Hi Jimbo,
Did you or NTS test all the relevant cases from 60601-1 subclause 8.4? For examble the patient auxiliary current? I have seen some cases when this is ignored totally (for internally powered devices) even though the standard clearly says the compliance is checked by mesurement (according to 8.7.4).
br Aki
Did you or NTS test all the relevant cases from 60601-1 subclause 8.4? For examble the patient auxiliary current? I have seen some cases when this is ignored totally (for internally powered devices) even though the standard clearly says the compliance is checked by mesurement (according to 8.7.4).
br Aki