FDA issues first draft guidance document on Mobile Medical Applications

Ronen E

Problem Solver
Staff member
Moderator
#11

Attachments

Elsmar Forum Sponsor

sagai

Quite Involved in Discussions
#12
Unless I have completely misunderstood something, let me share my thoughts.
Again, I am completely ignorant about your device, but let me have a theoretical example.

To make it very simple, let say the transferring device you have is collecting patient body temperature by Bluetooth communication , really keep it very simple, from multiple measurement devices attached to the patients and your device transfers these measurements to the central monitoring "server" for further processing and analysis.
So we get the body temperature of multiple patients via Bluetooth communication, we store temporarily it and than we transfer it, let say via Bluetooth to the analysis server station.

In my imagination I would raise the following potential hazards. It is not an extensive list, rarely the ideas first come into my mind.

1., Patient mix-up
There is a potential that our device transfers body temperature for a different patient than it was actually measured.
Severity:
The extent of such mistake should be evaluated by physician. Could be major, because patient may get drug that actually they do not have to and it cause a very not expected side effect in her/his condition, furthermore this drug may hide in a future the actual status of the patient, really I can not determine the extent of it.

Several hazardous situations can relate to this hazard, my thoughts would be:
1., when receiving the data, it's already corrupted for patient identifier and we are not aware of that we actually transfer the corrupted data,
2., when we are transferring the data, regardless we have received the correctly assigned body temperature data to the patient it is corrupted at the server side we are transferring to and patients are mixed up.

Probability, that these hazardous situations can occur also up to the physician who knows the extent and the landscape of this whole scenario.


The causes for these two could be:
1., a, sensors are mixing up patient data and we receive already corrupted data
1., b., Bluetooth communication is falsifying the correct data from sensor (due difference of different Bluetooth protocol, lack of false-safe communication for communication interruption cases, I am only guessing ... )
2., a., Due to our implementation (database storage, calculation, whatever mechanism we have that use the patient data) we corrupt the patient data.
It can be when due to our implementation consideration (our design and our corresponding code) of the Bluetooth protocol (may be we use a public library for this, but that again can fail to do it properly).
It can be, because the database we use corrupt the patient data (data format mix-up, truncation of patient identifier, anything).
It can be, because we do pre-calculation or formatting of the received data and our algorithm falsify the patient data, or the way than we store it does so.
2., b., Due to our implementation of the transfer Bluetooth communication, it mixes up the patient data. Or again, there is a probability, that regardless the communications are implemented correctly, the difference of the the different versions of the protocol may cause mix up.

Risk control measures to eliminate or to limit these causes:
I have no idea, it should be worked out with the engineers and the physician.
Because physician can decide if the planned control possibly reduce the probability of the hazardous situation to occur.
Moreover physician can decide if the device can be considered as safe for medical use. No on else can do so.

I would not like to be too long, but some more hazard category, I would raise for this case.
2., Measurement mix-up
Body temperatures can be measured in Celsius, Fahrenheit, whatever.
There is a hazard that we mix up this essential measurement.
3., Body temperature chronological order mix-up
There is a hazard, that we do not receive/store/transfer these data in a correct chronological order.
4., Bluetooth communication fraud.
Yes, somehow we need to consider the hazard when someone intentionally fraud the data due to security leak.
5., Data loss
There is several scenarios can drive to it, typical example of the boundary situation when our database gets full. Or when our maintenance mechanism in place on the database.

This would be my far not comprehensive example, just for a food of thoughts.

Well, it can happen, when someone opens up Pandora's jar though ;)

Cheers!

