We are currently dealing with an issue involving the order that steps in a Work Instruction are completed. Obviously there are some steps that must be completed in the exact order (remove part from cast THEN clean cast) But others that could be of any order (Clean part / Clean empty cast). It has been written in the procedures governing the WI that they must be done in order unless written otherwise. However I still haven't found a work instruction that uses that exemption.

This has created an issue. How do you handle yours? How exact do we need to be? Can't logic rule?

Note: This is for medical devices
 
Elsmar Forum Sponsor

Project Man

Involved In Discussions
#2
It appears as if the simple answer is to update the work instructions to use the exemptions. But, I think the bigger issue may be that the procedure allows it but the method of achieving it is not clear (i.e. bracketing the steps that can be done in any order).
Logic can rule but not everyone thinks logically hence, clear work instructions.
 

Ninja

Looking for Reality
Trusted Information Resource
#3
When dealing with both cases (sometimes order matters, sometimes not), play to the greater population.

In most of my procedures, order matters. All procedures and WI have numbered steps to be followed in order.
Sometimes (rarely) order doesn't matter...those sub-instructions say "The following steps can be done in any order"...and they are bullets, not numbered.

1. "Do this"
2. "Do that"
3. "The following may be done in any order
- thing
- thang
- this
- that
4. "do the next thing"
5. "do one more thing"

Logic ruling is pretty dependent on your employee base. That's a call for you to make (with consideration to medical requirements). Not everyone's logic is the same...but sometimes your culture fosters a working base level logic that works.
 

Candi1024

Quite Involved in Discussions
#4
Yes, I do agree with both of you. The problem is, we have approximately 1000 different work instructions. Up until this past year, not a great amount of focus has been put on being very specific in having correct order, or exemptions of order.

This is definitely something we will work towards though. However it's going to take quite awhile, and we have this classified as a CAPA which is already in the effectiveness check stage. They did one round of redlines, but I think there is still a lot of work left to do. Woo hoo!
 

Ninja

Looking for Reality
Trusted Information Resource
#5
...This is definitely something we will work towards though. However it's going to take quite awhile, and we have this classified as a CAPA which is already in the effectiveness check stage.
A great reason to play toward the majority...first step is to identify how many would actually get updated...it shouldn't have to be all of them. You might get lucky and only do 100.

If you've decided to make changes, and the changes aren't made yet...you might want to rethink what stage your CAPA is...should be easy enough to back it up to a work-in-process.

Never lose sight that the QMS is there to support you, not burden you. It is work...but choose carefully via risk assessment how high of a priority this gets. Are order of instruction mishaps causing significant damage daily?
 

Candi1024

Quite Involved in Discussions
#6
A great reason to play toward the majority...first step is to identify how many would actually get updated...it shouldn't have to be all of them. You might get lucky and only do 100.

If you've decided to make changes, and the changes aren't made yet...you might want to rethink what stage your CAPA is...should be easy enough to back it up to a work-in-process.

Never lose sight that the QMS is there to support you, not burden you. It is work...but choose carefully via risk assessment how high of a priority this gets. Are order of instruction mishaps causing significant damage daily?
No, we don't have any significant damage. However, it was found during an audit, then second audit, that we were not following procedure. At no time did this cause us quality issues. All audit findings become a CAPA. All CAPAs become all encompassing and overextending with a 1 year time limit. Aren't large companies grand? :sarcasm:
 
Last edited by a moderator:

Ninja

Looking for Reality
Trusted Information Resource
#7
At no time did this cause us quality issues.
Can't you just handle it in an overriding procedure by adding something like:
"Work instructions are not intended to dictate order of operations within a task"

If it is a paperwork chase, and adds no value to you, keep the solution as simple as you can and move onto something important.
 

Candi1024

Quite Involved in Discussions
#8
Can't you just handle it in an overriding procedure by adding something like:
"Work instructions are not intended to dictate order of operations within a task"

If it is a paperwork chase, and adds no value to you, keep the solution as simple as you can and move onto something important.
I've been trying to push that. It's not my project, I'm just doing one small part of it. However after my check I'll be making some recommendations based on what I saw, and that will be included.

The argument is that in some cases, (some say more than the majority of steps) it really does matter.
 

Ninja

Looking for Reality
Trusted Information Resource
#9
The argument is that in some cases, (some say more than the majority of steps) it really does matter.
Matters so much that you've never had an issue.
Doesn't sound like something I'd want to spend a lot of time on.

Good luck with your sales pitch!

Next thing I would look at is whether it would take me more time to change the documents, or to convince others that it isn't worth doing.
At that point you are outside of QMS and into personality management.
Sometimes it's just easier to do something useless than to avoid it.
 
