FDA Position about Stainless Steel & Ra?
cisko
#11
What FDA is really saying is that it is up to you to verify that the stainless steel that you choose to use has no adverse effect on safety, quality, identity, potency or purity for the product that you manufacture using your materials, machines, and methods. Testing to verify non-impact should be part of your process validation requirements.
Yep ok
& what I also wanted to check it's why we can find on many websites "stainless steel is GRAS" but it's not mentioned on FDA website :frust:
Ronen E
Problem Solver
Moderator
#12
Yep ok
a. The FDA is in charge of several fields - pharmaceuticals, medical devices etc. Not necessarily the same statements apply to all fields.
b. "Generally Regarded As Safe" - Please note the term "generally". This may very well be the case in many situations, but as long as it isn't so in *all*, I wouldn't expect the FDA to make the sort of declarations you're looking for.
c. Further, "Generally Regarded As Safe" - safe for what? The FDA typically works on an intended-use basis, so as long as this is kept vague, I wouldn't expect the FDA to be very clear.
Cheers,
Ronen.
L
Lugog
#13
Hi Ronen, thanks for your useful answer!
The flowchart is interesting but in date of 1995 :(
Currently we ask the EN 10204 2.1 or 3.1 certificat for alloy 316/316L, of course it's enough but we are looking further that
Any other ideas or opinions ? Also about the rugosity coefficient (Ra)
Cheers,
Cisko
Hi all,
Coming back on this one, is the 3.1 certificate or EN110204 2.1 really enough thought?
I am working on a similar definition for our plant (pharma industry) and am being asked again and again by my management if this is really sufficient.
What was your position in the end?
Regards
Lugog
cisko
#14
Hi Lugog,
We decided to ask a 3.1 certificat for all metals/polymer & 2.1 for ceramic & glass materials
Is it enought ? actually It depends of far do you trust your supplier :caution:
We know that chineses certificates are often suspects .. so we might use a PMI analyser for metals products & conducts internal tests on polymerics materials...
v9991
#15
1) asking for certificate is not adequate...especially for product contact surfaces, there ought to be a Inhouse test. ( i believe that even EN 10204 also outlines the need for "specific inspection"and "manufacturers" validation.
and its certainly not for your to depend on supplier...rather what are the processes in place to rely on supplier...and in some circumstances, limitations of performing certain tests!!!
2) as has been said, the suitability/compatibility has to be established through product specific challenges/studies.(which is part of your validations of equipment or process as appropriate)
this could b some times as simple identification of the specified grade of steel toverify the supplier certificate or in other cases, the hold time study/compatibility study on specific product.
L
Lugog
#16
Hi V9991
I'm interested in which kind of testing you'd suggest to perform.
We could not perform test for every material we are receiving even if they have product contact. The approach we took is to perform testing for polymeric components (leachability and extractability) as there is a confirmed risk.
For metal components we check the supplier qualification and material certificate for each incoming part. No further testing afterwards.
Would you have a reference to recommand?
JeantheBigone
#17
You might want to consider verifying your supplier cert at least occasionally. And if you are getting material that originates in China, I'd verify each and every batch.
Material downgrades on stainless are rampant, not only in China. You specify 316 but you get 304, they figure no one will notice.
I wonder if the surface roughness requirement on stainless is a carryover from food production? :cool:
http://www.foodquality.com/details/article/813533/Go_With_the_Flow.html?tzcheck=1
v9991
#18
:yes:
although there are many references, its always required to confirm/sync with the manufacturer;s procedure. (cause, even if either of you are not following standard procedure, the exercise would not serve purpose)
anyways, here's few
ASTM A493-95 :- Corrosion Resistance
Molybdenum test
http://www.fumeanddust.com.au/manuals/clinox/Easy Test_presentazione ENG.pdf
http://www.koslow.com/metal_test_kits/1542c.html
http://www.eng-tips.com/viewthread.cfm?qid=78040
:2cents:
though a little far fetched,, here's one reference where compatiblity is demonstrated in this protocol...
http://www.xcellerex.com/pdf/XDA-Biocompatibity.pdf
Michael Malis Managing Use Error Risk - FDA Position: Use Error is Manufacturer's responsilbility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 23
D FDA Guidance on Computer Software Assurance versus 21 CFR Part 11 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
G Does FDA allows remote approvals of quality documentation. Is there any specific guidance on signing any quality records remotely? Document Control Systems, Procedures, Forms and Templates 1
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 1
S Transitional Adolescent A and B - "CDRH PREMARKET REVIEW SUBMISSION COVER SHEET FORM FDA 3514" Medical Device and FDA Regulations and Standards News 1
P MSDS for MVQ FDA White, Vinyl Methyl Silicone Rubber EU Medical Device Regulations 4
C RA (Regulatory Assurance) Training (FDA) looking for resources Training - Internal, External, Online and Distance Learning 5
E FDA 513(g) Cover Letter US Food and Drug Administration (FDA) 5
S Records - Do's and don't' of record entries (FDA - 21 CFR 820) Records and Data - Quality, Legal and Other Evidence 13
B New Facility register with FDA Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
J FDA regulation on decorative contact lenses Medical Device and FDA Regulations and Standards News 5
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F FDA classification for a mobile app Medical Information Technology, Medical Software and Health Informatics 3
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
O Any info on release date of FDA “Computer Software Assurance for Manufacturing and Quality System Software” document? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
P Writing a presubmission to the FDA prior to the De novo submission US Food and Drug Administration (FDA) 4
K FDA - Can we have more than 1 Initial Importer Medical Device and FDA Regulations and Standards News 4
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
R Importers - For the FDA Registration/ Listing, is the CMO the importer? Medical Device and FDA Regulations and Standards News 6
Z Iterative development and FDA change requests IEC 62304 - Medical Device Software Life Cycle Processes 12
P FDA Approved Product Contact Parts ISO 13485:2016 - Medical Device Quality Management Systems 6
O Clarifying FDA definition of "finished device" and “capable of functioning” 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
JoCam FDA Registration for Sub-contract manufacturers Medical Device and FDA Regulations and Standards News 2
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
E Contract manufacturer FDA requirements foreign company US Food and Drug Administration (FDA) 6
P Choice of FDA reviewer US Food and Drug Administration (FDA) 6
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
T Definition Human Use (FDA) Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 7
K FDA 510k electrosurgery 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J Leveraging FDA 510k Clearance for International Registrations Other Medical Device Regulations World-Wide 2
shimonv FDA News FDA guidance on Multiple Function Device Products (8/2020) Other US Medical Device Regulations 1
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
J FDA notification of address change US Food and Drug Administration (FDA) 2
N Usability testing required for FDA IDE (investigational device exemption)? Human Factors and Ergonomics in Engineering 8
A FDA Class Classification for a cabinet 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
L FDA Biocompatibility Requirements - Transitory Contact Other US Medical Device Regulations 1
Edward Reesor FDA DM and /or Class I Life Saving US Food and Drug Administration (FDA) 0
Watchcat FDA will not tolerate fraud…meaning what, exactly? Other US Medical Device Regulations 8
K FDA Premarket Cybersecurity Guidance - 4 questions Other US Medical Device Regulations 5