Gage Calibration non QS-9000 Section

B

B. Maynard

#1
Good Morning,
We will be audited in Dec. 1999 for initial QS-9000 Registration. We have approximate 300 gages of which 225 are used in the mold making area which is not part of our QS-9000. We are only going for QS in the Final Product Production from the molds.
My question is !
Do we have to calibrate the gages in mold making side since it will not be involed in the QS Registration. They are highly skilled employees.
Upper management has decided that they should be labeled for referance only. The adjustment of Final Product is a Process problem and we will calibrate those gages that affect the Final Product.
I say all should be Calibrated but Management says "no".
Am I right or can we do what the Team says.
I have been through QS before and they have not.
Thanks,
The Necessary Evil 'Quality Assurance Mgr.)
Bill Maynard
 
Elsmar Forum Sponsor
D

Debra Woodford

#2
Hi Bill, Well what I read in 4.11.2b identify all i,m,& t equipment that can effect product quality, and calibrate and adjust them at perscribed intervals, or prior to use yada yada yada, with the *NOTE i,m, & t equipment includes equipment in tooling departments used to qualify or maintain production tools regardless fo ownership. Sounds pretty clear to me so I'm with you however I understand the "upper management" short cut thinking. Good Luck
 
D

Debra Woodford

#3
Hi Bill, Well what I read in 4.11.2b identify all i,m,& t equipment that can effect product quality, and calibrate and adjust them at perscribed intervals, or prior to use yada yada yada, with the *NOTE i,m, & t equipment includes equipment in tooling departments used to qualify or maintain production tools regardless fo ownership. Sounds pretty clear to me so I'm with you however I understand the "upper management" short cut thinking. Good Luck
 
T

TheOtherMe

#4
If it's ouside your registration scope, you do not need to do anything. It is off limits to the auditors.

----------

On the other hand, if you take a look at the T&E Supplement to QS9000 (Second edition), you will find it says (although the AIAG will probably want to sue me for quoting verbatum from the T&E Supplement): "4.11.4 Measuring System Analysis - NOTE: All hand held variable gages that do not require setup are exempt from this requirement (e.g., micrometers, calipers and height gages). These gages still require gage identification, calibration and traceability."

So, it's saying that yes - they have to be calibrated but you don't have to do Gage R&R and such. (Duh - gage R&R and measurement studies are only appropriate for production scenarios, anyway...)

If it was me, I'd have them all calibrated and in the cal database but to me that's common sense. However - the money issue rears its ugly head. And the money issue is real. For every time someone says "That's common sense" - well, let's spend some more money. I personally am particularly sensitive to the need for calibrated M&TE and the need for preventive maintenance - they have my votes. But then there's the production folks who believe they can save some serious bucks if all the lines are upgraded with some poke-yokes here and there. And the accounting folks. And there's the design group which says it will profit 2 more SPARC stations and two more software seats so they don't have to share. I know calibration is a fixed expense (should be about the same year to year unless the business is growing rapidly) as opposed to the other examples, but I hope you catch my drift... Damn! I'm glad I'm not upper management! I guess that's why they make the big bucks - big decisions.

----------

Bottom line: It's outside the registratiuon scope so it doesn't matter - the auditors have no reason to even visit that area.

I have had clients who only did the QS9000 dance on automotive lines so in the same room you had some equipment which was in the scope and some not. At one client's audit, when the auditor asked about a line or piece of equipment which wasn't in the scope, we dragged out our scope statement and showed the auditor that it was not in the scope and we moved on to equipment which was listed in the scope statement. Gages not in the scope were IDed as inventory by serial number and such but the procedure read that gages with no cal sticker would not be used in QS areas. Since all the gages in the QS areas were in the system, had a cal sticker and all the rest, there was no problem.

[This message has been edited by TheOtherMe (edited 25 August 1999).]
 

Howard Atkins

Forum Administrator
Leader
Admin
#5
We will be audited in Dec. 1999 for initial QS-9000 Registration. We have approximate
300 gages of which 225 are used in the mold making area which is not part of our
QS-9000. We are only going for QS in the Final Product Production from the molds.
___________________________________

Do you expect to get ISO 9000 for the mold department?
I think that even if the mold department is not included i.e. the manufacture, the audit will refer to maintenance of the molds and
the interface between injection and molding. The molds are also in some case customer owned and thus have to be treated as such as per customer supplied material.
___________________________

My question is !
Do we have to calibrate the gages in mold making side since it will not be involed in
the QS Registration. They are highly skilled employees.
------------------------------
I do not see the relevance of their qualifications. If they were not you would not employ them. How does their skills effect the reliability etc. of the gages.

------------------------------
Upper management has decided that they should be labeled for referance only. The
adjustment of Final Product is a Process problem and we will calibrate those gages that affect the Final Product.
-------------------------
IMHO calibration is necessary for gauges that are used to determine the quality of the product. If you can show that the
gauge is not a factor in the determination of final customer demands but just an aid to production I think you would not have
to calibrate them.
----------------------------------
I say all should be Calibrated but Management says "no".
Am I right or can we do what the Team says.
I have been through QS before and they have not.

