"Fast action" claim and MDR (Medical Device Regulations)
cam5603
Hello all :bigwave:,
I have a question concerning claims of medical device and regulation. The medical device is a treatment for wax plug and I would like to make sure that we can use “fast action” or “fast-acting” as a claim.
Do you know any requirements for this kind of claims?
An in-vitro study was performed on ear wax and the dissolution of the ear wax was twice faster with our product than with other solutions. Is it enough to use this kind of claim?
:thanx:
FoGia
Does "fast action" of the wax plug relate to a clinical safety or performance aspect? If the answer is yes it needs to be supported in your clinical evaluation report, otherwise not.
Now I see a dissonance between your claim and your data. Your in-vitro study shows that your products dissolves wax faster, not necessarily 'fast'. You could ask a panel of users of what duration they would define as being 'fast'. It could be that even twice as fast isn't fast enough for the users, or that current products are already considered fast and a drop of 50% wouldn't make a difference.
Even if it's not related to clinical performance, make sure that your evidence material is strong in case a competitor would have the idea to sue you for inaccurately claiming that your product dissolves was faster than theirs.
Mark Meer
In my opinion, your approach is fine....but I'm not aware of any authoritative guidance regarding subjective terms like this. If you really want to cover your bases, I suppose you could always use the caveat/asterisk in your labelling. For example:
"Fast-acting* wax removal!
*when compared with competing brands"
...You could ask a panel of users of what duration they would define as being 'fast'. It could be that even twice as fast isn't fast enough for the users, or that current products are already considered fast and a drop of 50% wouldn't make a difference....
This is an interesting point of discussion regarding subjective language.
Are we concerned about misleading claims liabilities? If so, I question the utility of having to assemble a panel to define a subjective term.
Consider, for example you take 3 people: one says "fast" means 3s, another says 8s, and the last says 10s. What are you to do with this data? Average it, and conclude that "fast" is 7s? Assuming your sample is representative, and you base your design requirements on this definition of "fast", you'd end up with something that 1/3 of users would not consider "fast".
Don't get me wrong: from a design-input standpoint, this data could still be very useful. But as far as being concerned with being cited for false advertising, I don't know how valuable this kind of undertaking would be...
FoGia
It all depends of the context, it may be that if one third don't have an opinion and the rest agrees then it's fine, it's up to you to define the criteria. But if one third of your panel strongly disagrees with the wording 'fast' in my opinion it may be necessary to reconsider the claim. I put that sort of claim in the same category as the product being 'user-friendly','reliable','confident'..., which are also purely subjective by nature. In that kind of situation when there is disagreement on the meaning of subjective words I typically advise against using such language in marketing material for the reason that they're hard to substantiate, they tend to confuse/deceive the target audience and they divert their attention from more important claims. Without going to a court case I've had similar discussions with business partners and notified bodies.
This is fine indeed, because it makes the claim more objective - the claim is now about the product being faster - however competitors may be swift to complain.
cam5603
Thanks all for your reply !
I have found a guidance document from UK that could help ("self care medical devices advertising guideline" on PAGB website). Even if we are not selling in UK we might be able to use it as an inspiration.
In this document, it is mentionned " 'Fast' claims
For most therapy areas 'fast' is currently taken to mean 'within 30 minutes'."
But as you said this is not an objective claim and it makes it hard to define...
Mark Meer
I put that sort of claim in the same category as the product being 'user-friendly','reliable','confident'..., which are also purely subjective by nature. In that kind of situation when there is disagreement on the meaning of subjective words I typically advise against using such language in marketing material for the reason that they're hard to substantiate, they tend to confuse/deceive the target audience and they divert their attention from more important claims...
Good point of discussion!
In practice, I think there is a risk assessment that needs to be considered. If we are talking about misleading advertising liabilities, there is certainly a risk/benefit call to make.
Consider two companies selling similar products.
Company A advertises "This is the best, top-rated, user-friendly device that does X.", and backs this up through some (tenuous) surveys they carried out themselves.
Company B, worried about potential false advertising, simply advertises "This device does X."
