Hi,
I am just thinking that for chemistry analyzers (IVD instruments) which do not have a direct impact to the patient safety, how do companies report their removals and corrections.

If there is any replacement of power boards or repair at the customer site or at the manufacturing site, without reducing any risk to health of the user, then are these considered to be reportable repairs (corrections or removals)?

For such IVD instruments, are all repairs reportable under CAR (Correction and Removal regulations)? I don't think so.
How does Companies/Industry practice?

Please share your thoughts!

Regards,
sreenu
 
Elsmar Forum Sponsor

TWA - not the airline

Trusted Information Resource
#3
I'm not in the field of IVDs, but IMHO 21CFR820.200 (c) applies here: Each manufacturer who receives a service report that represents an event which must be reported to FDA under part 803 of this chapter shall automatically consider the report a complaint and shall process it in accordance with the requirements of ?820.198.
 
M

MIREGMGR

#4
My view is that field upgrade/preventive-maintenance (PM) actions are not reportable as Corrections unless they are related to an Adverse Event per 21CFR 803, or constitute a substantial design change for which premarket notification would be required.

A "repair" on the other hand would seem to inherently involve patient risk, at least at the level of non-availability of service. If however the non-availability of service resulting from the action is minimal relative to system use-frequency due to good scheduling, ready availability of components, design-for-maintenance and so forth, then perhaps the action can be characterized as maintenance instead of repair and the patient risk regarded as insignificant.
 
Last edited by a moderator:

sreenu927

Quite Involved in Discussions
#5
When you say "availability"- how soon or is there any timeline? I know the regulations do not provide any specific timeline and is instrument or reagent kit or assay dependent based on "reasonably suggests"..but is there any industry guidance or practice that if the service is provided with in a reasonable timeline of x days, then it is ok, else considered to be a removal or correction case!

Thanks,
Sreenu
 
M

MIREGMGR

#6
I don't know of any guideline or common practice. As far as I know, the situation is unusual enough that every instance is dealt with individually.

In my view, non-availability for say three days due to a need for repair is reasonable if the equipment is used on average every six days. I could write a rationale for that circumstance.

If on the other hand the equipment is used on average once daily, and will be down for several days, then you have a potential patient risk situation due to non-availability that I wouldn't be able to rationalize, and I might feel obligated to Report as a field action.

Obviously this is all about judgement calls and how you write your rationale.
 

sreenu927

Quite Involved in Discussions
#7
Hi MIREGMGR,

Thank you. I got some clue from your response, "...equipment is used on average once daily, and will be down for several days, ..".

May be, I can think of something relevant to our instruments in this line to develop a strategy of reporting the field action (correction/removal).

Many Thanks,
Sreenu
 
C

cff920

#8
If you feel uncertain, write to FDA for clarification. This is the safest way to protect your firm from relevant legal exposure.
 
Thread starter Similar threads Forum Replies Date
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 0
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
D FDA Information - Revising the Instructions for Use US Food and Drug Administration (FDA) 0
S Transitional Adolescent A and B - "CDRH PREMARKET REVIEW SUBMISSION COVER SHEET FORM FDA 3514" Medical Device and FDA Regulations and Standards News 1
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
P MSDS for MVQ FDA White, Vinyl Methyl Silicone Rubber EU Medical Device Regulations 4
S Manufacturing Process FDA FOIA Medical Device and FDA Regulations and Standards News 3
S Manufacturing Process FDA FOIA US Food and Drug Administration (FDA) 4
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Supplier requirements - Major supplier is a Non-Profit registered with ICCBBA (FDA UDI) Supply Chain Security Management Systems 12
C RA (Regulatory Assurance) Training (FDA) looking for resources Training - Internal, External, Online and Distance Learning 5
E FDA 513(g) Cover Letter US Food and Drug Administration (FDA) 5
S Records - Do's and don't' of record entries (FDA - 21 CFR 820) Records and Data - Quality, Legal and Other Evidence 13
B New Facility register with FDA Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J FDA regulation on decorative contact lenses Medical Device and FDA Regulations and Standards News 5
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F FDA classification for a mobile app Medical Information Technology, Medical Software and Health Informatics 3
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
O Any info on release date of FDA “Computer Software Assurance for Manufacturing and Quality System Software” document? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
P Writing a presubmission to the FDA prior to the De novo submission US Food and Drug Administration (FDA) 4
K FDA - Can we have more than 1 Initial Importer Medical Device and FDA Regulations and Standards News 4
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
R Importers - For the FDA Registration/ Listing, is the CMO the importer? Medical Device and FDA Regulations and Standards News 6
Z Iterative development and FDA change requests IEC 62304 - Medical Device Software Life Cycle Processes 12
P FDA Approved Product Contact Parts ISO 13485:2016 - Medical Device Quality Management Systems 6
O Clarifying FDA definition of "finished device" and “capable of functioning” 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
JoCam FDA Registration for Sub-contract manufacturers Medical Device and FDA Regulations and Standards News 2
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Contract manufacturer FDA requirements foreign company US Food and Drug Administration (FDA) 6
P Choice of FDA reviewer US Food and Drug Administration (FDA) 6
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
T Definition Human Use (FDA) Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 7
K FDA 510k electrosurgery 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J Leveraging FDA 510k Clearance for International Registrations Other Medical Device Regulations World-Wide 2
shimonv FDA News FDA guidance on Multiple Function Device Products (8/2020) Other US Medical Device Regulations 1
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
J FDA notification of address change US Food and Drug Administration (FDA) 2
N Usability testing required for FDA IDE (investigational device exemption)? Human Factors and Ergonomics in Engineering 8
A FDA Class Classification for a cabinet 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
L FDA Biocompatibility Requirements - Transitory Contact Other US Medical Device Regulations 1
Edward Reesor FDA DM and /or Class I Life Saving US Food and Drug Administration (FDA) 0
Watchcat FDA will not tolerate fraud…meaning what, exactly? Other US Medical Device Regulations 8
K FDA Premarket Cybersecurity Guidance - 4 questions Other US Medical Device Regulations 5
gunnyshore Form FDA 3500A MedWatch eSubmitter 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
Watchcat FDA Webinar Series - N95 Respirators Other US Medical Device Regulations 0
M Off-Label Use - Clarification of FDA Policy US Food and Drug Administration (FDA) 1

Similar threads

Top Bottom