Filling totes and pails - when is a calibrated scale / flowmeter required in ISO 9001

Dan M

Involved In Discussions
#1
Hello all. We are a chemical products manufacturer who operates some warehouses (distribution centers). Our warehouses will routinely repack our products into smaller containers, such as 330-gallon IBC containers, drums, or pails. Some of the warehouses do not have the proper measuring equipment and rely on visually filling to a known volume. For example, if selling 200 gallons, they will fill the 330-gallon IBC to the 200 gallon graduation mark stamped on the plastic bladder.

There is another scenario where they may fill 5 gallon pails, visually approximating the volume by filling to the top of the container. In this instance, the product is sold as one (1) ea, not billed by the gallon.

The clause of ISO 9001 I am getting hung up on:

7.1.5.2 Measurement Traceability

When measurement traceability is a requirement, or is considered by the organization to be an essential part of providing confidence in the validity of measurement results, measuring equipment shall be:
a) calibrated or verified, or both, at specified intervals


So my questions are:
(1) Is there any precedent to use the visual markings on 330-gallon IBC to measure the volume filled and bill the customer based on this volume?
(2) If is acceptable under ISO 9001 to fill containers such as 5-gallons pails or 55-gallons drums to full (visually) if the containers are sold as one unit (one each) and we are not billing by pounds/gallons.


My rhetorical question:
So who decides when measurement traceability is a requirement if our customer has not defined it as a requirement? What if our organization does not consider measurement traceability to be an essential part of providing confidence in the validity of measurement results when it comes to the examples I stated above?
 
Elsmar Forum Sponsor

Jen Kirley

Quality and Auditing Expert
Leader
Admin
#2
I once had a tape measure that measured about 1/8 inch short in every foot. Really! I wish I kept it... Measuring devices sold for cooking are known to vary somewhat in their accuracy. I would therefore not just trust markings on bladders or pails, and would wonder too if there is a fold in the bottom of that bladder that would affect its fill accuracy unless that was addressed during its insertion or before filling.

If you are selling a particular quantity, ISO is asking you to plan for a way to ensure you have controls in place to ensure your containers are holding what you claim they are.

My clients have almost all done this by weight, because scales are easy to calibrate and can be installed in conveyors. Fill height to container can also be done if you have made sure the containers and their fill levels are, in fact accurate. I suggest you can do that yourself but you will need to document your methods and keep a record of having performed them in order to show you have youor controls in place.

I hope this helps.
 

John Predmore

Trusted Information Resource
#3
In a plant quality context, the word calibration includes checking or adjusting a measurement device to a national standard. ISO 9001 7.1.5.2 allows for verification or calibration, or both. The answer to your question is your organization decides how and how often to assure fitness of purpose for the measuring device.

In my factory, I do a sort of mental FMEA risk assessment to decide between outside calibration service and internal verification...
Of all potential product uses for a device, pick the one demanding the highest degree of accuracy and continue.
Severity: What would be the impact if the reading were high or low by 0.5%? 1%? 5%? How much is unacceptable?
Detection: How could it happen the device if off by that much? Would it happen gradually or suddenly? How soon before someone notices?
Occurrence: How frequently would you imagine it might happen that the device is off by that much? (This guides you to set verification interval.)
The answers to these questions help me make a risk-based decision.
 

dwperron

Trusted Information Resource
#4
My rhetorical question:
So who decides when measurement traceability is a requirement if our customer has not defined it as a requirement? What if our organization does not consider measurement traceability to be an essential part of providing confidence in the validity of measurement results when it comes to the examples I stated above?


You are moving into the specialized territory of Legal Metrology. Those measurement requirements are determined by your local Weights and Measures officials, which can operate at the local, county, or state level. I would recommend that you talk with your local Weights and Measures people and find out what they require.

At the federal level, the guiding document for "Legal for Trade" metrology is from NIST:

NIST Handbook 44 - Current Edition
 
Thread starter Similar threads Forum Replies Date
S New to FAIR, need help in filling it out AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
P Drug packaging / filling operations Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 8
S Any experience on filling form, FDA 3654 for the DICOM standard? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
G Filling out inspection report using calibrated pins Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
K Production Batch Documentation based on Filling Manufacturing and Related Processes 5
G Are there any requirements to filling out AS 9102 Forms AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
B Assistance in filling out an Appearance Approval Report for PPAP Quality Assurance and Compliance Software Tools and Solutions 3
L IQ and OQ Validation for the Filling Machine Capability, Accuracy and Stability - Processes, Machines, etc. 5
C Best Material for Casting a Pig Mold Tub used for filling up Liquid Lead Manufacturing and Related Processes 4
S Filling Up Porosity - Grey Cast Iron Casting Manufacturing and Related Processes 4
R Filling in electronic forms for medical devices Other Medical Device and Orthopedic Related Topics 4
B Filling Out FDA Form 3654 when Testing is Incomplete 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
P Determining Sample Size - Filling a bag with three different types of candy Inspection, Prints (Drawings), Testing, Sampling and Related Topics 31
Q FMEA Form - Clarification on correct way of filling up (key/fmea date, rev) FMEA and Control Plans 1
M FDA - "Net Quantity of Contents" - Manufacturing Filling Process Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 12
Q Is anyone using FormDocs (Electronic Form-Filling Software)? Document Control Systems, Procedures, Forms and Templates 12
M Digital Filling System - What do I need to Validate Here? Software Quality Assurance 1
A In-process check frequency - Class I devices on the filling and labeling lines ISO 13485:2016 - Medical Device Quality Management Systems 2