Hi All,
Hoping for a little help on the best route forward for a reliable roundness measurement.
We are producing two PC rings sample #1 & #2 (Dia both around: 14" (+/-0.01" Roundness: <0.016)
Measurement device: Micro VU + touch Probe (also done same study with CMM)
Roundness: 16 points. Least Square Circle method.
Sample #1 - Ring has Internal support
Ring placed on fixture, location is positioned in x,y
Diameter pass GRR
Roundness requires part to be measured 3x and average taken - Passes GRR
Sample #2 - No center support
Diameter passed GRR
Roundness measured 3x and average taken - Can not pass.
I believe that the variation, is caused due to product loading into fixture, and in-part-variation. Due to if I repeat the measurement without replacing the part the results are incredibly consistent.
I know the clear thing may be looking to fixate the part better but I am wary to restrain the part too much to effect the result and i feel the part is incredibly sensitive to touch and any forces applied on it.
How do other people measure roundness on flexible parts that have in-part-variation?
Is 16 points too little based on my diameters?
Any advise would truly be appreciated.
Thanks,
John Predmore
You say flexible part. I don't know what PC is, polycarbonate? Without knowing anything about your part beyond what you said here, yes, fixturing a part or even the weight of the part could cause a measurable deflection. In your case, I think the part should be laid flat, unfixtured, on a flat surface and use non-contact methods. It sounds like that is where you are at. Sixteen points is a good number to start, but if the measurement is not repeatable, then more might be needed. If you get a repeatable measurement in repetitions without moving the part, but measurements vary once the part is moved, that seems to indicate the method is capable but roundness changes when the part is moved. How can you handle the part without the forces of handling the part? Can you float the part on a liquid, then drain the liquid, so the part comes to rest with no external forces causing deflection? Can you float the part into position on an air table?
Matt's Quality Handle
How is the part used in its intended application? As a matter of practice, on flexible parts, I would always recommend fixturing in a similar manner to actual use conditions.
With the disclaimer that the drawing must explicitly allow constraining for measurement purposes. IIRC, ASME 14.5 says that dimensions apply in the free state at 68 degrees F, unless otherwise noted.
B Setting flexible packaging specification limits ISO 14971 - Medical Device Risk Management 4
T Gage R&R Study- Measuring a flexible part (compressed part) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P IEC 60601 Clause 57.3 / 8.11.3.2 - PVC Flexible Cords IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F Flexible Quality Management System (Multiple Standards) ISO 13485:2016 - Medical Device Quality Management Systems 8
B Flexible & OOR O.D. Measuring - "Richard's Gage" for measuring flexible items Quality Manager and Management Related Issues 1
C Procedures or Work Instructions on Roll Flagging for a flexible packaging company Document Control Systems, Procedures, Forms and Templates 5
S Lean - Ways to Measure Equipment Flexibility - Flexible Manufacturing System (FMS) Lean in Manufacturing and Service Industries 2
B How do I set up GMP's for a flexible packaging converter? Manufacturing and Related Processes 13
B How do I begin implementing HACCP? Flexible food packaging Food Safety - ISO 22000, HACCP (21 CFR 120) 3
C QMS Performance Trends - Based on fixed or flexible periods - Clause 5.6.1.1 IATF 16949 - Automotive Quality Systems Standard 6
Marc Boeing's New 787 Wings are Amazingly Flexible World News 4
C Flexible Accreditation Scopes - Electrical safety testing to EN standards General Measurement Device and Calibration Topics 2
K Medical Devices QC Testing Procedure - Flexible Medical Solution Bags ISO 13485:2016 - Medical Device Quality Management Systems 5
Q Measuring flexible materials like sponge, epdm, and felt General Measurement Device and Calibration Topics 8
Gabble FMEA start point - should it be flexible? FMEA and Control Plans 2
S 17025 Flexible Scopes of Accreditation - Laboratory to 'self-assess' new test method ISO 17025 related Discussions 2
R Cpk calculation for short runs - Manufacturing FFC (flexible flat cable) Capability, Accuracy and Stability - Processes, Machines, etc. 3
P Flexible controlled software method? Software Quality Assurance 2
S What is the best method for checking the flatness of a flexible part? General Measurement Device and Calibration Topics 4
S Comparing measurement results for a part from two ring gage General Measurement Device and Calibration Topics 3
I Class VI certification on seals, o-ring and plastic for a WFI system Qualification and Validation (including 21 CFR Part 11) 0
J Using ring gauges General Measurement Device and Calibration Topics 2
D O Ring capability and measurement - What is the automotive 'norm' for capability studies on O Rings? General Measurement Device and Calibration Topics 3
C Thread Tolerance - Can I use a 3A Ring on a part that the drawing calls out for a 2A? Manufacturing and Related Processes 5
S ISO 5356-1 - 22 mm ring gauges and test laboratories for an Anesthetic connector Other Medical Device Related Standards 1
S Ring Gage Thermal Expansion General Measurement Device and Calibration Topics 4
S Minor diameter of threaded ring gage General Measurement Device and Calibration Topics 1
S Poka Yoke Inspection Retaining Ring (C-Clip) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
S Interesting Discussion Threaded ring gage calibration and tendonitis in the elbows General Measurement Device and Calibration Topics 8
S Thread Ring Gauge verification General Measurement Device and Calibration Topics 6
M Calibration of Thread Plugs and Ring Gages - Various Thread Standards General Measurement Device and Calibration Topics 3
S Clarification of threaded ring gage calibration procedure/requirements General Measurement Device and Calibration Topics 2
S Looking for a Threaded Ring Gage Calibration Laboratory General Measurement Device and Calibration Topics 13
C O-Ring Groove Measurement Method in CNC Machined Parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
Q Is it normal practice to lubricate Ring Gages before use? General Measurement Device and Calibration Topics 17
GURU SDI Leak Testing - O-ring installation in the groove of Bore ID Problem Solving, Root Cause Fault and Failure Analysis 4
C Manufacturing Thread Ring Gages Manufacturing and Related Processes 3
C Measuring Solid Thread Ring Gages General Measurement Device and Calibration Topics 17
P Best Method of Cleaning of Thread Ring Gages Calibration Frequency (Interval) 5
P No-go Ring Over the specifiaction limit Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
kedarg6500 Availability of 5/8" 17.75 TPI Thread Ring Gauge & Die Nut Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
P How to add Voip features (ring group, virtual receptionist) into Cisco SIP PBX in C#? Medical Information Technology, Medical Software and Health Informatics 1
V How to Calibrate Ring Force Gauges Service Industry Specific Topics 1
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
M Thread Ring Gauge Specifications General Measurement Device and Calibration Topics 3
K Calibration Requirements for Thread and Ring Gages AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
S Thread Timing - Ring Gage Frustration General Measurement Device and Calibration Topics 11
C Test Uncertainty Ratio (TUR) for Ring Gage Calibration Measurement Uncertainty (MU) 7
Z Error of 30 microns checking a Ring Gauge on a CMM Machine Inspection, Prints (Drawings), Testing, Sampling and Related Topics 15
C Adjustable Thread Ring Gage Calibration General Measurement Device and Calibration Topics 8