Frequency and sample size for lot release of EtO sterilized products

shimonv

Trusted Information Resource
#1
Hi everyone,
I would like to do some benchmarking with you:
I was given to understand that EtO residuals and biological indicators are tested for each lot and bio-burden and LAL are done periodically (quarterly or after a certain number of manufactured LOTs).

What is your experience with that and what sample size to you use (I suspect there is no explicit guidance on that)?

Thanks,
Shimon
 
Elsmar Forum Sponsor

Cthames

Involved In Discussions
#2
Hello Shimon,

From my experience (this was for one company, not sure how others do it), ETO residuals are validated and not measured every lot. This validation usually is the final piece of a sterilization validation and is repeated when new risks are introduced or every 1-2 years. Granted, the company I was working for had 600 SKUs which fit into 4 finished product families. We didn't specifically validation individual SKUs, so I'm not sure what the standard practice is for that.

We used 34 biological indicators with each sterilization run of 14 pallets of medical devices. We placed these inside worst cases devices (within pouches) and had the sterilizer place them in strategic locations within the sterilization chamber. The consultant (who worked closely with the sterilizer) came up with this sample size. I don't have access to that validation anymore, but I believe it was based off a combination of how many biological indicators we tested during our validation runs, and how many strategic locations we had to account for (gas entry points, gas exit points, center of room, hardest to reach locations, etc).

There may be guidance in ISO11135:2014, but I cannot remember.
 

chris1price

Trusted Information Resource
#3
Hi, as Cthames said, you validate the residuals, so there is no requirement for monitoring each lot. However, I am aware that some companies do perform this on every lot.

I do not think there is any standard for LAL frequency, I have seen it done for every lot, once per month and once per quarter. I suspect this is very dependent on the risk posed by the product and the manufacturing processes.

Bioburden is again validated, so monthly should be sufficient, there is nothing in the standard but 10 samples per quarter would match the frequency used for gamma sterilization. ISO11737-1 A.8.1 says "It is common practice to use a sample size of between three to ten items for routine monitoring of bioburden levels."

ISO11135, table C.3 lists the number of BIs to use. for routine monitoring of N cubic meters of load space over 10m3, it is (30+(N-10))/2
 
Thread starter Similar threads Forum Replies Date
H How to calculate sample size and sample frequency Statistical Analysis Tools, Techniques and SPC 2
Y FMEA link to Control Plan frequency and sample size FMEA and Control Plans 20
P Frequency and Sample Size Requirements for an MSA Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
B Customer Inspection Frequency and Rejection Level Criteria - Sample Lot Size AQL - Acceptable Quality Level 6
L Criteria for Establishing Control Plan Sample Size and Inspection Frequency FMEA and Control Plans 7
Manix GM's GP12 Increased Frequency/Sample Size - B1 of the plan development section FMEA and Control Plans 8
A Sample Size and Frequency - In-process monitoring in a mass production shop Statistical Analysis Tools, Techniques and SPC 2
N What frequency and sample size is 'normal' for SPC? Statistical Analysis Tools, Techniques and SPC 4
G Determining Sample Size and Frequency for Features on Control Plans FMEA and Control Plans 8
B Sampling Plan to determine sample size/frequency based on a Given Rate of Production Inspection, Prints (Drawings), Testing, Sampling and Related Topics 40
R How to determine the correct Sample Frequency for SPC Features Statistical Analysis Tools, Techniques and SPC 66
P Acceptable quantity levels/Sigma levels for reducing the frequency of quality tests. Six Sigma 7
A SaMD: Metric for release frequency EU Medical Device Regulations 5
Q Control Plan and the frequency of sampling FMEA and Control Plans 11
M PMCF Frequency EU Medical Device Regulations 32
S AMS 2769 - weekly frequency test AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
N MSA Frequency related literature Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
B CQI-15 Welding Health Chart Frequency IATF 16949 - Automotive Quality Systems Standard 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M Bioburden Frequency (sterile MD) Other Medical Device Related Standards 2
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
R Procedure, Frequency and Acceptance Criteria for Replicate, Recalibration, Before-After and Intermediate Checks ISO 17025 related Discussions 8
qualprod To set frequency to review documents in ISO 9001 7.5? Document Control Systems, Procedures, Forms and Templates 13
M Defining frequency of measurement tools callibration Calibration and Metrology Software and Hardware 3
M Define voltage and frequency to perform tests 61010-1 and 61326-1 for CE certification CE Marking (Conformité Européene) / CB Scheme 4
L Calibration Frequency for Surftest SJ-400 General Measurement Device and Calibration Topics 0
D Determining Calibration Frequency schedule for items used in production Manufacturing and Related Processes 2
G ISO 9001 - 9.3.1 Management Review - Content and Frequency Management Review Meetings and related Processes 12
H Frequency of shelf life/sterilization/bench studies conducted on a medical device Other Medical Device Related Standards 1
M RFID (Radio Frequency Identification) Registration in Europe and in MENA countries EU Medical Device Regulations 1
M Leakage current measurement with non-frequency-weighted device IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
S Requirements for Interval Measurement test & Frequency Response test clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B Radiation Sterilization Dose Audit: Bioburden frequency Other Medical Device Related Standards 3
bobdoering SPC Charting frequency calculator for X hi/lo chart Imported Legacy Blogs 0
D IATF 16949 surveillance audit frequency IATF 16949 - Automotive Quality Systems Standard 16
C AS9100 rev D 8.5.1 c 2 - Defining the Machine in-process frequency per ANSI/ASQ Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
J 95% reliability for calibration frequency Calibration Frequency (Interval) 7
R Internal Audit Frequency - Remote Site (a Design Center) IATF 16949 - Automotive Quality Systems Standard 3
V Periodic Audit Trail Review - Scope, Content & Frequency Qualification and Validation (including 21 CFR Part 11) 13
JoCam Required Frequency of Notified Body reviews of Technical Files EU Medical Device Regulations 6
Q AMS 2750E System Accuracy Test (SAT) Frequency AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
B MHRA Unannounced Audits Frequency Changed? EU Medical Device Regulations 2
P IATF 16949 - Internal Audit Frequency Requirements Internal Auditing 2
Q Objective and Goals - ISO 9001:2008 Frequency Revision Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Q RFID (radio frequency identification) registration for Medical Device Other Medical Device Regulations World-Wide 6
G Sequence and Frequency of Calibration for Thread Gauges General Measurement Device and Calibration Topics 2
N Tool frequency change - Data Analysis needed Statistical Analysis Tools, Techniques and SPC 9
GStough Adequately Defining Which Suppliers to Audit and Frequency Supplier Quality Assurance and other Supplier Issues 8

Similar threads

Top Bottom