Ps.:
One more thing ...
The more or less expected content of the hazard analysis somehow should detail the following:
1., Identified hazards
2., Hazardous situations when the hazard can occur and lead to patient harm (1:n relation per hazard)
3., Severity per hazardous situation
4., Potential causes of the hazardous situation (some of us may say of the hazard, I have some doubts if the hazard itself can be elaborated into causes at the first place), its also 1:n relation to the hazardous situation
5., Initial Probability of the hazardous situation (initial probability, considering none of the actual implementation)
6., Risk evaluation of the hazardous situation based on the severity and initial probability.
7., Description of the Risk control measure in place to reduce the risk.
It is a common debate, the some of us considers verification as a risk control measure.
NO.
It needs to be a kind intelligence built in the device. Professional literature stipulates to consider design, protective measure, information on safety as a potential method in this order.
8., Residual severity evaluation. I have doubts if it can be different than the initial severity. I know, there is a tendency to say severity can be reduced, but clearly, as long as we have that behaviour in place in the device than how on earth can the severity to be reduced?
9., Residual probability, when the risk control measure in place.
10., Residual risk based on 8., and 9., .
11., Risk/Benefit analysis and decision of the control measure. (moreover there is a risk/benefit assessment for the device as a whole at the end looking into all identified hazards and work carried out as related to those)
12., Verification method and particular way of it as regard to the applied risk control measure to ensure its effectiveness.

I think more or less that's all.
 
Last edited:
R

regqual

#13
Medical Mobile App-classification

Read the Draft Guidance over three times, forward, backward including upside down with a bottle of wine and it still doesn't help me in deciding our approach. And of course following discussion at the Cove.

"Mobile apps that are an extension of one or more medical device(s) by connecting17 to such device(s) for purposes of controlling the device(s) or displaying, storing, analyzing, or transmitting patient-specific medical device data."

We have an app that connects with a standard USB cable from a Class II medical device to a bluetooth device and then transmits information obtained to your smartphone. The information is not diagnosed, controlled, analyzed. It is displayed in a more user-friendly tracking method (daily/weekly/monthly graph). The information is stored by the app only if the user chooses for a maximum of 90 days.

While I reserverdly agree this is a medical mobile app, my dilemna is what device classification? The argument mainly is this is low risk, user beneficial, non threatening application. We want in the hands of the user as soon as possible. If this is a medical mobile app, then it is a class 1 and not a class 2.

Thoughts-arguments-advice-guidance-comments-all helpful.
 

sagai

Quite Involved in Discussions
#14
Re: Medical Mobile App-classification

Let me roll back the time wheel and the wisdom of knowledge a wee.

Is the smartphone application a medical device as medical device is defined in the Food and Drug and Cosmetic Act?
Cheers!
 
I

isoalchemist

#15
Re: Medical Mobile App-classification

:2cents:I think you have a few issues treating this as an unregulated product. Even though you are not physically changing the data, you are presenting it is a way not initially intended by the medical device. Displaying it is not an issue, but storing for later analysis brings you back into the gray area again even if it is for a limited time. Lastly starting with a Class II device it makes it a little more diifcult to claim this is just a better way to track the information.

I think you crossed the line into a Medical Device.

The guidance makes this line unclear at times. Capturing data from a Class II device and sending it to a phycian for analysis makes an MDDS Class I, but if they are video conferening and get the data directly the MDDS follows the class of the unit sending it and the MDDS becomes Class II because it is considere "Real-Time"

Good Luck!:2cents:
 
R

regqual

#16
Thank you sagai, Thank you isoalchemsit! :agree1: Sometimes you just need a wee bit of perspective to get back on track. Was out doing my homework and I got it. Yup, also got confirmation from the FDA even in the process. We are going with MDDS. So much brainpower and wine spent on this decision.
 