Thread starter Similar threads Forum Replies Date
M Change Order Process - Stake holders not following through - Suggestions wanted Manufacturing and Related Processes 5
T Root Cause Failure Analysis - Not following Customer packaging Specification Problem Solving, Root Cause Fault and Failure Analysis 5
M Informational The USFDA Continues to Remind Facilities of the Importance of Following Duodenoscope Reprocessing Instructions: FDA Safety Communication Medical Device and FDA Regulations and Standards News 0
M Medical Device News Brexit - Withdrawal Agreement and Political Declaration laid before Parliament following political agreement. EU Medical Device Regulations 4
D Design FMEA for a component - Should I make the following assumptions? FMEA and Control Plans 7
S. Moore Record Retention - How long must a company keep the following records? Records and Data - Quality, Legal and Other Evidence 17
S Supplier Classifications - Please give me a good definition of the following terms ISO 13485:2016 - Medical Device Quality Management Systems 5
K Nonconformance on training - Not following own processes (IATF 16949) Internal Auditing 14
P Is the next revision of ISO 15378 following the High Level Structure? Other ISO and International Standards and European Regulations 5
M Please explain the following Automotive Acronyms IATF 16949 - Automotive Quality Systems Standard 3
M CAPA due to Consumer not following Maintenance Instructions... Nonconformance and Corrective Action 9
Crimpshrine13 ISO/TS 16949 CBs & Auditors not following up on the schedules IATF 16949 - Automotive Quality Systems Standard 43
R Which of the following indicators is important for Operations - Ppk, Cpk, or Cpm? Statistical Analysis Tools, Techniques and SPC 15
T What to do when Employees are not following Instructions Human Factors and Ergonomics in Engineering 65
F ANSI ASQ Z1.4 - Is the following process acceptable? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
M QSR - Design Control on Existing Device not Designed following QSR 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
R QA Manager and Disciplining Employees for not following Processes Quality Manager and Management Related Issues 28
I Medical Device Registration details for the following Countries ISO 13485:2016 - Medical Device Quality Management Systems 5
M Sampling Inspection at QA - Inputs on the following sampling plan Statistical Analysis Tools, Techniques and SPC 3
N My FDA auditor asked the following: Other US Medical Device Regulations 12
Q How to determine the average baseline for the following data in SPC Quality Tools, Improvement and Analysis 3
M 7.6.2 Revisions following Engineering Changes - What is the right meaning? IATF 16949 - Automotive Quality Systems Standard 4
Marc Software gives visual representation of who?s following you online After Work and Weekend Discussion Topics 1
Ronen E European Commission calls for action following PIP's Breast Implant Saga EU Medical Device Regulations 1
T Employee not following Purchasing Process Other Medical Device Related Standards 27
Marc US east coast in chaos following rare October snowstorm - 2011 Coffee Break and Water Cooler Discussions 14
T Is the following flow for Training Needs Analysis correct? Training - Internal, External, Online and Distance Learning 6
L Consultant not following Process Approach ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 34
Q Can I use Regression Analysis in the following scenario Statistical Analysis Tools, Techniques and SPC 18
quality1 Supplier Disqualification - High Rejects and not following established Procedures ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
D Manufacturing Execution Systems (MES) - People Not Following Work Process (WPs) Manufacturing and Related Processes 5
L Fitted Splined Connections following DIN 5480 - Inspection Problems Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
eternal_atlas Not following the Procedures is a Nonconformance? Sales department General Auditing Discussions 61
H Meeting following an interview, what should I expect? Career and Occupation Discussions 15
T Labeling Control and trouble with labelers not following procedure ISO 13485:2016 - Medical Device Quality Management Systems 11
Ajit Basrur Teva recalls drug following Jackson death probe Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
A Audit NC (nonconformance) for not following guidelines on calibration! General Measurement Device and Calibration Topics 9
S Consequences of Shipping product with label not following CE rules EU Medical Device Regulations 2
Q Corrective Action - CAPA - Root Cause - Not following the training procedure! Problem Solving, Root Cause Fault and Failure Analysis 13
S Why people not following SPC charts regularly Training - Internal, External, Online and Distance Learning 10
G Not sequentially following steps in a work instruction Process Audits and Layered Process Audits 85
Q Work Environment - Which of the following is to be addressed as nonconformity? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 25
Z How to Establish Quality Manual Not Following the Structure of the Standard? Quality Management System (QMS) Manuals 3
D How do we interpret the following XmR Trend Chart data? Statistical Analysis Tools, Techniques and SPC 3
L Corrective action following a wrong answer of the auditee ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
T In the following statement, what does "Z" stand for? ISO 13485:2016 - Medical Device Quality Management Systems 7
Q Failure where the Root Cause was the Material Handler Not Following Procedure Problem Solving, Root Cause Fault and Failure Analysis 20
Q Personnel Not Following Procedures - How do you write up an audit finding? Internal Auditing 43
A Non-conformance logged following a TS16949 Stage 2 audit - APQP Control Plan IATF 16949 - Automotive Quality Systems Standard 15
S New PPAP on Sale - Includes the following changes IATF 16949 - Automotive Quality Systems Standard 2

Similar threads

Top Bottom