My question relates to inventory organization with relationship to the FDA and pest control, specifically, placement of racks and/or pallets against internal and/or external walls.

Background: My company contract-manufactures for multiple industries, including Class I medical devices. We don't dabble in pharma and our products are pretty low risk items; as a result, our facility has never been inspected by the FDA, despite several years of registration. A couple of our customers hire a third-party auditing company to perform annual surveillance audits to 21 CFR 820, cGMP, etc. Frequently this third party hires former FDA inspectors as auditors. Two separate auditors with the same FDA background have made the same comment; we shouldn't push pallets too close to the walls, as it obstructs the path of pests, who normally will stick close to the wall, and shoots them off into the production area. We've never had a finding on this matter, but having the same observation come up during our facility tour has made me curious.

Is this something that could come up as a finding in the event that we finally undergo an FDA inspection? If so, what FDA regulation are we not meeting?
 
Elsmar Forum Sponsor

v9991

Trusted Information Resource
#2
Hope u have already referenced the 21 CFR part 211.; depending upon the observation, Observation on creating an space between pallet & wall could relate to ease-of cleaning- maintenance etc
Sec. 211.42 Design and construction features.
(a) Any building or buildings used in the manufacture, processing, packing, or holding of a drug product shall be of suitable size, construction and location to facilitate cleaning, maintenance, and proper operations.

Sec. 211.56 Sanitation.
(a) Any building used in the manufacture, processing, packing, or holding of a drug product shall be maintained in a clean and sanitary condition, Any such building shall be free of infestation by rodents, birds, insects, and other vermin (other than laboratory animals). Trash and organic waste matter shall be held and disposed of in a timely and sanitary manner.
(b) There shall be written procedures assigning responsibility for sanitation and describing in sufficient detail the cleaning schedules, methods, equipment, and materials to be used in cleaning the buildings and facilities; such written procedures shall be followed.
(c) There shall be written procedures for use of suitable rodenticides, insecticides, fungicides, fumigating agents, and cleaning and sanitizing agents. Such written procedures shall be designed to prevent the contamination of equipment, components, drug product containers, closures, packaging, labeling materials, or drug products and shall be followed. Rodenticides, insecticides, and fungicides shall not be used unless registered and used in accordance with the Federal Insecticide, Fungicide, and Rodenticide Act (7 U.S.C. 135).
(d) Sanitation procedures shall apply to work performed by contractors or temporary employees as well as work performed by full-time employees during the ordinary course of operations.
 

Schkund

Involved In Discussions
#4
Correct me if I'm wrong, but isn't Part 211 for pharmaceutical devices? We don't manufacture any pharmaceutical products (no products with active ingredients or anything that could otherwise be classified as a "drug"). We don't have any pharmaceutical products registered with the FDA, only Class I medical devices, and no combination products. I wouldn't think we would be required to meet Part 211 requirements, would we? Our products don't fall within its scope.
 

Schkund

Involved In Discussions
#5
Ajit, thank you for the attachment, that does provide some clarification. I'll delve into the attached forum discussion further.
 

v9991

Trusted Information Resource
#6
Correct me if I'm wrong, but isn't Part 211 for pharmaceutical devices?
We don't manufacture any pharmaceutical products (no products with active ingredients or anything that could otherwise be classified as a "drug").

We don't have any pharmaceutical products registered with the FDA, only Class I medical devices, and no combination products.

I wouldn't think we would be required to meet Part 211 requirements, would we? Our products don't fall within its scope.
For the sake legality - true; but its difficult to argue just alone based on it.
But the response to be more based on assessment viz., pest-controls in facility; trends if any on pest/rodents; process taking care of any possible infestation; any requirements & controls at end with customer / manufacturer before its used/dispatched to market etc
besides the point, isn't it only a good practice to ensure "cleanability", pest/rodent control etc.
 

Schkund

Involved In Discussions
#7
Certainly, and our facility does have procedures for ensuring cleanliness as well as a third party to conduct pest control activities, which we review, but I guess my point is that I'm still not sure we would get an observation based on placement of our pallets near interior walls based on Part 211 requirements. If we aren't meeting an FDA regulation, I want to know what that regulation is to ensure there aren't other gaps I should be addressing. It seems to be a specific point brought up by former FDA inspectors, "nothing should be within 12 to 18 inches from interior walls"... where is this coming from? The forum that Ajit attached at least shows that I'm not the only one who is a bit confused by this specific requirement.
 

v9991

Trusted Information Resource
#8
following are pertinent references. Hope this helps...
next search is about FDA embracing / endorsing the following approach...

2013 Pest Management Standards For Food Plants
2.2 Rodent Program-Interior Survey All areas of the plant interior shall be surveyed for rodents and conditions which may lead to rodent infestation. The following observations must be identified and documented using a diagram and/or notes:
• Open doors, gaps beneath doors, or other holes, gaps or cracks which could permit rodent entry
• Clutter, debris or other potential rodent harborage location
• Lack of 18-inch inspection aisle along walls
• Areas of rodent infestation based on sightings, droppings, harborage or other evidence
• Spillage or other potential food sources which could lead to infestation
• Storage practices which are conducive to rodent infestation
National Pest Management Association, Inc. 2009 Pest Management Standards for Food Plants
Plants should have an inspection aisle of at least 18 inches along walls
.

