(I thought this news was posted in February when the news got out, but I could not find it here, thanks Sidney)
The big question is 'when?' I've not heard any plans / timelines.
Oh, there is some plans: FDA Considers Shift on Device Quality System Regulations | RAPS
There's even a formal, proposed rulemaking at (Reginfo is not working for me at the moment:
Harmonizing and Modernizing Regulation of Medical Device Quality Systems rule
Spring 2018 Unified Agenda of Regulatory and Deregulatory Actions
Sam Lazzara
Trusted Information Resource
#4
If the requirements for signatures go away (since 13485 does not require them) then 21 CFR Part 11 requirements for electronic signatures will not apply.
I am wondering if FDA will apply 13485 with some additional requirements like signatures and also include the exceptions that are currently in 820.180(c). I find it hard to believe 13485 will be applied verbatim without modification.
Sec. 820.180 General requirements.
All records required by this part shall be maintained at the manufacturing establishment or other location that is reasonably accessible to responsible officials of the manufacturer and to employees of FDA designated to perform inspections. Such records, including those not stored at the inspected establishment, shall be made readily available for review and copying by FDA employee(s). Such records shall be legible and shall be stored to minimize deterioration and to prevent loss. Those records stored in automated data processing systems shall be backed up.
(a)Confidentiality. Records deemed confidential by the manufacturer may be marked to aid FDA in determining whether information may be disclosed under the public information regulation in part 20 of this chapter.
(b)Record retention period. All records required by this part shall be retained for a period of time equivalent to the design and expected life of the device, but in no case less than 2 years from the date of release for commercial distribution by the manufacturer.
(c)Exceptions. This section does not apply to the reports required by 820.20(c) Management review, 820.22 Quality audits, and supplier audit reports used to meet the requirements of 820.50(a) Evaluation of suppliers, contractors, and consultants, but does apply to procedures established under these provisions. Upon request of a designated employee of FDA, an employee in management with executive responsibility shall certify in writing that the management reviews and quality audits required under this part, and supplier audits where applicable, have been performed and documented, the dates on which they were performed, and that any required corrective action has been undertaken.
ScottK
Not out of the crisis
Staff member
Super Moderator
#5
The way I'm reading this is that 21 CFR 820 will be modified to better match 13485 and reference it, but will still be the law.
Which doesn't affect part 11 - I doubt that's going anywhere.
Sidney Vianna
Post Responsibly
Staff member
Admin
#7
Would this bring Certification Bodies/Registrars into the mix or would FDA still do its own thing?
Regulatory bodies can never relinquish their responsibilities to organizations they have zero control over, as the law would not permit that.
Nevertheless, as such oversight bodies also struggle for resources to monitor the whole (immense) supply chain, I can envision entities such as the FDA using market driven data (such as ISO 13485 certification status AND audit results) when using a risk-based approach to steer their inspection resources.
Another aspect of what I had envisioned when I started the poll/thread Should Regulatory Agencies require Accredited Certification in their Respective Areas
Mark Meer
#8
How does this move affect the FDA's commitment to MDSAP?
Is it likely a dead/redundant endeavour? ...or might (mandatory) MDSAP be the means by which the FDA transitions to an ISO 13485 framework?
Marcelo
#9
It's easy to see the MDSAP is one of the reasons FDA is considering using 13485, and it's also being brewed as thew "future" of medical devices QMS auditing.
