FDA Premarket Submissions for Embedded Software - Level of Concern

kreid

Involved In Discussions
#1
Hello,

I have been looking into the level of concern for a piece of embedded software under development.
Question 4 of Table 1 of the FDA Premarket Submissions for Software guide, asks:
Prior to mitigation of hazards, could a failure of the Software Device result in death or serious injury, either to a patient or to a user of the device?
As part of the design of the device a Preliminary Hazard Analysis was carried out which showed that a software failure might lead to a serious injury. As a result of the PHA the device is being designed so that there are hardware safe-guards to prevent any software failiures leading to serious injury.
My question is, does the "Prior to mitigation" part of Q4 mean that if the design is created (based on the PHA) that prevents software failures causing serious hazards, does this constitute a risk mitigation, and hence still qualify it as software of a major concern or can we say that the design is such that software failures cannot contribute to serious harm.

In effect, at what point in the design and development lifecycle should Q4 be asked?

Thanks
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
While this may seem to be skirting the question a bit, IEC 62304 does make this rather clear:

If the RISK of death or SERIOUS INJURY arising from a software failure is subsequently reduced to an acceptable level (as defined by ISO 14971) by a hardware RISK CONTROL measure, either by reducing the consequences of the failure or by reducing the probability of death or SERIOUS INJURY arising from that failure, the software safety classification may be reduced from C to B; and if the RISK of non-SERIOUS INJURY arising from a software failure is similarly reduced to an acceptable level by a hardware RISK CONTROL measure, the software safety classification may be reduced from B to A.

Given that and the fact that 62304 is a recognized consensus standard by the FDA, I think you'd be in a pretty defensible position if you did so.
 

kreid

Involved In Discussions
#3
Hi Yodon,
Thanks for your response.

My problem is that for the FDA if the software failure could cause a seriuos injury pre-mitigation then it is classsed as a major level of concern and this has an impact on registration and other activities.

The excerpt you quote from 62304 is a common approach in safety critcal software systems but the FDA approach would seem to penalise someone following the good engineering practice of changing a design to limit the effects of software failure.

Kreid
 

kreid

Involved In Discussions
#4
As an aside the passage you quote from 62304 was removed in the last update.

It was replaced by a flow chart that I think is very confusing, and makes it difficult to conclude your software is anything but Class A!
 

yodon

Staff member
Super Moderator
#5
As an aside the passage you quote from 62304 was removed in the last update.

It was replaced by a flow chart that I think is very confusing, and makes it difficult to conclude your software is anything but Class A!
I think it's generally the same; they just replaced hardware control with control external to the software system.
 
Thread starter Similar threads Forum Replies Date
M Informational US FDA Final guidance – Metal Expandable Biliary Stents – Premarket Notification (510(k)) Submissions Medical Device and FDA Regulations and Standards News 0
Ronen E FDA News FDA issues guidance on bench testing in premarket submissions Other US Medical Device Regulations 0
S Is Software code documentation required for FDA premarket submissions? IEC 62304 - Medical Device Software Life Cycle Processes 2
M Medical Device News FDA Releases Draft Recommendations on Premarket Submissions for Management of Cybersecurity in Medical Devices Other US Medical Device Regulations 0
M Medical Device News FDA news 14-09-18 - Use of Voluntary Consensus Standards in Premarket Submissions Other US Medical Device Regulations 3
S Transitional Adolescent A and B - "CDRH PREMARKET REVIEW SUBMISSION COVER SHEET FORM FDA 3514" Medical Device and FDA Regulations and Standards News 1
K FDA Premarket Cybersecurity Guidance - 4 questions Other US Medical Device Regulations 5
M Informational US FDA Final Guidance – Special Premarket Notification [510(k)] Pathway Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Final Guidance – Consideration of Uncertainty in Making Benefit-Risk Determinations in Medical Device Premarket Approvals, De Novo Classificati Medical Device and FDA Regulations and Standards News 0
M Medical Device News FDA news -11-09-18 - Review of Cybersecurity into Premarket Review Other US Medical Device Regulations 0
P FDA - Evaluating Substantial Equivalence in Premarket Notifications 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S FDA CDRH Premarket Review Submission Cover Sheet - adding more products 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
bio_subbu US FDA issues Draft Guidance on Premarket Approval (PMA) Acceptability Other US Medical Device Regulations 0
R FDA CDRH Premarket Review Requirements - Submission Cover Sheet 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Michael Malis FDA issued modifications to the list of standards used in Premarket Review Process US Food and Drug Administration (FDA) 0
D FDA Guidance on Computer Software Assurance versus 21 CFR Part 11 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
G Does FDA allows remote approvals of quality documentation. Is there any specific guidance on signing any quality records remotely? Document Control Systems, Procedures, Forms and Templates 1
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 1
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
D FDA Information - Revising the Instructions for Use US Food and Drug Administration (FDA) 0
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
P MSDS for MVQ FDA White, Vinyl Methyl Silicone Rubber EU Medical Device Regulations 4
S Manufacturing Process FDA FOIA Medical Device and FDA Regulations and Standards News 3
S Manufacturing Process FDA FOIA US Food and Drug Administration (FDA) 4
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Supplier requirements - Major supplier is a Non-Profit registered with ICCBBA (FDA UDI) Supply Chain Security Management Systems 12
C RA (Regulatory Assurance) Training (FDA) looking for resources Training - Internal, External, Online and Distance Learning 5
E FDA 513(g) Cover Letter US Food and Drug Administration (FDA) 5
S Records - Do's and don't' of record entries (FDA - 21 CFR 820) Records and Data - Quality, Legal and Other Evidence 13
B New Facility register with FDA Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J FDA regulation on decorative contact lenses Medical Device and FDA Regulations and Standards News 5
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F FDA classification for a mobile app Medical Information Technology, Medical Software and Health Informatics 3
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
O Any info on release date of FDA “Computer Software Assurance for Manufacturing and Quality System Software” document? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
P Writing a presubmission to the FDA prior to the De novo submission US Food and Drug Administration (FDA) 4
K FDA - Can we have more than 1 Initial Importer Medical Device and FDA Regulations and Standards News 4
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
R Importers - For the FDA Registration/ Listing, is the CMO the importer? Medical Device and FDA Regulations and Standards News 6
Z Iterative development and FDA change requests IEC 62304 - Medical Device Software Life Cycle Processes 13
P FDA Approved Product Contact Parts ISO 13485:2016 - Medical Device Quality Management Systems 6
O Clarifying FDA definition of "finished device" and “capable of functioning” 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
JoCam FDA Registration for Sub-contract manufacturers Medical Device and FDA Regulations and Standards News 2
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Contract manufacturer FDA requirements foreign company US Food and Drug Administration (FDA) 6
P Choice of FDA reviewer US Food and Drug Administration (FDA) 6

Similar threads

Top Bottom