For an MDI, is leak-test a stability parameter ?

v9991

Trusted Information Resource
#1
While establishing the product stability of a metered dose inhaler, would you consider leak-test to be part of stability program?

My suggestion, is that it is to be considered as part of stability program...(because, the formula assumes that the rate-of-leak (as measured at T0 time point) is constant; )

If it is not part of normal stability program; the leak test need to be established a part of design-verification of device. (simulate all the stress-conditions of product shelf life)
 
Last edited by a moderator:
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
While establishing the product stability of a metered dose inhaler, would you consider leak-test to be part of stability program?

My suggestion, is that it is to be considered as part of stability program...(because, the formula assumes that the rate-of-leak (as measured at T0 time point) is constant; )

If it is not part of normal stability program; the leak test need to be established a part of design-verification of device. (simulate all the stress-conditions of product shelf life)
Hi,

I'm not 100% sure what are your concerns, but anyway -

If you develop a device that you suspect is subject to a leak mechanism (as a normal characteristic or as a failure mode), and the leak creates a concern over you device's performance, then yes, you probably should, as part of your design verification / shelf-life studies, establish the leak characteristic under typical and extreme shelf-life and use-life stresses;

If on top of that you are also concerned with the stability of a drug that is supposed to work with your device, and you suspect that this stability is affected by the above characterized leak mechanism, then in my opinion it would make sense to test this stability under the worst-case leak scenario over the entire applicable product life (device / drug / shelf / use etc.).

Cheers,
Ronen.
 
Last edited:

v9991

Trusted Information Resource
#3
here's specific reference citing the need for establishing leak test over stability...

The leak rate test is important in stability studies because it may provide information on pressure loss and may predict, at subsequent test stations, failures in testing for dose content uniformity through container life (see section III.F.1.j). It should be noted, however, that leak rates are not necessarily constant over time.
4.2.1.16 Performance after temperature cycling (CTD 3.2.P.2.4)
A study should be conducted to determine the effect of temperature cycling on the performance of the product. Containers should be stored in various orientations and cycled between recommended storage conditions and a temperature below freezing (0°C). For suspensions, cycling between the
recommended storage conditions and a high temperature (e.g., 40°C) should be considered, and may be combined with the low temperature cycling study. Storage time should be at least 24 hours under each condition, and containers should be stored under each condition at least five times. The containers should be examined visually for any obvious defects, and tests such as leak rate, weight loss, delivered dose uniformity, fine particle mass, related substances and moisture content should be performed. Any changes from initial results should be assessed for their significance.
when there is point of decision, where ther's difference in expertise... it all comes down to references...:eek:
 

Attachments

Thread starter Similar threads Forum Replies Date
S Capability or Gage R&R Study for Leak Tester? Reliability Analysis - Predictions, Testing and Standards 15
N MSA Study for a Leak Testing Device Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
M Proper ISO standard for Pressure Testing and Leak Rates Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
H Validating Nondestructive leak testing machine Qualification and Validation (including 21 CFR Part 11) 3
F Leak Tester Gage R&R Sheet Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
triathlonx13 Leak test correlation - Bubble to Air Reliability Analysis - Predictions, Testing and Standards 8
O Capability or Gage R&R Study for Leak Tester Reliability Analysis - Predictions, Testing and Standards 11
GURU SDI Leak Testing - O-ring installation in the groove of Bore ID Problem Solving, Root Cause Fault and Failure Analysis 4
L Environmental Gas Leak Detection Equipment Recomendations Miscellaneous Environmental Standards and EMS Related Discussions 2
T Would a Leak be a Failure Mode on a Design FMEA? Design and Development of Products and Processes 2
GURU SDI Measurement - Pressure Drop vs. Leak Rate Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
S Destructive Attribute MSA for Simple Water Leak Check in Trailers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
somashekar Why does a tap leak... Funny Stuff - Jokes and Humour 2
B Validation of a Pressure Decay Leak Tester for a Medical Device Design and Development of Products and Processes 2
Claes Gefvenberg Massive gas leak off Scotland World News 12
H Determining Calibration Tolerance for new Leak and Occlusion Equipment Measurement Uncertainty (MU) 2
N Tyvek Pouch Seal - Pressure needed to conduct Vacuum Seal Leak test? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Zyglo Fluorescent Dye for Leak Testing - Sloooooow acting - Seeking advice Reliability Analysis - Predictions, Testing and Standards 4
R Sample Size for P Chart - 100% Leak Test for Condensor Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
R Seal Packaging - Burst and Leak Specifications for a Foil Seal applied to a PTFE well Manufacturing and Related Processes 4
T Gage Repeatability for an Automated Leak Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
W Oven Leak Test Method - Integrity of the chamber to hold gas Qualification and Validation (including 21 CFR Part 11) 3
D Leak Test Method for Balloon Catheters EU Medical Device Regulations 2
B MSA (Gage R&R) vs. Capability Cpk Study - Helium Leak Check Station Statistical Analysis Tools, Techniques and SPC 2
D Halogen Leak Detector - Calibration source needed General Measurement Device and Calibration Topics 1
G Leak Tester Calibration - Sentinal C-20 controller with a calibrated orifice General Measurement Device and Calibration Topics 1
G Leak Test - Building a master pump to calibrate pump tester Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
ScottK Oxygen Cleaning and Surfactants used for leak detection Various Other Specifications, Standards, and related Requirements 1
T Calibration of vacuum type leak detectors manufactured by Vacuum Technology Inc General Measurement Device and Calibration Topics 5
H GR&R (Gage R&R) on leak test equipment is there a deviation requirment? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
R Calibrating Leak Testers - Old fashioned water tank type tester General Measurement Device and Calibration Topics 2
C Master Validation Frequency in production - Absence of leak, thread presence, etc. Reliability Analysis - Predictions, Testing and Standards 4
W Leak Detection - Oven Seals General Measurement Device and Calibration Topics 1
M Developing a Leak Test - Housing part for an automotive customer Various Other Specifications, Standards, and related Requirements 7
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
E Tensile test machine, tinius olsen 602 Manufacturing and Related Processes 1
M Test failure Root cause not found Customer Complaints 10
Moumen H Variations between ASTM A29 Standard for steel bars and Mill test certificates specs Manufacturing and Related Processes 1
D Tip test 10° -> 5° IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Sponsor name presented in the test report EU Medical Device Regulations 4
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2