Thread starter Similar threads Forum Replies Date
bio_subbu US FDA plans to launch new website US Food and Drug Administration (FDA) 4
J Sub-supplier change from manual to automated process - same specs - Report to FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
C FDA Establishment registration - Buying some medical devices from another manufacturer Medical Device and FDA Regulations and Standards News 2
I How to classify a medical device based on FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
dinaroxentool Question about FDA Classification of a Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S The US FDA requirements on Disposal of a medical device US Food and Drug Administration (FDA) 1
S FDA Requirements for Medical Device Label Reconciliation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S What is considered a "core algorithm"? (From an FDA guidance document) Medical Information Technology, Medical Software and Health Informatics 4
B FDA-Medical Device Reporting (MDR )procedure compliant with 21CFR section 803 US Food and Drug Administration (FDA) 0
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D ISO 13485, FDA 21 CFR 820 and Auditing the Accounting Department ISO 13485:2016 - Medical Device Quality Management Systems 5
S FDA's E submitter - For clinical electronic thermometer Medical Device and FDA Regulations and Standards News 6
J FDA wants electrical safety testing on battery powered medical device US Food and Drug Administration (FDA) 7
A FDA and NB audit of Engineering Drawings in DHF and DMR. Medical Device and FDA Regulations and Standards News 1
D FDA De Novo Cover Letter - What is expected US Food and Drug Administration (FDA) 3
Ed Panek 21 CFR Part 820 - FDA Label Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J Mislabeling - Consider this an FDA notified recall? CFR 806.10 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
O Hospital-Developed, In-House Device - FDA Clearance? Other US Medical Device Regulations 19
D FDA Approved Class II Medical Device? Other US Medical Device Regulations 11
R When does the FDA consider a component a medical device? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
N Risk Management besides mandated FDA requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Informational US FDA – Ear, Nose, and Throat Devices; Classification of the Self-Fitting Air-Conduction Hearing Aid Medical Device and FDA Regulations and Standards News 0
L UDI Requirments Gap Analysis - EU MDR vs FDA CFR EU Medical Device Regulations 8
M Informational US FDA – Modifications to the List of Recognized Standards, Recognition List Number: 052 Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Draft Guidance – Breast Implants – Certain Labeling Recommendations to Improve Patient Communication Medical Device and FDA Regulations and Standards News 0
M Looking for third party reviewers for FDA submissions US Food and Drug Administration (FDA) 2
M Informational Work in progress at the FDA for biological evaluation – Color Hazard and RISk calculator (CHRIS) Medical Device and FDA Regulations and Standards News 0
M Informational US FDA – CDRH Proposed Guidances for Fiscal Year 2020 (FY 2020) Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Final Guidance – Coronary, Peripheral, and Neurovascular Guidewires – Performance Tests and Recommended Labeling Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Final Guidance – Intravascular Catheters, Wires, and Delivery Systems with Lubricious Coatings – Labeling Considerations Medical Device and FDA Regulations and Standards News 0
T Clinical trial - Medical device product not cleared (without FDA approval) in a drug trial Other US Medical Device Regulations 9
W How Often Does FDA State Consultant Recommended in a Pharma WL? US Food and Drug Administration (FDA) 1
M Informational US FDA – URGENT/11 Cybersecurity Vulnerabilities in a Widely-Used Third-Party Software Component May Introduce Risks During Use of Certain Medical Dev Medical Device and FDA Regulations and Standards News 0
M Informational US FDA paper – Epidemiological Evidence on the Adverse Health Effects Reports in Relation to Mercury from Dental Amalgam: Systematic Literature Review Medical Device and FDA Regulations and Standards News 0
M Informational US FDA paper – Biological Responses to Metal Implants Medical Device and FDA Regulations and Standards News 0
V Presubmission FDA for 510k US Food and Drug Administration (FDA) 4
M Informational US FDA Draft Guidance – Clinical Decision Support Software Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Final and update guidances – Software Medical Device and FDA Regulations and Standards News 0
M Informational From Medtech Insight – QSR/ISO 13485 Harmonization Update: FDA Enforcement Discretion Likely When New Rule Stands Up; Draft Reg Coming By Year’s End; Medical Device and FDA Regulations and Standards News 0
F FDA-registration and inspection cost 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Informational US FDA Draft Guidance – Providing Regulatory Submissions for Medical Devices in Electronic Format – Submissions Under Section 745A(b) of the Federal F Medical Device and FDA Regulations and Standards News 0
D Framing FDA Questions - Validation US Food and Drug Administration (FDA) 1
M Informational US FDA CDRH’s Health of Women Program Strategic Plan Medical Device and FDA Regulations and Standards News 0
M Informational US FDA – Frequently Asked Questions (FAQs) About the Accreditation Scheme for Conformity Assessment (ASCA) Pilot Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Draft Guidance – Patient Engagement in Design and Conduct of Medical Device Clinical Investigations Medical Device and FDA Regulations and Standards News 0
M Informational Several US FDA draft guidances, including some specific device guidances for the Safety and Performance Based Pathway Medical Device and FDA Regulations and Standards News 0
MDD_QNA How to deal with FDA after not reporting anything for several years Other US Medical Device Regulations 14
M Informational Device Shortages Update – The US FDA Announces Two New Innovation Challenges on Device Sterilization Medical Device and FDA Regulations and Standards News 0
M Informational FDA draft guidance – Safer Technologies Program for Medical Devices Medical Device and FDA Regulations and Standards News 0
K Mammography Machine Bench Testing - FDA 510(K) US Food and Drug Administration (FDA) 5
Similar threads
Top Bottom