Thread starter Similar threads Forum Replies Date
Ronen E FDA issues draft guidance on form & content of UDI Other US Medical Device Regulations 0
Ronen E FDA issues draft guidance on additive manufacturing Other US Medical Device Regulations 1
M FDA issues draft guidance to replace G95-1 (Application of ISO 10993) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
bio_subbu US FDA issues Draft Guidance on Premarket Approval (PMA) Acceptability Other US Medical Device Regulations 0
Ajit Basrur FDA issues draft guidance on nanotechnology US Food and Drug Administration (FDA) 1
bio_subbu FDA issues draft PMA (Pre-Market Approval) Clinical Study Guidance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
Ronen E FDA News FDA issues guidance on bench testing in premarket submissions Other US Medical Device Regulations 0
M FDA News The FDA Issues Final Rule on Medical Device Classification Procedures Medical Device and FDA Regulations and Standards News 0
Marc FDA Issues Guidance On Cybersecurity of Medical Devices Other Medical Device and Orthopedic Related Topics 1
bio_subbu FDA issues Guidance on UDI - Frequently Asked Questions, Vol. 1 Other US Medical Device Regulations 1
bio_subbu FDA issues Small Entity Compliance Guidance for UDI Other US Medical Device Regulations 1
Ajit Basrur FDA issues Guidance Document - Benefits-Risks Factors to consider for 510(K) US Food and Drug Administration (FDA) 1
Ajit Basrur FDA issues Final Guidance on Mobile Medical Apps (Applications) US Food and Drug Administration (FDA) 2
AnaMariaVR2 St. Jude Medical does the unthinkable: Warns of FDA warning letter before FDA issues US Food and Drug Administration (FDA) 5
Ajit Basrur FDA issues Warning Letters for misleading advertising of Lap-Band US Food and Drug Administration (FDA) 3
Ajit Basrur FDA issues 2 Guidance Documents on Nov 10, 2011 US Food and Drug Administration (FDA) 0
bio_subbu FDA Issues Guidance Document on Electrocardiograph (ECG) Electrodes Other US Medical Device Regulations 0
Ajit Basrur FDA Issues Final Regulation on Dental Amalgam 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
Ajit Basrur FDA Issues concerns on "e-cigarettes" US Food and Drug Administration (FDA) 4
D FDA Compliance - CMC quality and regulatory issues & cGMP - 21CFR parts 210 and 211 Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 5
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 0
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
D FDA Information - Revising the Instructions for Use US Food and Drug Administration (FDA) 0
S Transitional Adolescent A and B - "CDRH PREMARKET REVIEW SUBMISSION COVER SHEET FORM FDA 3514" Medical Device and FDA Regulations and Standards News 1
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
P MSDS for MVQ FDA White, Vinyl Methyl Silicone Rubber EU Medical Device Regulations 4
S Manufacturing Process FDA FOIA Medical Device and FDA Regulations and Standards News 3
S Manufacturing Process FDA FOIA US Food and Drug Administration (FDA) 4
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Supplier requirements - Major supplier is a Non-Profit registered with ICCBBA (FDA UDI) Supply Chain Security Management Systems 12
C RA (Regulatory Assurance) Training (FDA) looking for resources Training - Internal, External, Online and Distance Learning 5
E FDA 513(g) Cover Letter US Food and Drug Administration (FDA) 5
S Records - Do's and don't' of record entries (FDA - 21 CFR 820) Records and Data - Quality, Legal and Other Evidence 13
B New Facility register with FDA Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J FDA regulation on decorative contact lenses Medical Device and FDA Regulations and Standards News 5
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F FDA classification for a mobile app Medical Information Technology, Medical Software and Health Informatics 3
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
O Any info on release date of FDA “Computer Software Assurance for Manufacturing and Quality System Software” document? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
P Writing a presubmission to the FDA prior to the De novo submission US Food and Drug Administration (FDA) 4
K FDA - Can we have more than 1 Initial Importer Medical Device and FDA Regulations and Standards News 4
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
R Importers - For the FDA Registration/ Listing, is the CMO the importer? Medical Device and FDA Regulations and Standards News 6
Z Iterative development and FDA change requests IEC 62304 - Medical Device Software Life Cycle Processes 12
P FDA Approved Product Contact Parts ISO 13485:2016 - Medical Device Quality Management Systems 6
O Clarifying FDA definition of "finished device" and “capable of functioning” 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
JoCam FDA Registration for Sub-contract manufacturers Medical Device and FDA Regulations and Standards News 2