7
I AS9100 8.4.2 Type and Extent of Control - External provider test reports Manufacturing and Related Processes 24
P AS9120B Control of External Providers for Franchised Distribution AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S Tools and equipment provided by customer - Considered as external provider? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
T Transportation - External Service Provider ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S ISO9001:2015 Clause 9.1 - What the external auditor will look at? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
N IATF 16949 7.1.5.3.2 External Laboratory - Calibration of Nikon machine in factory IATF 16949 - Automotive Quality Systems Standard 24
Ashok sunder How to determine the internal and external issues in ISO 45001 standard? Occupational Health & Safety Management Standards 2
Q How to efficiently and compliantly reference external standards/regulations Other Medical Device Related Standards 2
B IATF 16949 7.1.5.3.2 FAQ #7 Audit Finding - External Calibration Laboratory IATF 16949 - Automotive Quality Systems Standard 7
P IATF checklist for External Supplier Audit IATF 16949 - Automotive Quality Systems Standard 9
J Ignoring a Failed Gage R&R and upcoming external audit IATF 16949 - Automotive Quality Systems Standard 22
sswaim IATF 16949 7.1.5.3.2 External Calibration Laboratory Requirelents IATF 16949 - Automotive Quality Systems Standard 15