______________________________________
The team is not necessarily right
QS9000 4.11.1 " The supplier shall ...........to demonstrate the
conformance of product to the specified requirements"
 
Thread starter Similar threads Forum Replies Date
D Gage Calibration - Non calibrated gages are not used for final dimensions? General Measurement Device and Calibration Topics 3
B Gage Calibration non QS Area 2nd Topic QS-9000 - American Automotive Manufacturers Standard 3
B Gage calibration Gonogo (calcul of cgk) General Measurement Device and Calibration Topics 0
H Air Pressure Gage Calibration General Measurement Device and Calibration Topics 6
C Gage Block Grade 1 - Calibration Measurement Uncertainty (MU) 15
Q Gage calibration in disarray ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
S Interesting Discussion Threaded ring gage calibration and tendonitis in the elbows General Measurement Device and Calibration Topics 8
R Master Gage Block Calibration (We have 2 sets) question General Measurement Device and Calibration Topics 3
S Clarification of threaded ring gage calibration procedure/requirements General Measurement Device and Calibration Topics 2
S Looking for a Threaded Ring Gage Calibration Laboratory General Measurement Device and Calibration Topics 13
S Calibration of a Coating Thickness Gage General Measurement Device and Calibration Topics 3
D Calibration of a Width Gage Slot with an Attribute Gage - Maximum Width Condition General Measurement Device and Calibration Topics 16
D Gage Calibration Tracking in MS Excel Calibration and Metrology Software and Hardware 9
K Height Gage Compensation Block Calibration General Measurement Device and Calibration Topics 1
K Gage Calibration - Corporate Scheme IATF 16949 - Automotive Quality Systems Standard 3
C Black Oxide Gage Pin Calibration Requirements General Measurement Device and Calibration Topics 4
X When a Gage is past the Calibration Due Date General Measurement Device and Calibration Topics 6
T Gage Block Calibration Requirements question General Measurement Device and Calibration Topics 19
C How to Clean Gage Blocks and Gage Pins prior to Calibration General Measurement Device and Calibration Topics 4
I Gage Block Calibration/Measurement Control Question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
M Trusting Gage Manufacturer Calibration Certifications General Measurement Device and Calibration Topics 3
K Qualifications I need to adhere to for Thread Gage Calibration Calibration Frequency (Interval) 1
D New to Gage Block Calibration Reports Measurement Uncertainty (MU) 6
W Setting up a Zmike for Pin Gage Calibration General Measurement Device and Calibration Topics 3
M Thread Gage Calibration - Who to use General Measurement Device and Calibration Topics 5
C ZZ Pin Gage Tolerances and Calibration General Measurement Device and Calibration Topics 2
S Pin Gage Calibration - Pins seem to be "growing" General Measurement Device and Calibration Topics 9
S Purchase Order requirements Gage Blocks sent out for Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
K Pin Gage Use vs. Calibration Frequency Calibration Frequency (Interval) 14
S Gage Calibration Requirement and Frequency/Extended Frequency ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
G Gage Recall Analysis - Threaded Plug Gage has failed Calibration General Measurement Device and Calibration Topics 10
M Calibration Procedure for Gage Blocks General Measurement Device and Calibration Topics 10
C Test Uncertainty Ratio (TUR) for Ring Gage Calibration Measurement Uncertainty (MU) 7
S Mechanical/Electrical Forces Gage Calibration Methods Calibration and Metrology Software and Hardware 1
M Calibration Cycle for Standard Iron & Steel Plate Gage General Measurement Device and Calibration Topics 17
A Grade 1 Block Gage Calibration Master General Measurement Device and Calibration Topics 2
C Gage and Calibration Software Comparisons Calibration and Metrology Software and Hardware 6
A Calibration Standards - Rod Standards vs. Gage Blocks Calibration Frequency (Interval) 2
M Thread Plug Gage - Calibration Frequency ISO 13485:2016 - Medical Device Quality Management Systems 1
C Adjustable Thread Ring Gage Calibration General Measurement Device and Calibration Topics 8
L Calibration Procedure for Internal Spline Plug Gage Needed General Measurement Device and Calibration Topics 4
V Gage Block Calibration Frequency Requirements Calibration Frequency (Interval) 20
B Temperature Gage on an Oven was Out of Calibration Period General Measurement Device and Calibration Topics 15
F Process Monitoring Gage Calibration Requirements - 7.6.3.2 External laboratory IATF 16949 - Automotive Quality Systems Standard 1
R Calibration of a portable FARO gage portable CMM # F04L2 gauge General Measurement Device and Calibration Topics 6
J In-House Calibration Requirements - Use of Grade 2 set of gage blocks General Measurement Device and Calibration Topics 2

Similar threads

Top Bottom