Do Your Own Establishment Inspection A Guide to Self Inspection for the Smaller Food Processor and Warehouse
z Are products stored on pallets and at least 18 inches away from the walls? It is important to leave space for inspection aisles so that rodent and insect activity can be seen more readily. You might consider painting a white line on the floor along the walls to indicate inspection aisles.
 

Schkund

Involved In Discussions
#9
Interesting.... perhaps the FDA considers these references a "best practice" for pest control. Clearly this inspection aisle is widely used in the food industry; I could see the FDA expanding these guidelines to pest control in general. Thank you so much for your help! I'm glad to have a direction to point my inquiries!
 

Edward Reesor

Trusted Information Resource
#10
I'm not one to pick a fight or argue endlessly with an auditor, but I am confident that "best practices" in one industry does not always translate to "best practices" in others (or else that industry would be subject to those standards from the onset). Does this infer that the food industry would then be expected to conform to specific best practices applied to Class I medical devices? This is the very reason why there are so many standards spread across so many industries.

Unless the auditor can site the standard that applies to the Class I medical device manufacturer being audited, I would dispute the NC to the head office and (if need be) request an auditor who is more familiar with the industry they are auditing.
 
Thread starter Similar threads Forum Replies Date
P FDA inspections and electronic signatures Other US Medical Device Regulations 2
M Informational USFDA draft guidance – Nonbinding Feedback After Certain FDA Inspections of Device Establishments Medical Device and FDA Regulations and Standards News 0
A Can Internal Audit Observations be kept confidential from State FDA inspections ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
S FDA Enforcements for 2011 - RAPS article on FDA Inspections and Enforcements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
A Is taking Photos by the Auditor during FDA Inspections allowed? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
R FDA Inspections - Questions from a Device Manufacturer Submitting 510k Other US Medical Device Regulations 3
S FDA Webinar on Foreign Inspections - May 17, 2011 US Food and Drug Administration (FDA) 1
T What do "OAI" "VAI" and "NAI" used in FDA QSIT manual mean in inspections? ISO 13485:2016 - Medical Device Quality Management Systems 3
Scott Catron Any difference in FDA inspections since the risk-based approach was announced? US Food and Drug Administration (FDA) 6
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
D FDA Information - Revising the Instructions for Use US Food and Drug Administration (FDA) 0
S Transitional Adolescent A and B - "CDRH PREMARKET REVIEW SUBMISSION COVER SHEET FORM FDA 3514" Medical Device and FDA Regulations and Standards News 1
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
P MSDS for MVQ FDA White, Vinyl Methyl Silicone Rubber EU Medical Device Regulations 4
S Manufacturing Process FDA FOIA Medical Device and FDA Regulations and Standards News 3
S Manufacturing Process FDA FOIA US Food and Drug Administration (FDA) 4
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Supplier requirements - Major supplier is a Non-Profit registered with ICCBBA (FDA UDI) Supply Chain Security Management Systems 12
C RA (Regulatory Assurance) Training (FDA) looking for resources Training - Internal, External, Online and Distance Learning 5
E FDA 513(g) Cover Letter US Food and Drug Administration (FDA) 5
S Records - Do's and don't' of record entries (FDA - 21 CFR 820) Records and Data - Quality, Legal and Other Evidence 13
B New Facility register with FDA Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J FDA regulation on decorative contact lenses Medical Device and FDA Regulations and Standards News 5
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F FDA classification for a mobile app Medical Information Technology, Medical Software and Health Informatics 3
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
O Any info on release date of FDA “Computer Software Assurance for Manufacturing and Quality System Software” document? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
P Writing a presubmission to the FDA prior to the De novo submission US Food and Drug Administration (FDA) 4
K FDA - Can we have more than 1 Initial Importer Medical Device and FDA Regulations and Standards News 4
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
R Importers - For the FDA Registration/ Listing, is the CMO the importer? Medical Device and FDA Regulations and Standards News 6
Z Iterative development and FDA change requests IEC 62304 - Medical Device Software Life Cycle Processes 12
P FDA Approved Product Contact Parts ISO 13485:2016 - Medical Device Quality Management Systems 6
O Clarifying FDA definition of "finished device" and “capable of functioning” 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
JoCam FDA Registration for Sub-contract manufacturers Medical Device and FDA Regulations and Standards News 2
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Contract manufacturer FDA requirements foreign company US Food and Drug Administration (FDA) 6
P Choice of FDA reviewer US Food and Drug Administration (FDA) 6
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
T Definition Human Use (FDA) Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 7
K FDA 510k electrosurgery 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J Leveraging FDA 510k Clearance for International Registrations Other Medical Device Regulations World-Wide 2
shimonv FDA News FDA guidance on Multiple Function Device Products (8/2020) Other US Medical Device Regulations 1
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
J FDA notification of address change US Food and Drug Administration (FDA) 2
N Usability testing required for FDA IDE (investigational device exemption)? Human Factors and Ergonomics in Engineering 8

Similar threads

Top Bottom