There is no doubt that Company A's marketing is more effective, and they do technically have evidence to back up their claims. They would be considering the probability that (a) a complaint is made, and (b) that their evidence is rejected, and weighing it against the benefits of better marketing.
Company B, on the other hand, recognizes that these are all ultimately subjective terms, so opts to avoid them altogether in their marketing. They eliminate the risk of misleading advertising liability, but do so at the cost of less effective marketing (compared to their competition - Company A).
It seems that you would, in general, advise to take Company B's strategy. However, I think there is a strong case to be made that Company A's strategy makes better business sense...
FoGia
This is a discussion I'm sure many people are familiar with, here are a few counterarguments to the use of 'light subjective fluffy claims' ('top-rated', 'best in class', etc):
  • We are all familiar with such claims because they're used commonly for consumer goods in markets that are heavily commoditized (furniture, cars, washing powder, frozen pizzas, etc) where company try to differentiate themselves from others by appealing to customer's perceptions instead of objective value. I recognize that there are numerous examples of medical devices that are in such a position, but overall, we're dealing with medical devices, products that have an impact on health which require more scrutiny. You can't use the same marketing methods for consumer goods and medical devices (or drugs).
  • Other stakeholders(CB, insurances, purchasing organization, clinicians, patients, ...) on the market are more and more in demand for products that have a clear benefit risk profile, to those people light claims are just annoying noise signal and you're not making your life easier
  • The more emphasis you put on such claims in your marketing material the more you give the impression that you try to cover up for weak/inexistent clinical claims
  • Light claims can be perceived to insult the intelligence of your customers if not well formulated, they can backfire
  • And from a broader perspective it depends also how you want your company to be perceived by the market, is your marketing strategy blowing a lot of air, or do you want to build an image of effective clinical solution provider?
I don't say that I'm always against light claims, my first response was not to dismiss the claim of the OP, but I'm going against the perception that "the shinier the marketing material is, the stronger sales will be". In a proper context they can be of value, but don't stand a chance against robust clinical claims.
To be clear, these are in no way commentaries about the product of the original poster, my post is a commentary on the broader topic of subjective claims.
Mark Meer
Agreed. But for the sake of discussion we are assuming that the clinical claims are the same.
I'd like to agree with you regarding the potential to "backfire" with respect to using such terms in marketing. I'd say possibly but not probably...Perhaps medical device consumers are generally more informed or have a particular focus on objective metrics, but at the end of the day they are still consumers and, I suspect, still prone to persuasion when it comes to subjective/emotive language to some degree.
Again, there is a risk/benefit calculation here, and I'd love to see some data to help with the calculation. To what degree do consumers find "top-rated" or "user-friendly" claims to be an insult to their intelligence, or perceive that it is a cover-up for other product deficiencies? To what degree are such terms perceived as "annoying noise signal(s)"? ....I suspect not much...but I'm open to being proven wrong.
Ultimately, I'm for letting the market sort it out in time. To whatever degree use of such terms is prone to backfire, this will reflect in sales and people will adjust their marketing accordingly. It may be the case that such terms are so ubiquitous that we are all numb to them, and it makes no difference either way. Only time will tell, I suppose...
