Flame Rating Class for Plastic Materials in Enclosure of Medical Equipment
temujin
Dear Forum,
I am in the process of building up the Technical File for the CE-Marking of a medical device.
Our Notified Body has required me to provide certificates/information on the Flame Rating (UL Rating) of all plastic material in the enclosure.
However, reading 60601-1, I cannot find any explicit requirements regarding Flame Rating for enclosures. My contact at the NB is also not sure where the requirement originates, he simply wants to have it.
Does anyone have experience with this requirement+
regards
t.
Sunset - 2011
Re: Flame Class for Enclosure of Medical Equipment
Hi temujin,
your enclosure is a critical safety component.
If I remember right, materials for medical devices must have a special flame rate class according to the standard ANSI/UL 94. I'm sorry, I actually forgot the mimimum requirement.
But you can go on the UL-website and search under certification for plastic material or plastic material companies or the UL code QMFZ2, if you click on one certificate, you'll find on top a link to general requirements for plastic components. May you find more information there.
The flame class also gives the maximum temperature for the material and maximum temperature requirements are under point 42 in the IEC 60601-1, I think. Please check.
I hope that helps a bit!
Anyone else if I'm wrong, please correct me.
Cheers,
Sunset
temujin
Thanks Sunset,
As a European manufacturer we are not required to follow the ANSI/UL standard. Unfortunately I have only the 3rd edition of 60601-1. Does anyone know where to find "temperature requirements are under point 42 in the IEC 60601-1" in the 3rd edition?
regards
t.
Sunset - 2011
HI temujin,
I'm sorry, I haven't got the third Edition, but if you follow through the standard, there will be a point with temperature requirements. There is a table which shows maximum requirements for components in medical devices.
Point 42 is of the 2nd Edition.
May someone else can help?
Sunset
Marcelo
There´s a new concept in the third edition called "fire enclosure", in clause 11.3. This is an alternative way to deal with the hazards described in 13.1.2 (emissions, deformation of enclosures or exceding maximum temperature). Keep in mind that just the flammabiulity classification (FV-1, FV-2) is not enough, you need to comply with all requirements of the clause, for example, the baffle construction if the equipment have openings as seen in figures 38 and 39. As you said that
it seems that your NB does not know that (not a surprise, but still a stupid request as he seems to not know what he´s talking about).
temujin
Thanks to all,
I am aware of the section on FIRE ENCLOSURE. However, what happen if the FIRE ENCLOSURE itself is covered by a second plastic ENCLOSURE? Would there be any specific requirements regarding the material of this secondary ENCLOSURE?
regards
t.
Marcelo
If there were a second plastic ENCLOSURE, then the first ENCLOSURE would not fit the definition of ENCLOSURE anymore so it would not be a fire ENCLOSURE anymore! An ENCLOSURE is an external part. In this case, the plastic enclosure would have to comply with all requirements of the clause for fire ENCLOSURE.
Steve McQuality
Quality Engineer
Hi Temujin:
Just got one of the Medical Industry Trade magazines today and there is an article in the most recent MD&DI that talks about plastic molding. I remembered reading your post this morning and thought I'd send this tid-bit along to you. Within the article, they talk about the UL Fire Rating. It might be worth a quick perusal and may help define the requirements as they stand alone. As far as their relationship to IEC 60601-1, I can't help you there...
Here's the link - http://www.devicelink.com/mddi/archive/08/07/006.html
Good luck,
-Steve
( I just realized this was my 100th Post - yea me!)
jmalm
In terms of "enclosure", does anyone know if the flammability ratings extend for cables and applied parts as well?
Example, say I have an esophageal temperature probe and a cable that connects to a measurement device. Do the external surfaces of the probe and cable need to have 94V-2?
I know many applied parts carrying electrical signals use Pebax as the outer shell (eg some ablation catheters); Pebax is rated 94-HB so wouldn't seem to be covered by 60601's definition of "enclosure" for the purposes of applying flammability ratings (unless the Pebax they use has enhanced flame ratings).
noorain
I'm aware that its to late to answer this, but assuming that other people might also be looking for the same answer, the auditor is looking for FV-2/ FV-1 flammability requirements.