H Operational planning and product and service provision when things happen so fast ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
O EMC - EFT (Electrical Fast Transients) on SIP couples erroneously with other cable CE Marking (Conformité Européene) / CB Scheme 1
Sam Lazzara Expedited, fast-track Notified Body Technical File reviews EU Medical Device Regulations 3
C QC and Compliance for a fast growing company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
R What Documentation is done by QA in FMCG (Fast Moving Consumer Goods) Misc. Quality Assurance and Business Systems Related Topics 2
F Fast Track Registration in Costa Rica for Medical Devices registered in Canada Canada Medical Device Regulations 3
K Fast Response Tracking Board (QSB+) Customer and Company Specific Requirements 6
Ajit Basrur Paul Walker of 'Fast and Furious' dead at 40 - RIP - Nov 2013 Coffee Break and Water Cooler Discussions 1
N NCR and CAPA in the FMCG (Fast Moving Consumer Goods) Industry Nonconformance and Corrective Action 1
S Scheduling Jobs through a fast paced Sheet Metal Job Shop Manufacturing and Related Processes 5
A Does anyone know of a WHO fast-track program for Medical Devices? Other Medical Device Regulations World-Wide 3
N RS485 stand to meet the EFT (Electrical Fast Transients) test according to IEC 60601 ISO 13485:2016 - Medical Device Quality Management Systems 3
N Fast Track for Medical Devices Registration in Mexico that are already CE Marked Other Medical Device Regulations World-Wide 11
S Fast growing small O&G Company QMS implementation - My path forward ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
P Statistical Sampling of Packaging Material Testing? Fast Moving Consumer Goods Supplier Quality Assurance and other Supplier Issues 4
S Ribbon Thickness Gauge - Need fast response please Manufacturing and Related Processes 3
Q Fast Response Boards - Questions and a Discussion Nonconformance and Corrective Action 3
R GM (General Motors) QSB Fast Response Program Customer and Company Specific Requirements 12
S Fast Talking - People who talk fast Coffee Break and Water Cooler Discussions 24
ScottK Soliciting recommendations for a fast turnaround calibration house - Mass Flow Meter Manufacturing and Related Processes 1
R Extremely fast TS 16949 implementation IATF 16949 - Automotive Quality Systems Standard 8
Sidney Vianna OASIS database - deadline for setting up an administrator is fast approaching AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
Stijloor Chrysler's Campi (new purchasing boss) will move fast to buy global. World News 11
D Food Sampling Matrix for Microbial Testing (number per batch) for Fast Food Processor Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
Marc How fast is the fastest you have ever driven a land vehicle? Coffee Break and Water Cooler Discussions 105
Sidney Vianna Crowdsourcing - Fast, Collaborative thinking After Work and Weekend Discussion Topics 2
L Estimating process - Material "fast quote" - Current material prices on the web Lean in Manufacturing and Service Industries 4
Marc Super-fast DSL, unlimited phone service and 100 TV channels - $38 a month After Work and Weekend Discussion Topics 2
L GM's new Fast Response process - Anyone with experiences to share? Customer and Company Specific Requirements 21
Sidney Vianna Fast gift Funny Stuff - Jokes and Humour 2
K Fast ISO13485:2003 Implementation - What would you do first? ISO 13485:2016 - Medical Device Quality Management Systems 27
Marc Electric cars as fast as a Ferrari World News 1
M Where can I buy (purchase) copies of the ISO/TS 16949:2002 Standard fast? IATF 16949 - Automotive Quality Systems Standard 3
I Fast Runs - Four lots, one day each, every month - What chart to use? Statistical Analysis Tools, Techniques and SPC 2
A Corrective Action Training ISO 13485:2016 - Medical Device Quality Management Systems 1
R Not every INCIDENT report will lead to a corrective action EU Medical Device Regulations 6
V ISO 10360-5: 2020 Gap analysis and Action plan Excel .xls Spreadsheet Templates and Tools 1
J Action Not defined in procedure - NC - Incoming inspection AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
L Separate Corrective Action and Preventive Action Procedures ISO 13485:2016 - Medical Device Quality Management Systems 13
T AS9100D Clause 10.2.1g Supplier Corrective Action for each and every nonconformity? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R Non conformance (NC) or Corrective & Preventive action (CAPA) CE Marking (Conformité Européene) / CB Scheme 7
I Setting action completion dates Nonconformance and Corrective Action 3
A Medical device Mode Of Action CE Marking (Conformité Européene) / CB Scheme 2
D Corrective Action Timing - IATF CB Deadline IATF 16949 - Automotive Quality Systems Standard 2
R AS 9100D - Containment and Corrective action AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
P Customer Corrective Action Requests in OASIS? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
C Corrective action for failure in documents control ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
nadhar2 Classification of Action Items Misc. Quality Assurance and Business Systems Related Topics 3
C If it doesn't prevent a non-conformance, is it a preventive action? IATF 16949 - Automotive Quality Systems Standard 13
G Assignable cause/corrective action list for SPC Software Statistical Analysis Tools, Techniques and SPC 3