D Flame Rating for Internal Plastic Support inside of a Fire Enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E IEC & UL 60601-1 Plastics Flammability - Guidance on flame-rated plastics usage IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
N How can you tell if Flame Retardant "kicks off" in plastic? Manufacturing and Related Processes 4
I How to provide Lower Flame limit (LFL) and Delta Hv(cal/g) Lean in Manufacturing and Service Industries 3
G Materials Requirements - Automotive industry - Smoke, Flame and Toxicity APQP and PPAP 3
M Reduce occurrence rating based on the PMS data and customer complaint data ISO 14971 - Medical Device Risk Management 2
W FMEA - Current control and occurrence rating FMEA and Control Plans 3
D FAA inspection at my repair station - Rating Check Federal Aviation Administration (FAA) Standards and Requirements 0
S Complexity Rating - CB adding another audit day for "high complexity" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
P Design FMEA - Detection Rating criteria ISO 14971 - Medical Device Risk Management 3
S Power rating for medical devices in the US - Radiotherapy devices US Food and Drug Administration (FDA) 0
P Detection rating for a user (surgeon) related failure mode in DFMEA APQP and PPAP 3
M Should Potential Customer Complaint Outcome Define Registrar NC Rating? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
R Determination of IP (International Protection Marking) Rating IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Supplier 8D Report-Rating Procedure Problem Solving, Root Cause Fault and Failure Analysis 10
A Supplier Rating from SAP QM Module Data Supplier Quality Assurance and other Supplier Issues 2
B Third Party Service Provider Risk Rating Spreadsheet Risk Management Principles and Generic Guidelines 1
S FMEA Visual Inspection Detection Rating FMEA and Control Plans 5
V Where to find Red/White Rust Rating Requirements? APQP and PPAP 4
L Flammability Rating - is FT4 a better rating than FT1? Other US Medical Device Regulations 2
B FMEA Ranking Rating Best Practices FMEA and Control Plans 10
P Occurrence Rating in AIAG FMEA Manual 4th Edition FMEA and Control Plans 5
R How much Protective Current Rating of Building Branch Circuit should be considered? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W Severity Rating CORRECTIVE ACTIONS - Opinions Wanted Please Nonconformance and Corrective Action 15
P Risk Rating approaches (OHSAS 18001 - For Office Environments) Occupational Health & Safety Management Standards 6
S Smartphone based software with automatic "rating" from remote servers Medical Information Technology, Medical Software and Health Informatics 4
T Enclosure Flammability Rating of Molded-In Custom Color Liquid Concentrates IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J Determining Fuse Rating and Labeling - Medical Device Crash Cart IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M FMEA Occurrence Rating for Customer Claims (Complaints) FMEA and Control Plans 2
K Supplier Performance and Supplier Rating Process - AS9100 Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
J How to Increase Customer Satisfaction Rating? Customer Complaints 20
M Supplier Rating Form with Quality, Cost, Delivery wanted Document Control Systems, Procedures, Forms and Templates 4
C Supplier Performance Index (Spi) - Supplier Rating Supplier Quality Assurance and other Supplier Issues 10
S How to determine what rating an article falls under - 14 CFR Part 145 Federal Aviation Administration (FAA) Standards and Requirements 5
L Gap in FMEA Occurrence Rating Standard FMEA and Control Plans 2
S About the rating of fuse - Can I employ 2 fuses of different rating in my equipment? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A PFMEA (Process FMEA) Detection Rating - Actions to Reduce RPN FMEA and Control Plans 3
I Service Supplier Rating where Objective Pass/Fail Data is not Available Service Industry Specific Topics 7
V In Design FMEA, how to give Occurrence Rating? FMEA and Control Plans 4
P Detection Rating in PFMEA and Process Controls FMEA and Control Plans 6
M FMEA Rating Scales for the Insurance Industry FMEA and Control Plans 4
Q FMEA: Visual Detection Rating IATF 16949 - Automotive Quality Systems Standard 25
N All-or-Nothing Rating System - Internal Audit Checklist - ISO 9001 Internal Auditing 5
Chennaiite FMEA Severity Rating for Potential Manufacturing Effect mentioned in the FMEA manual FMEA and Control Plans 9
R Supplier Rating System Ideas and Suggestions Quality Manager and Management Related Issues 3
R Supplier Quality Rating Scorecard Template (excel spreadsheet) Supplier Quality Assurance and other Supplier Issues 4
D How to use the Supplier Rating Calculator Supplier Quality Assurance and other Supplier Issues 2
S Supplier Quality Rating - What to evaluate Supplier Quality Assurance and other Supplier Issues 11
Chennaiite PFMEA rating - Occurence Rating and Detection Rating - Some clarification FMEA and Control Plans 15
A Vendor Rating System - IS:12040-1987 (India Standard) - Where to buy ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3